Hi Frank!
Thanks for the hints on the 3458A measurements, unfortunately the 3458A is in a permanent action during working hours this summer, thus I've made measurements with the Fluke 8508A only.. 3458A in October/November this year most probably. There is still LAB1 with the Fluke 8588A and with a planned measurement in August this year.
Anyhow, I incorporated your hints into below measurements, at least I hope so
BTW - I've got an info from ADI the ADR1001 will most probably be released in FEB 2024..
####
A new LAB measurement today - this time in LAB2 and with Fluke 8508A.
The previous measurement with the same 8508A in the LAB2 (April 18th):
Vref = 9.9998697
Vref = -9.9998745 at T_amb=21.1C, T_board=35.49C, AC powered.
This time aprox 5.5uV higher at T_amb=23.0C, T_board=35.52C, battery powered.
Below the results - the measurements were performed in a sequence from "A" to "O".
The green 10xFast measurements - the math has been performed inside the Fluke.
The shot as usual for demo purposes, during the measurements the myADR box was placed about 50cm off the meter.