Need to repeat the INL measurements at ±1/3FS for 1 & 10NPLC when a lower noise/drift adjustable voltage source is available.
| Mods | ADC TC (CAL72 vs CAL175) in ppb/K | Comment |
| unmodified | -347 | |
| ADR1000 | -341 | |
| ADR1000, U160 ADA4522 | -406 | U151+U160+U165 socketed, U160: +12ref amp, TC change likely from LT1001 TC (maybe additionally from sockets) |
| ADR1000, U160+U165 ADA4522 | -336 | U151+U160+U165 socketed, U165: -12ref amp, back to original TC indicates LT1001 TCs were matched (or coincidence?) |
| ADR1000, U160+U165+U151 ADA4522, U170 OPA2206 | -335 | U151+U160+U165+U170 socketed, U151: +5ref amp, U170: DC_BUF & FET MOD, no significant TC change expected |
| ADR1000, U160+U165+U151 ADA4522, U170 OPA2206, 4x10µF | Unreliable: downtime/high drift (-163) | U151+U160+U165+U170 socketed, 4x10µF: MKS2 C151+C152+C160+C165, from my estimation there should be negligible TC effect (10µF measured <300pA leakage 10V/21°C) |
| ADR1000, U160+U165+U151 ADA4522, U170 OPA2206, U110 OPA205, 4x10µF | Temp range <1°C (-239) | U110+U151+U160+U165+U170 socketed, U110: integrator, no effect on TC expected as offsets are rejected by AZ |
| ADR1000, U160+U165+U151 ADA4522, U170 OPA2206, U110 OPA205, 4x10µF, RN | -73 | U110+U151+U160+U165+U170 socketed, RN: LT5400 10k for -12ref (replaces internal RN), most of TC improvement is expected from RN |
| ADR1000, U160+U165+U151 ADA4522, U170 OPA2206, U110 OPA205, 4x10µF, RN, LPF | -117 | U110+U151+U160+U165+U170 socketed, LPF A9 ADR100 output: 10Ω - 4700µF (low leakage) || 100nF, should not have an effect on TC -> implausible |
| ADR1000, U160+U165+U151 ADA4522, U170 OPA2206, U110 OPA205, 4x10µF, RN, LPF, 2x1nF | Temp range <1°C (-118) | U110+U151+U160+U165+U170 socketed, 2x1nF C0G: ZR_HI/ZR_LO & U110 +IN(ZR_LO)/AGND, no effect on TC expected |
| ADR1000, U160+U165+U151 ADA4522, U170 OPA2206, U110 OPA205, 4x10µF, RN, LPF, 2x1nF, ZGJC disabled | -11 | U110+U151+U160+U165+U170 socketed, ZGJC disabled: R184 (ZJUMP 619k: 0V or +400mV offset) & R185 (3Meg/-12ref: -200mV offset) disconnected from rusn1, TC effect plausible (see details below) |
| ADR1000, U160+U165+U151 ADA4522, U170 OPA2206, U110 OPA205, 4x10µF, RN, LPF, 2x1nF, ZGJC disabled (R185 connected) | -26 | U110+U151+U160+U165+U170 socketed, R185 (3Meg/-12ref: -200mV offset) connected back to rusn1, R184 still disconnected (ZJUMP 619k), TC effect plausible (see details below) |
| ADR1000, U160+U165+U151 ADA4522, U170 OPA2206, U110 OPA205, 4x10µF, RN, LPF, 2x1nF, ZGJC R185 ±12Vref | -24 | U110+U151+U160+U165+U170 socketed, ZGJC R185 ±12Vref: R185 (3Meg) switched between + and -12ref with ADG1219 steered by ZJUMP (±200mV offset), TC effect plausible (see details below) |
For the TC & drift fitting, penalized linear least squares was used (linear TC & cubic B-spline drift with discontinuous segment offsets after gaps and 3h warm-up exclusion).
Exchanging the RN for -12ref amp was primarily done to get rid of the excess noise (see previous posts), but for this specific U180 it occasionally was the major contributor to the ADC TC.
This is not applicable in general: it will only decrease TC for specific U180 and additionally LT5400 has some spread on tracking TC itself.
The effect on TC after adding the LPF at the output of the ADR1000 makes no sense, any ideas?
At first, disabling the ZGJC was only done to measure the INL glitch at 0V, but glad I decided to have a look at the effect on TC additionally.
Quite a surprise, first thought the large effect on the ADC TC made no sense at all.
After some deeper investigation, it became clear that the high TC of the rusn1 switch in conjunction with the resistors R184/R185 of ZGJC introduces TC around -100ppb/K.
This revealed another design flaw, which can simply be fixed by adding a SPDT switch and removing R184.
R185 (3Meg) switched between + and -12ref with switch (ADG1219) steered by ZJUMP (±200mV offset).
$$ TC_{\rm ZGJC} = -\frac{ R_{\rm sn} \cdot TCR_{\rm sn} \cdot \left(\frac{1}{R_{\rm jp}}+\frac{1}{R_{\rm jn}}\right) }{ \left(1+R_{\rm sn}\cdot\left(\frac{1}{R_{\rm in}}+\frac{1}{R_{\rm jp}}+\frac{1}{R_{\rm jn}}\right)\right) \cdot \left(1+\frac{R_{\rm sn}}{R_{\rm in}}\right) } \quad{\rm with}\quad R_{\rm sn}=14.4\,\Omega,\quad TCR_{\rm sn}=4000\,{\rm ppm/K},\quad R_{\rm in}=50k,\quad R_{\rm jp}=619k,\quad R_{\rm jn}=3Meg $$