In the meantime i have done some multiplexer measurements on three more Nomca 1603 8x 5K arrays. They got mounted on SMD adapters with gold plated pins and cleaned in ultrasonic bath. Then i setup 4 wire R measurements using a HP3456A at 0,1 mA measurement current with averaging to get below ppm resolution. Temperature span was 18 °C. Error on a single resistor TC measurement is about 0,03 ppm/°C.
Results on TCs for each part (average( 8 ), stdev( 8 ), span( 8 )):
D 4,34 +/- 0,62 ppm/°C, max-min = 1,85 ppm/°C
E 11,23 +/- 0,63 ppm/°C, max-min = 1,74 ppm/°C
F 16,20 +/- 1,55 ppm/°C, max-min = 4,51 ppm/°C
According to my measurements those are pretty serious parts, well suited to produce temperature compensated voltage dividers. And it makes complete sense to measure TCs of individual resistors in order to determine a good combination - better than statistics! Combining resistors properly each of my three parts should give me a 13:1 divider with less than 0,03 ppm/°C on the division ratio. At least that's the result of a simulation using the measured curves of the individual resistors.
Regards, Dieter