Author Topic: AD587LW - 10V precision travel standard  (Read 98821 times)
Vgkid, p-wave, tszaboo and 52 Guests are viewing this topic.
Online AndreasTopic starter
Re: AD587LW - 10V precision travel standard
« Reply #125 on: October 08, 2021, 09:25:21 pm »
Ups I did it again:
(putting one of my references into the freezer for ~8 hours).
see also:
https://www.eevblog.com/forum/metrology/ad587lw-10v-precision-travel-standard/msg2449794/#msg2449794
AD587LW#1 had drifted back to the old value before the first freezing.
So I wanted to know wether I could trim it back near inital calibration by freezing again
or wether the reference is hardened by first freezing so no new change would happen.
Temperature logger shows that -20 deg C where reached.
And: we have again a step in output voltage: this time by -6.1 ppm. (previously -4.8 ppm)
with best regards
Andreas
Offline iMo
  • Country: li
« Reply #126 on: October 09, 2021, 10:43:37 am »
Why so steady drift upwards after so many years? It is a metal can, isn't it?
Wouldn't be better to get rid of all those settings resistors and thermistor around it and live with something like 10.034.256?? It is only a number..
Offline 3roomlab
« Reply #127 on: October 09, 2021, 12:37:06 pm »
so is it possible that there is a temperature where it does 0ppm drift?
maybe @ 5C ?  :-//
Offline iMo
  • Country: li
« Reply #128 on: October 09, 2021, 12:46:23 pm »
It is a travel standard so messing with ovens/coolers is not the right way to go, imho. I asked above as the drift could be caused by surrounding resistors and trimmers. I think a naked 587 (with the buffer) and a diode or thermistor for measuring the temperature close to the chip with a calibration graph glued on the top of the box would be less prone to such long term steady drifts..
« Last Edit: October 09, 2021, 12:59:15 pm by imo »
« Reply #129 on: October 09, 2021, 02:21:56 pm »
Hello,
it is a CERDIP8-package not a metal can.
Without those trimmers and termistors around we would have 1-2 ppm/K temperature drift.
So we would have annual variations of more than 20 ppm with my lab temperature range of about 18 - 33 deg C.
With the compensation I have only problems at very hot days (see those "dips" at above 30 deg C).
from my 4 samples 2 are drifting with about 2 ppm/year (which is not too much) the other 2 are much lower in drift compared to my LTZ-references.
with best regards
Andreas
The following users thanked this post: iMo
« Reply #130 on: October 09, 2021, 04:02:17 pm »
Hello,
I am dreaming of a small processor with 12 Bit A/D to measure temperature and outputting a correction voltage either per PWM or a 12 Bit DAC (LT1257). So with a 3rd order polynomal one could correct +/- 2ppm/K of selected AD587 over a 18-38 deg range with a resolution of  ~100 nV.
with best regards
Andreas
Offline iMo
  • Country: li
« Reply #131 on: October 09, 2021, 05:16:54 pm »
fyi - I've been using the vintage ADS1100 (6pin complete i2c 16bit adc) for converting temp from an LM35 sensor, compensating my 34401A :)
The following users thanked this post: Andreas
Offline DavidKo
  • Country: cz
« Reply #132 on: April 22, 2022, 10:27:50 am »
Hello,
with best regards
Andreas
Maybe you can use similar approach as in AN86, only you need to add the code for the temperature measurement and correction.
« Reply #133 on: April 22, 2022, 01:48:56 pm »
The more interesting question is here: how can I monitor that the poor processor is actually doing that what is intended.
And not hanging e.g. into a reset-loop or forgetting the calibration constants while outputting the minimum/maximum possible voltage?
with best regards
Andreas
Offline guenthert
« Reply #134 on: April 22, 2022, 03:41:11 pm »
with best regards
Andreas
     ???  The same as with any component.  If your VR is stable, you're going to assume all is peachy, if it jumps or drifts excessively you investigate.
     As for how to investigate software bugs, I don't think this application is much different from other MCU applications.   At first thought it seems even simple enough to be implemented as FSM, in which case the states could be enumerated and the index of current one output on otherwise unused pins (if that extra noise is deemed acceptable).
     And don't be afraid of reset-loops: the firmware of e.g. Keithley 181 always resets after each measurement cycle.
Offline miro123
« Reply #135 on: April 22, 2022, 05:18:22 pm »
with best regards
Andreas
i have few questions
1.Do you want only to monitor the temperature or do you want to do compensation? or just say it with other words does your application requires exact 10.00000V
2. How you gonna sense the temperature? I asking this question because i struggling to get it right. I've build TC oven I can easily achieve sub-milikelvin sensitivity 'mK' and reasonable long term stability. unfortunately the mesurement erors are several orders of magnitudes worse - e.g placing the probe few cm further away gives +-100..200mK, small smd component cannot be measured directly
I come to conclusion that sensing of temperature can only be done on silicon die - e.g. LM399, LTZ1000 or REF50xx
« Last Edit: April 22, 2022, 05:24:09 pm by miro123 »
« Reply #136 on: April 22, 2022, 06:32:03 pm »
Hello,
with the AD587LW project I am compensating T.C. according to the suggestion from Lars Walenius in the analog domain.
This works good for a small temperature range and you can do linear and partly 2nd Order temperature compensation.
Of course a oven will give a much better stability but I also want to have a 4-7 days continuous operation on battery.
Since the temperature range of analog compensation is too small for my "lab" temperatures it would be good to have a 3rd order temperature compensation.
But if you do this you also have to be shure that the compensation is "bullet proof".
So I actually want to monitor that the temperature compensation is working right.
(on a oven you also have a LED indicating that the setpoint temperature is reached):
For temperature sensing a good thermal isolation of the NTC/Voltage reference area against environment is essential.
Most of the temperature is distributed by the component leads. (so at best connect NTC ground to Ground pin of AD587)
And also selecting devices which have a low T.C. (< 2 ppm/K) in the temperature range of interest helps to maintain stability.
SMD components have usually lesser stability than DIP/CerDIP packages.
with best regards
Andreas
Offline magic
« Reply #137 on: April 24, 2022, 06:09:21 am »
You could try to measure chip temperature by monitoring its quiescent current, but I don't know how much TC it has and it could be too low to be practical.
Other than that there seems to be no hope for AD587. Somebody torn down a few units and even the "factory test pins" had no internal connection.
« Reply #138 on: April 27, 2022, 07:58:02 pm »
Hello,
somewhat off topic here but has to do with AD587:
Branadic asked me wether the zener reference voltage (visible on Pin 8 ) is typically "better" in T.C. than the output voltage.
I found this question very interesting. Especially because there is the Joe Geller Patend using a LM399 connected to Pin 8 to improve the T.C. of a AD587 to below 1 ppm/K.
https://patents.google.com/patent/US7382179B2/en
So I made a measurement.
My K2000 (temperature compensated) is connected to Pin 8 (which is filtered by a 1uF MKS02 capacitor).
My ADC27 is connected to the 10V output of the reference.
My 34401A is connected to both voltages measuring the ratio between VSense = 10V and Input = 7V
Here the results of 2 samples
AD587JQ#04 has a relative high T.C. of ~7ppm/K within 10-40 deg C
AD587JQ#08 has a relative low  T.C. of <2 ppm/K within 10-40 deg C
AD587JQ#4
Vout  ADC27  225.3 ppm / 32.4 deg C = 6.96 ppm/K  9998.078 mV @ 25 deg C
Vz    K2000  229.8 ppm / 32.4 deg C = 7.09 ppm/K  7132.876 mV @ 25 deg C
Ratio 34401A  4.44 ppm / 32.4 deg C = 0.14 ppm/K
calc  ratio   4.81 ppm / 32.4 deg C = 0.15 ppm/K
AD587JQ#8
Vout  ADC27  54.8 ppm / 32.9 deg C = 1.67 ppm/K 9999.375 mV @ 25 deg C
Vz    K2000  63.2 ppm / 32.9 deg C = 1.92 ppm/K 7136.059 mV @ 25 deg C
Ratio 34401A  8.6 ppm / 32.9 deg C = 0.26 ppm/K
calc  ratio   9.0 ppm / 32.9 deg C = 0.27 ppm/K
here the initial drift was rather high so I had to wait 2 temperature cycles until the values stabilized somewhat.
In both cases the output voltage divider has a relative low T.C. (well below 1 ppm/K).
With best regards
Andreas
The following users thanked this post: branadic, Mickle T., alm
Offline branadic
  • www.qoool-sensing.org
« Reply #139 on: April 27, 2022, 08:39:35 pm »
Thanks Andreas for performing this measurement and presenting the results. So that literally means, that the third order component in the t.c. is actually coming from the temperature compensation at the zener diode itself and is not added by the output stage, which makes it hard if not impossible to compensate for, sad news.
-branadic-
Measuring is like guessing, but more advanced.
« Reply #140 on: April 30, 2022, 03:31:13 pm »
Hello,
some update on ageing drift data of my 4 AD587JQ references now after ~ 4 years (>1400 days)
The drift data is relative to my LTZ references.
Most stable reference is LTZ#4 (green line in charts).
#01: this device I had 2 times over night in freezer (which gave ~-5ppm jumps) otherwise the drift would have been ~10 ppm in 4 years.
#02: ~4 ppm in 4 years against LTZ#4. so pretty stable.
#03: was 1 time in freezer (see jump) otherwise near zero ageing drift but some seasonal changes due to internal temperature changes.
#04: also nearly no ageing drift against LTZ#04 but rather sensitive to temperature when internal temperature rises above ~30 deg C
with best regards
Andreas
The following users thanked this post: alm
Online mawyatt
« Reply #141 on: April 30, 2022, 04:22:24 pm »
Hello,
Andreas
The Joe Geller idea is very clever, and your measurements suggest that the AD527 doesn't contribute much additional temp drift with the amp and gain resistors, as the dominate TC source seems to be the buried zener.
Wonder why National did not have a single chip that basically combined the LM399 with on-chip heater and the AD587 amp resistor network to get to 10 volt output? Or maybe they did and I just didn't know about it?
Anyway, thanks for the post and now you've created an intriguing question of whether this LM399 and AD587 combo might be an attractive alternative to a LM399 with discrete amp and resistors??
Best,
~Wyatt Labs by Mike~
« Reply #142 on: April 30, 2022, 05:41:51 pm »
Hello,
with the AD587 output ratio I do not have the flexibility that I want.
(converting from 6.6 to 7.2 V to exactly 10V).
Even if the ratio stability is rather good.
So I´d rather use a high resolution DAC with specified ratio drift for that task.
(So similar to Ian Johnstons approach).
with best regards
Andreas
Online Kleinstein
« Reply #143 on: April 30, 2022, 06:11:19 pm »
The LM399 is running quite hot and this can effect the stability of resistors.
A stable 10 V reference is not such a common application that it needs a special chip for this. There is still the option to have a discrete zener, resistors and temperature regulation that includes the resistors.
NS had a LM369 zener reference with 10 V scaling, but without a heater. So somewhat similar to the AD587.
Online mawyatt
« Reply #144 on: April 30, 2022, 06:19:10 pm »
50 years ago or so National had a Hybrid 10V reference but don't think it had a heater either. We used these a lot, but recall they were somewhat expensive.
I'm sure there was a very good reason for not including the heater, otherwise they would have done it!!
Best,
Offline kleiner Rainer
  • Rainer DG1SMD JN48ts
« Reply #145 on: May 01, 2022, 06:50:46 am »
Best,
National Semiconductor LH0070 - there are two versions of it: a hybrid module (there comes the "LH" for hybrid from) built around a LM329 reference, some laser trimmed resistors and a LM741 and a newer version built around a LM169, but still on a hybrid.
I got a sample of the older version during my internship, forty years ago - it is still working and of course well aged.
Link:
https://www.richis-lab.de/REF09.htm
I had a link with pictures of the first version, but it seems to have disappeared.
Greetings,
Rainer
« Last Edit: May 01, 2022, 06:52:33 am by kleiner Rainer »
The following users thanked this post: chickenHeadKnob, Andreas, mawyatt
« Reply #146 on: July 23, 2023, 06:17:39 pm »
Hello,
obviously I missed to repeat 1/f noise measurement after population of the transient zeners in ~April 2019
So after doubt of IMO I repeated the noise measurement to see if there is any increased noise by the zeners.
I could not see any significant change in 0.1 - 10 Hz noise.
The values in the table are averaged values out of 15-19 repeated measurements with 100 sec measurement time each.
with best regards
Andreas
The following users thanked this post: alm, iMo
« Reply #147 on: October 20, 2024, 07:07:14 pm »
Hello,
I wanted to know how the trimmers which are used for adjusting the output voltage and the T.C. of the cirquit affect stability of the output voltage of the AD587LW cirquit.
https://www.eevblog.com/forum/metrology/ad587lw-10v-precision-travel-standard/?action=dlattach;attach=402882
I am using 2 different trimmer types. One is a Bourns 3296W 50K trimmer specced with 100 ppm/K. The other a cheaper RKt 50K trimmer with 200 ppm/K.
As sanity check I also tested a divider out of two 27K metal film resistors with 50 ppm/K.
I did a T.C. measurement over ~10 .. 45 deg C with 5V supply measuring at the wiper at 25%/50%/75% hoping that the T.C. over the whole trimmer resistance would compensate somehow leading to better overall stability.
The two 27K resistors showed ~72 uV change over the temperature range so below 1 ppm/deg C @2.5V giving a good compensation.
To my surprise the Bourns sample was rather "jumpy". Whereas the RKt sample showed more T.C.
All in all the deviations at the wiper output were around 1-3 mV.
So even when only regarding a relative small temperature range you cannot expect to have better stability as ~5mV@5V or 0.1%.
(Datasheet spec of the Bourns is ~1% over the whole temperature range. The RKt has no stability spec except the 200 ppm/K)
Calculating with 5mV for the voltage trimming (with a 10 Meg series resistor to the trim pin) gives ~2.6 uV change of the 10V output so 0.26 ppm.
The T.C. trimmer is more complicated to calculate. There are two 680K NTCs in series and a nominal 1.5Meg series resistor which has to be adjusted to the 3rd order correction of the T.C. of the AD587. But generally the trimmer influence is larger on this trimmer.
For AD587LW#3 and AD587LW#4 I calculated up to 24 uV or 2.4 ppm for a 5 mV trimmer voltage change.
So perhaps this also is a explanation for the 5 ppm shift after putting the AD587LW devices into the freezer.
So all in all I think that the idea of using DACs/PWM dividers instead of trimmers would be more suitable for precision devices.
with best regards
Andreas
« Last Edit: October 20, 2024, 07:08:52 pm by Andreas »
Offline iMo
  • Country: li
« Reply #148 on: October 20, 2024, 07:52:52 pm »
I saw the jumpiness with 10t cermet trimmers too. What I found helps a little bit is to set the wiper in a position where it is in the middle of the "dead zone" (there is quite large dead zone where the teeth of the inner plastic wheel carrying the wiper do not touch the screw gear).
Offline bastl_r
« Reply #149 on: October 21, 2024, 02:01:44 am »
My experience with the trimmers on my Fluke 343 is that if you don't know in which direction they were last adjusted, it makes sense to first adjust the trimmers significantly and then slowly feel your way to the desired setting from there.
regards