Author Topic: Handheld meter robustness testing  (Read 1169235 times)

Jon.C and 12 Guests are viewing this topic.

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1050 on: October 24, 2016, 10:42:30 pm »
A message body...

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1051 on: October 30, 2016, 04:47:41 pm »
Sitting here is a new in the box AIMO MS8211. It's marked 600V CAT III and is similar to the MS8211D in that it's a pen type as well.   It has a back light, non-contact voltage detector and even a little flash light in the tip.   It does not have the logic probe like the D model.   Unlike the D model, they also use a mechanical 250mA fuse rather than the PTC to protect the current input.  They claim the diode test is 1mA at 2.8V.  This would also be a step up from the D model.   

Keep in mind, like the D I doubt it will hold up to the testing well.  Again, these pen probes just seem like a bad idea.

As usual, if there is any specific test you would like to see ran feel free to ask.

Offline SeanB

  • Super Contributor
  • ***
  • Posts: 16284
  • Country: za
Re: Handheld meter electrical robustness testing.
« Reply #1052 on: October 30, 2016, 06:38:37 pm »
How about doing the high voltage testing one time with the bare board in the case, with a mirror ( well insulated) so we can see both sides of the board at the same time to show where the arc over occurs.  Need to have the cover on for this, and latch it down well for your own safety, as it will make copper vapour as it blows.
 

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1053 on: October 30, 2016, 07:50:44 pm »
The mirror is an interesting idea.  Cheaper than a second camera!  :-DD

I don't talk about safety much when I run these tests.  The energy levels are normally so low, there is very little risk.  The generator has a fair amount of safety features designed in as well.  The half cycle generator adds a little more risk.  The cameras are about two and a half feet back to avoid any damage to them.  I am no where near the device I am testing when they go off.    I keep the output shorted up until I am ready to run the test.  If I test with fuses, light bulbs or other things with glass I will normally cover them up to at least contain the fragments.  Once the transient has fired, I shut the system down and short the outputs before doing anything else.  Running an open board would pose less of a risk than having something that can contain the pressure, like my fuse powered rocket.   

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1054 on: November 07, 2016, 02:48:24 am »
Some things just never get old.....    Video is finished with some bonus material.  About a half hour total. 

Offline Muttley Snickers

  • Supporter
  • ****
  • Posts: 2341
  • Country: au
  • Cursed: 679 times
Re: Handheld meter electrical robustness testing.
« Reply #1055 on: November 07, 2016, 02:54:40 pm »
Joe just recently posted another video on his Youtube channel so here it is linked below.   :-+

 

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1056 on: November 08, 2016, 01:26:36 am »
Sad to see the pen meter fail with the 220V line test but on the plus side at least it survived the grill starter unlike the vast majority of UNI-T meters I have looked at. 

We will see what Fluke has to say about the VHS tape.   

In the mean time, I have a little nicer analog meter I plan to look at next.  It's a Ketai Instruments KT-7050 sold under the Tekpower brand.  I may see if the schematics are available and just see if I can made it more robust rather than just damaging it. 

Offline joseph nicholas

  • Frequent Contributor
  • **
  • Posts: 408
  • Country: mx
Re: Handheld meter electrical robustness testing.
« Reply #1057 on: November 08, 2016, 01:45:01 am »
My cheap meter stopped working after only 9 months on all its voltage scales. I decided to buy a MF-47 analogue miltimeter kit.  At least if it gets damaged I know it would be easy to fix if needed from handy through hole parts.  It will probably be my go to meter of choice for trouble shooting.  I was inspired to do this after seeing the abismal quality of construct of the digital I bought.  The circuit board looked like it was stored in a wet basement for several years.  It was corroded and mouldy.  I bought it as new.  When I viewed the you tube video of a teardown of the same meter the internals were nothing like the video showed.  The exteriors were exactly the same.  When I got the meter the  a wire fell out of one of the probes for no reason.   
 

Offline 3db

  • Frequent Contributor
  • **
  • Posts: 331
  • Country: gb
Re: Handheld meter electrical robustness testing.
« Reply #1058 on: November 10, 2016, 11:51:40 am »
@Joeqsmith
Take a bite out of UNI-T  :-DD :-DD
 

Offline joseph nicholas

  • Frequent Contributor
  • **
  • Posts: 408
  • Country: mx
Re: Handheld meter electrical robustness testing.
« Reply #1059 on: November 10, 2016, 01:41:37 pm »
I found mine on ebay.  I haven't gotten the kit yet to assemble but I'll tell you how it goes.  This is perhaps the uglyest multi-meter ever made but it has a bale, and what could be some decent test leads for a change.  I paid something like 10.00usd for it but after I paid the site raised the price so go figure.  While I am waiting for the analog meter I tried to fix the cheapy.  It too has some schematics on line but unfortunatly no joy.  I doubt I cooked the chip because the thing still counts and works in the other modes.  Eventually I know I will fix it and it too should last. 

jnicholas
 

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1060 on: November 11, 2016, 04:24:40 am »
This is perhaps the uglyest multi-meter ever made but it has a bale, and what could be some decent test leads for a change.
jnicholas

 :-DD  It looks cleaner than that first analog meter I tore apart.  If that wasn't ugly enough, I really made a mess of it with all my mods.   :-DD
Looking forward to seeing the kit. 

Offline rsjsouza

  • Super Contributor
  • ***
  • Posts: 5986
  • Country: us
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: Handheld meter electrical robustness testing.
« Reply #1061 on: November 11, 2016, 04:33:05 pm »
Joe, that was a pretty cool mushroom in the high speed camera.
Despite it is stronger than your UNI-Ts, I can't help but imagine this is actually much more unsafe than a DMM on the table: the potential to burn the hand that holds the pen while this happens is pretty high.

Anyways, that was another excellent video! Thanks for doing this.
Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1062 on: November 12, 2016, 02:02:11 am »
Joe, that was a pretty cool mushroom in the high speed camera.
Despite it is stronger than your UNI-Ts, I can't help but imagine this is actually much more unsafe than a DMM on the table: the potential to burn the hand that holds the pen while this happens is pretty high.

Anyways, that was another excellent video! Thanks for doing this.

Thanks for the compliment.   I do like that camera.  It captures things that we just don't see with the slower one. 

I get what you are suggesting about how this meter could be less safe than the UNI-T's even though it held up a little better.  Keep in mind, while the grill starter did not end it's life, that little rectified AC waveform did.  IMO, its a pretty poor design.  We can look at the mushroom cloud and think it's worse but I really don't know.  The energy is just so low I am really not sure what it tells us.  If we ran fuse tests on every meter using an actual CAT III line, what would actually happen?  Meters we think look safer may indeed come apart with some a much more dramatic effect.   I just don't know.   We need to leave safety to the experts. 

When I made those couple of videos showing how stupid it is to jumper out an HRC fuse, at one point I stuck in some 28 AWG or something of that sort.  When I hit it with my generator, I did not even have enough energy to blow that wire.  It's pretty pathetic relative to photonicinduction's sort of destruction.   

Offline rsjsouza

  • Super Contributor
  • ***
  • Posts: 5986
  • Country: us
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: Handheld meter electrical robustness testing.
« Reply #1063 on: November 12, 2016, 03:05:48 am »
Well, my reason to assume the relative safety levels is based on the distance between the operator's body and the source of the explosion...

However, there's always a chance the dial of the meter on the table comes flying after the event and hits the operator in the head...  :-DD
Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1064 on: November 12, 2016, 04:41:21 am »
Well, my reason to assume the relative safety levels is based on the distance between the operator's body and the source of the explosion...

However, there's always a chance the dial of the meter on the table comes flying after the event and hits the operator in the head...  :-DD

I'm not sure I follow.  Are you suggesting because it is a pen meter that the operator would be holding it where a meter with leads, they would be holding the leads and not the meter and this means the pen is less safe?  There could be some merit to that. 

That knob flew about 8' across the room and hit the wall. 

Offline rsjsouza

  • Super Contributor
  • ***
  • Posts: 5986
  • Country: us
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: Handheld meter electrical robustness testing.
« Reply #1065 on: November 12, 2016, 12:58:45 pm »
Well, my reason to assume the relative safety levels is based on the distance between the operator's body and the source of the explosion...

However, there's always a chance the dial of the meter on the table comes flying after the event and hits the operator in the head...  :-DD

I'm not sure I follow.  Are you suggesting because it is a pen meter that the operator would be holding it where a meter with leads, they would be holding the leads and not the meter and this means the pen is less safe?  There could be some merit to that. 
Joe, yes I am suggesting that.

One ergonomic aspect that I noticed with the pen DMM I have is that it is impossible to do the "hand on the pocket" rule if measuring high energy circuits, unless an alligator clip on the negative side is used. With a traditional table DMM you can also hold both probes in one hand. Anyhow, I digress.

That knob flew about 8' across the room and hit the wall.
I have experienced situations where flying remnants of average energy short circuits (mains) were targeted at me. Fortunately I used glasses at the time, which saved my eyes from any damage.
Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1066 on: November 12, 2016, 05:12:24 pm »
That make sense.  There was a couple of clips included with the last meter that will fit over both ends.   In practice, I would never use a meter like this in the first place.  One wrong click on the dial and they are in current mode.   

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1067 on: November 19, 2016, 09:07:03 pm »
All tucked in and waiting for things to get stable....


Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1068 on: November 19, 2016, 09:43:38 pm »
Attached shows the previous data as well as the test I ran today.   Note that instead of running them from -10 to 40C, I ran them from -20 to 60C.   I also changed from a 1V reference to 1mV.  We can see running with 1mV we can now actually get some idea about their drift. 

While I have heard how poor the UT61E was, it is not as bad as the $300 Extech! Now again, I have been inside that Extech but I very much doubt anything I have done would have changed its TC.   The UT181A that I damaged, repaired, modified the PCB out performed my BM869s.   

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1069 on: November 20, 2016, 04:40:33 pm »
Comparing 5KY's UT61E with mine. Even thought the two meters are similar, I have a feeling UNI-T belongs to a resistor of the month club.   The JML datasheet shows +/-200ppm.  Not sure what the others are.  May just swap out the few passive parts and see what difference it makes.

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1070 on: November 21, 2016, 04:07:35 am »
Can you make a $50 UT61E more temperature stable than Dave's BM235 without buying an expensive reference?  What about making it even more stable than the BM869S and keep the parts cost under $1.00? 

Stay tuned for the next video where I will show how I pull it off.

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1071 on: November 21, 2016, 11:22:37 pm »
Here you go, for less than 35 cents in parts in single qtys, the UT61E takes on my two Brymen meters.   


 
The following users thanked this post: thisguy, 3db, MosherIV

Offline Vgkid

  • Super Contributor
  • ***
  • Posts: 2710
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1072 on: November 21, 2016, 11:44:27 pm »
Best Eleotron, couldn't even afford a "c".
If you own any North Hills Electronics gear, message me. L&N Fan
 

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11747
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #1073 on: December 21, 2016, 04:10:05 am »
2017 is shaping up... 

Offline Muttley Snickers

  • Supporter
  • ****
  • Posts: 2341
  • Country: au
  • Cursed: 679 times
Re: Handheld meter electrical robustness testing.
« Reply #1074 on: December 21, 2016, 08:34:56 am »
2017 is shaping up...

And shipping out by the look of it, what's in the box ?.   :-//

The suspense is killing me meters.   :-BROKE
 