Here's my last drift diagram of ADR#3. After adding a small transformer in my 3458A, it will go in there, making it a 9 digt instrument, at 6ppb StD.
You can see the EMP event by a nearby thunder strike at July 1st, and I have assembled it finally in a proper double shielded case, which caused the negative shift.
I already reported on its creeping characteristic, requiring 2 weeks to recover from an elongated time-out.
This would render the ADR1000 useless for any use in a DMM, or you would need to add a standby power supply.
Unfortunately, no other volt-nuts have repeated my tests, so I don't know yet, if this behavior is unique to my sample.
When I recently made all the changes during testing noise performance of my 3458A, and heating it 5°C higher, I recognized, that even after a 1 day long interruption the ADR shows a long termed creeping effect as well.
The +5°C change lets it creep even after 2 days.
I discussed with John R. Pickering about his tests on modified F7000 reference assemblies.
He also tried to anneal the die attach of the ADR by using his modified heating process in his mothers oven.
This did not work, as well the F7000 hysteresis removing method did not work, especially as the ADR does not show any hysteresis effect after a big temperature excursíon.
So I conclude, that creeping behavior is a common characteristics of all ADR1000.
The die attach of the LTZ1000 is an epoxy, and obviously has the characteristics of being solid, non elastic.
The baking process stresses and cracks weak inner bonds in the glue, so it does not change any more it's elastic properties.
The ADR1000s die attach can be described as being a viscious, elastic fluid, like tar.
Such a fluid does not show hysteresis, but it takes a lot of time to creep from one tension equilibrium to the other.
I did not yet do a reasearch, which die attach material shows exactly this behavior.
I will ask Analog and make the proposal of an ECR, to change to the LTZ1000s adhesive.