Author Topic: Statistical arrays  (Read 9748 times)

0 Members and 1 Guest are viewing this topic.

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Statistical arrays
« on: December 09, 2018, 08:49:16 pm »
Dear colleagues,

First of all we heared about interesting properties of TaN resistors statistical arrays.
There are some CERN- related papers describing developments made by great Mr John Pickering.
He seems to like using this technique as it does the job at reasonable price and in relatively simple, elegant way.
Inspired by him I made simple setup to see what can be achieved using off the shelf component.
I used 8x10k NOMCA in SOIC 16 package and wired the circuit as shown on the drawing below.
It is thermally balanced divider by 10.

The setup takes advantage of HP 34401A in ratio mode, Data Precision 8200 as (4-wire) voltage source, thermocouple thermometer.


P.S.  please post examples of statistical arrays applications (schematics) that You know about

« Last Edit: December 30, 2018, 12:30:39 pm by doktor pyta »
 
The following users thanked this post: lukier

Online Echo88

  • Frequent Contributor
  • **
  • Posts: 571
  • Country: de
Re: Statistical arrays
« Reply #1 on: December 09, 2018, 08:57:21 pm »
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #2 on: December 09, 2018, 09:27:18 pm »
I was always concerned about the tracking t.c. of 5ppm/K for Vishay TDP16031002 or NOMCA16031002. I wonder if TaN statistical arrays are any better. But even if they were, they seem to be unobtanium. I once searched for more information, but wasn't able to find the real hints. So any link to related papers are welcome.

https://www.ebay.de/itm/1pcs-TDP16031002BUF-Vishay-Resistor-Networks-Arrays-10Kohms-16/263962507025
https://www.ebay.de/itm/1pcs-NOMCA16031002ATS-Vishay-Resistor-Networks-Arrays-16-pin-10/263957703582

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #3 on: December 09, 2018, 10:25:08 pm »
Now watch this:

1. division ratio error: +12ppm (no adjustment!)

2. T amb=25'C, T heat=50'C
average T.C. of division ratio: 0.32ppm/'C

3. Thermal shock 25'C; -20'C; 25'C
division ratio change: 0ppm

4. Thermal shock 25'C; 100'C; 25'C
division ratio change: 0ppm

Keep in mind 1ppm is a resolution of my simple but very stable setup (further measurements need to be done).


To be precise I used P/N NOMCA16031002ATS bought from Farnell.
« Last Edit: December 11, 2018, 09:04:16 am by doktor pyta »
 
The following users thanked this post: Andreas, EmmanuelFaure

Offline d-smes

  • Regular Contributor
  • *
  • Posts: 88
  • Country: us
Re: Statistical arrays
« Reply #4 on: December 10, 2018, 11:12:09 am »
When you did thermal shock, did you go from 25C to -20C, back to 25C, and then make the measurement?  Or did you go from 25C to -20C and then make the measurement?  I just wondering why the TC didn't show up in the ratio measurement...
 

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #5 on: December 10, 2018, 11:56:52 am »
2. and 3.
3. and 4. describe situation after thermal shock. So measurements were performed @25'C.
« Last Edit: December 10, 2018, 07:53:15 pm by doktor pyta »
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #6 on: December 10, 2018, 04:55:41 pm »
So this means it's not a temperature coefficient [ppm/K] but hysteresis value [ppm] you measured for any case @ 25°C, doesn't it?

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #7 on: December 10, 2018, 06:31:22 pm »
Using more resistors is a known way to improve TC matching and similar in a kind of statistical way. Knowing how the resistors are located on the substrate could go beyond just the statistical improvement. The more normal case is to have a gradient in TC over the chip.
 

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #8 on: December 10, 2018, 07:54:15 pm »
So this means it's not a temperature coefficient [ppm/K] but hysteresis value [ppm] you measured for any case @ 25°C, doesn't it?

-branadic-
3. and 4. is hysteresis in ppm

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #9 on: December 10, 2018, 08:04:27 pm »
Using more resistors is a known way to improve TC matching and similar in a kind of statistical way. Knowing how the resistors are located on the substrate could go beyond just the statistical improvement. The more normal case is to have a gradient in TC over the chip.

True, but here in metrology section only few (2...3?) of us ever tried using this technique.
Besides I haven't found their results.
I needed numbers to have reference point for my further designs.

Kleinstein, could You please point a literature, papers which explains theory behind statistical arrays ?

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 811
  • Country: de
Re: Statistical arrays
« Reply #10 on: December 10, 2018, 11:18:27 pm »
Just for my naive understanding:

I wonder how much improvement you're getting from that "mixed" wiring and if that very good behaviour here is only luck?

Such arrays are normally specified with a tracking and an absolute TC. The absolute is worse than the tracking, because the tracking is more or less only the difference in TCs of the resistors. I think with such a layout used here one can improve the thermal homogeneity but I wouldn't expect an order of magnitude improvement only by that.

Which point I'm missing here?


A second point is the increase of the amount of resistors. Averaging out the TC over many resitsors only works if the mean of many resistors is really zero (or nearby). I have no experience with such tan arrays, but I measured a lot of thick film HV resistors (Caddock USF) and resistors with negative TC are very rare.
« Last Edit: December 10, 2018, 11:24:17 pm by e61_phil »
 

Online pigrew

  • Frequent Contributor
  • **
  • Posts: 515
  • Country: us
Re: Statistical arrays
« Reply #11 on: December 11, 2018, 12:09:18 am »
A second point is the increase of the amount of resistors. Averaging out the TC over many resistors only works if the mean of many resistors is really zero (or nearby). I have no experience with such tan arrays, but I measured a lot of thick film HV resistors (Caddock USF) and resistors with negative TC are very rare.

My naïve understanding:

These are used as voltage dividers. It doesn't matter if the mean TC is zero (or not). It matters that the effective TC of each group is matched. If one group has a +14.5ppm/C and the other is +14.6ppm/C, then the ratio will be very close to constant with temperature, even though the value of the resistors changes significantly with temperature

These resistors are specified as having a tracking coefficient of 5 ppm/C. This means that the ratio of any two resistors will change no more than 5 ppm/C.

With enough resistors, the sample mean TC approaches the population mean TC (assuming random sampling), so the TC of the two samples of resistors should approach equality. In this case, we think that neighboring resistors will have better matched TC and more identical temperatures, so we try to distribute likely-matched resistors into separate groups, so the groups will match each other with fewer resistors required than if we assumed a random sampling of resistor TCs.

 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 811
  • Country: de
Re: Statistical arrays
« Reply #12 on: December 11, 2018, 12:12:08 am »
Ahh  :palm:

Yes, of course you only need to average out the tracking, because your whole divider is in that chip (not like mine, where only the HV part consists of the Caddock resistors) :)

Thanks!
 

Offline Magnificent Bastard

  • Regular Contributor
  • *
  • Posts: 123
  • Country: aq
Re: Statistical arrays
« Reply #13 on: December 11, 2018, 12:36:33 am »
Since (on the NOMCA), the resistors are on a ceramic substrate, and the package material is epoxy, one would have to assume that the two materials will behave differently with changes in temperature and humidity (with humidity being the more evil of the two effects).  This would most likely result in a curvature lengthwise-- suggesting that the resistors more near the center of the package will track better than the resistors near the ends of the package.  For better statistical tracking, we can also combine multiple units.  for example, we start with some NOMCA 10K 16-pin (8-resistor) networks.  Now place them side by side on a PCB.  We wire R1 in series on all three packages (resulting in a 30K resistor R1')-- and then do the same with the other 7 resistors in each package.  Now we have 8 resistors (R1' through R8') that have a value of 30K-- with "statistically better" absolute value and TCR tracking.  Let's now make the same 10:1 divider.  We need (3) 30K resistors in series, and (3) 30K resistors in parallel, making 90K and 10K.  It is assumed that resistors next to each other in the network will TCR/humidity track each other better than resistors farther apart from each other-- so, the best arrangement is to "interdigitate" the resistor sets (the 90K and the 10K).  With this in mind, we ignore R1' and R8'.  R2', R4', and R6' are wired in series making 90K, while R3', R5', and R7' are wired in parallel, making the 10K.  Now, wiring the 90K in series with the 10K, we end up with a 10:1 ratio that will track very well-- (at least theoretically, anyway).  My "gut feel" for this arrangement is that it will track within +/-0.1ppm over temperature and humidity.  Not bad actually-- and you can do this over production without selecting parts!
 
The following users thanked this post: TiN

Offline JimmyJo

  • Contributor
  • Posts: 9
  • Country: hk
Re: Statistical arrays
« Reply #14 on: December 11, 2018, 03:27:51 pm »
Dear Doktor Pyta,
I see that you have interleaved the divider resistors of different power dissipation.  Would a current step be useful to determine the power coefficient of ratio?

p.s. Has anyone actually seen the internal construction of of these resistor networks? are they really just 8 resistors side by side on the substrate? 

 

Offline Magnificent Bastard

  • Regular Contributor
  • *
  • Posts: 123
  • Country: aq
Re: Statistical arrays
« Reply #15 on: December 11, 2018, 04:43:12 pm »
Dear Doktor Pyta,
I see that you have interleaved the divider resistors of different power dissipation.  Would a current step be useful to determine the power coefficient of ratio?

p.s. Has anyone actually seen the internal construction of of these resistor networks? are they really just 8 resistors side by side on the substrate?

Actually, in his original post, they were NOT described as being "interleaved" (or "interdigitated"), which they should have been.
Edit:  I was looking at the schematic and not the layout.  The layout is done correctly-- (using the inner-most resistors only, and interdigitating the series and parallel connected resistors-- this provides the best chance of tracking).

Yes, the resistors are made on a ceramic substrate one after another.  There is literature on the Vishay website that describes how they are made.  There are also thin-film networks that use a silicon substrate, and these are less desirable for this application (because they flex more).
« Last Edit: December 11, 2018, 05:07:02 pm by Magnificent Bastard »
 

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #16 on: December 11, 2018, 07:29:51 pm »
Would a current step be useful to determine the power coefficient of ratio?

Other test would be more suitable.
I've just applied 7.07V instead of 10.00V (half of power is dissipated at 7.07V)
34401A in ratio mode changed reading from:
100.0012m @ 10V to
jumping between 100.0011m and 100.0012m @7.07V

so the change is below 1ppm and it is not necessary 100% related  to DUT.
« Last Edit: December 11, 2018, 07:46:54 pm by doktor pyta »
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #17 on: December 11, 2018, 08:06:18 pm »
I have no reference for a chance to get more than just the statistical advantage. The point is more from simple reasoning:
The TC of the individual resistors will likely not be purely random but likely will also have some position dependent part, like changing in one direction from one side to the other. In this case the use of interleaved resistors is more efficient in compensating these differences than just averaging over random values.

Also for self heating the interleaving helps to get faster and better temperature matching between the resistors, even with self heating.

I really like the idea of using the array.
 

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #18 on: December 11, 2018, 10:47:08 pm »
Finaly I found a good paper related to our subject.
'20A trapezoidal reference current pulse generator for the evaluation of current transducers'  DOI: 10.1109/I2MTC.2013.6555434


My next step will be testing an array to replace temperature setting divider in LTZ1000 circuit.
From my measurements I expect that one 8-resistor array may do the job.
I have an idea shown on the schematic below.
Divider by 13 could replace 1k/12k divider which might be used only with LTZ1000CH.
Divider by 16 could replace 1k/15k divider which might be used with LTZ1000ACH, unfortunately it will be hot inside. In this case thermal layout seems to be slightly worse.
Off course more possibilities can be obtained using two arrays, but I didn't came up with any universal solution like thermally balanced divider by 14.

If someone has better ideas please share.
« Last Edit: December 11, 2018, 11:47:29 pm by doktor pyta »
 
The following users thanked this post: Magnificent Bastard

Offline Magnificent Bastard

  • Regular Contributor
  • *
  • Posts: 123
  • Country: aq
Re: Statistical arrays
« Reply #19 on: December 12, 2018, 01:48:52 am »
Finaly I found a good paper related to our subject.
'20A trapezoidal reference current pulse generator for the evaluation of current transducers'  DOI: 10.1109/I2MTC.2013.6555434


My next step will be testing an array to replace temperature setting divider in LTZ1000 circuit.
From my measurements I expect that one 8-resistor array may do the job.
I have an idea shown on the schematic below.
Divider by 13 could replace 1k/12k divider which might be used only with LTZ1000CH.
Divider by 16 could replace 1k/15k divider which might be used with LTZ1000ACH, unfortunately it will be hot inside. In this case thermal layout seems to be slightly worse.
Off course more possibilities can be obtained using two arrays, but I didn't came up with any universal solution like thermally balanced divider by 14.

If someone has better ideas please share.

This looks good-- no need for the 8th resistor-- you could use a 14-pin device.  3K would be ideal, but not available (ever!) as Vishay NOMCA-- so you can use 2K or 5K.  If you want to special order, you can buy some series SOIC-C at 3K from IRC/TT-Electronics (through Mouser)-- you will have to ask Mouser for a quote, and there will be a minimum buy quantity-- (~100?)  A 14-pin 2K or 5K NOMCA will work just fine, and you might find some in stock (somewhere).

Edit:
Another possiblity is to use 14-pin 1K NOMCA networks (3 of them).  As before, wire all R1's in series to make R1' (a 3K resistor).  Now do the same for R2-R7 making R2' through R7'.  Now wire these 3K resistors (R1' though R7'), as before:  R1', R3', R5', and R7' in series to make 12K; and R2', R4', and R6' in parallel to make 1K.  You now have a 12K:1K divider that will TCR and humidity track within 0.1ppm/K-- and this will be true over production without special parts selection.  Pretty cool!


Aren't Hamon dividers fun and useful?
« Last Edit: December 12, 2018, 02:04:43 am by Magnificent Bastard »
 
The following users thanked this post: doktor pyta

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #20 on: December 12, 2018, 12:05:39 pm »
Has anyone actually seen the internal construction of of these resistor networks? are they really just 8 resistors side by side on the substrate?

zlymex once posted a picture of TDP series array.
https://www.eevblog.com/forum/metrology/any-resistor-series-that-are-stable-wrt-timetemp-but-not-precision-values/msg1056895/#msg1056895

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #21 on: December 12, 2018, 03:09:02 pm »
I wonder how they perform in a 4k : 10k or at least 10k : 25k configuration for the 7V --> 10V boost circuit. Any real world numbers on that?

EDIT: The only used experience seems to be with LM399

-branadic-
« Last Edit: December 12, 2018, 04:04:26 pm by branadic »
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #22 on: December 12, 2018, 03:24:51 pm »
This will be next step. Will be tested for sure.

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #23 on: December 12, 2018, 07:11:03 pm »
Have ordered a few NOMCA16035001 for experiements as NOMCA14035001 are equal to NOMCA14032001 and NOMCA16032001 simply unobtanium. So the 8th resistor can be used in parallel to some low ohmic resistor for fine adjusting the 10V output voltage.

-branadic-
« Last Edit: December 12, 2018, 07:14:52 pm by branadic »
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline razvan784

  • Contributor
  • Posts: 20
  • Country: ro
Re: Statistical arrays
« Reply #24 on: December 12, 2018, 08:57:44 pm »
Played with this (in LTSpice) some time ago as a replacement for 13k-1k.
Tried a layout that is as symmetrical as possible.
Didn't manage to built it yet.
 
The following users thanked this post: doktor pyta, Magnificent Bastard

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #25 on: December 12, 2018, 10:04:08 pm »
The 13:1 divider is a really nice solution. However it would provide only a limited amount of averaging. So I am not so sure it would be really stable and low TC.

For the 15:1 ratio, there is an easier solution: 5 in series and 3 in parallel.
 

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #26 on: December 12, 2018, 10:44:54 pm »
razvan784, thanks for the inspiration!

Below is my version of divider by 14.
1.6k output resistance should not be a problem.
The resistors are grouped in pairs with same power dissipation.

What do You think guys ?


Edit: tempco measured for comparision.
Let's call it layout v.1.

T1=25.2'C
T2=60.4'C
mean T.C. of division ratio -3.9ppm/'C

and

division coefficient measured: 14.00088
« Last Edit: December 31, 2018, 07:11:35 pm by doktor pyta »
 

Offline razvan784

  • Contributor
  • Posts: 20
  • Country: ro
Re: Statistical arrays
« Reply #27 on: December 13, 2018, 03:38:44 pm »
Thank you doktor pyta. Yes, your design is better, more symmetric.
Kleinstein, I agree that averaging is limited; these designs would work well however if the individual TCs were not just randomly distributed but more like a gradient, as you noted earlier. On the other hand I have done no such measurements; I finally abandoned this design in favor of just using two selected foil resistors.
 

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #28 on: December 13, 2018, 09:34:03 pm »
Update: divider /14 with slightly corrected layout.


Edit: tempco finally measured.

T1=25.2'C
T2=60.0'C
mean T.C. of division ratio 0.41ppm/'C

and

division coefficient measured: 14.00055

« Last Edit: December 30, 2018, 12:39:38 pm by doktor pyta »
 
The following users thanked this post: EmmanuelFaure

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #29 on: December 14, 2018, 09:04:42 am »
For the layout, selecting the resistor positions, the power dissipation should not be that important, if used at a reference, as the power dissipation would be constant. The power dissipation could be important if the divider is used at variable voltage, e.g. to set an amplifier gain.

So It is not sure the 2 nd version is actually better than the 1 st. It is likely more about a gradient in the resistor properties, so to have the resistors well mixed and evenly distributed. Here the 1st version is still not that good as it has the resistors towards ground are in the center - so a difference from center to edges would show up. The 2nd version is more like a little biased left to right.
 
The following users thanked this post: doktor pyta

Offline Magnificent Bastard

  • Regular Contributor
  • *
  • Posts: 123
  • Country: aq
Re: Statistical arrays
« Reply #30 on: December 14, 2018, 05:51:09 pm »
The idea is to keep the temperature as uniform as possible across the ceramic substrate.  This will allow for better tracking over time and temperature-- (the Ta2N resistors "age" by heat and oxygen availability-- keeping them close to each other in temperature will aid in uniform aging characteristics).  Also, the farther away from the center of the package, the less the resistors will track; and resistors next to each other will track better than resistors far apart; and this is due to the difference in TCE between the ceramic substrate and the epoxy package; and also-- humidity will have a similar (more pronounced) effect.  The layout should be as symmetrical as you can get it while observing the above constraints.  As oxygen attacks the tantalum nitride compound-- tantalum pentoxide is formed, which is not conductive (and is unaffected by water molecules-- even under power).  Since these are thin-film resistors, that leaves only one drift mechanism-- and they always drift "up" in value.  When a thick enough layer of tantalum pentoxide forms-- the resistors will slow their drift rate to almost zero.  This is what the 90-day burn-in at 125oC is all about-- to artificially age these resistors so that the subsequent drift over time is almost zero-- doing this in an ozone-rich environment will speed up this process.
 
The following users thanked this post: e61_phil, doktor pyta

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #31 on: December 14, 2018, 06:31:28 pm »
I'm pretty sure the resistors are processed in a panel for economic reasons and cut into pieces after PVD process. So one can expect the resistors to be very uniform within one ceramic die. So it's all about heat distribution across the ceramic die.

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #32 on: December 14, 2018, 06:36:38 pm »
I wonder if the temperature distribution inside statistical array could be evaluated objectively before practical measurements.
Some simple freware thermal simulation software would be useful but without knowing dimensions, thermal conductivities specifc to the array it may not be good enough.
Any ideas?
« Last Edit: December 14, 2018, 07:15:11 pm by doktor pyta »
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #33 on: December 14, 2018, 07:20:28 pm »
Well, it's not that you now nothing about the array... you can estimate the size of the ceramic, you now it's alumina (Al2O3) substrate and know about thermal conductivity and that there are 7 or 8 resistors on it. But I'm sure there is no need to simulate the heat distribution, but to think about how heat will spread.
On the other hand you can use any thermal simulation program available to get an idea of what is thermally going on. If you want to do a real world simulation you need to sacrifice one unit, crack it open and measure the dimensions including the thickness of the TaN layer.

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline Magnificent Bastard

  • Regular Contributor
  • *
  • Posts: 123
  • Country: aq
Re: Statistical arrays
« Reply #34 on: December 14, 2018, 08:08:57 pm »
OR, you can just measure it!  Measure R2-R8 while applying power to R1 (vs. no power).  Now do the same by powering R2, and measuring the other resistors.  Same with R3 (and so on) until you have a complete set of measurements.  With some mathematical magic, you should be able to understand how each resistor's dissipation affects the other resistors.  Intuitively, it "feels" like having the higher-dissipation resistors on the ends of the package will help keep the substrate temperature gradients lower (as opposed to having the higher dissipation resistors in the center of the package).  This can be verified with the above experiment.  An additional interesting question is: "Do the temperature gradients of the substrate change when mounted to a PCB vs. a package in "free air"?

Fun stuff!
 

Offline JimmyJo

  • Contributor
  • Posts: 9
  • Country: hk
Re: Statistical arrays
« Reply #35 on: December 15, 2018, 06:53:54 am »
Inspired by Zlymex, I present to you this 1K version of the NOMCA1603.  Upper right corner is Pin1 in both pictures.
 
The following users thanked this post: TiN, Echo88, Andreas, e61_phil, EmmanuelFaure, doktor pyta, razvan784, Magnificent Bastard, syau

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #36 on: December 15, 2018, 10:25:08 am »
Nice job JimmyJo!

Below distances measured from the image.

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #37 on: December 15, 2018, 01:09:15 pm »
Since Vishay is not a european company the distance between the resistors is most likely 0,3175mm or 0,0125". There seems to be a small parallax error from the microscope.

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #38 on: December 15, 2018, 07:50:01 pm »
Just curious if NOMCA16035001 could perform as temperature stabilized 10k resistance reference just as good as Fluke SL935 from TiN? There are enough 5k resistors inside one package to realize that.

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #39 on: December 15, 2018, 08:12:19 pm »
One could use 1 or 2 of the resistors for heating, but even than the resistor is encapsulated in epoxy. So there can be humidity effects.

Extra heating might be useful to compensate for self heating, if the resistors are used for something like a divider with a variable voltage.
 

Offline splin

  • Frequent Contributor
  • **
  • Posts: 806
  • Country: gb
Re: Statistical arrays
« Reply #40 on: December 16, 2018, 03:42:48 am »
One issue with statistical arrays is that they multiply the number of leakage paths and also makes guarding more of a problem. If, for example, you have 10 resistors in series forming the bottom leg of a divider, then to fully guard it you would need 10 guard voltages.

In practice, given the lower voltage drop across each resistor, you would probably compromise by guarding fewer nodes at the expense of potentialy a bit more leakage. The calculations might be interesting if the leakage resistance of the PCB is significantly non-linear with voltage - perhaps due to humidity and/or surface contamination (I've no idea if it is).
 
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #41 on: December 16, 2018, 11:42:54 am »
I was thinking about a pcb with the 5k resistor array connected as a 10k reference on the component side of the board and an oven control circuit on the back or the pcb with a big copper area on the back mounted to some BPR10 resistor forming an oven that is running at 35°C ... 40°C.
The thermistor could be some SMD longterm stability type mounted on the component side close to the resistor array. If permanently powered there should be no problem with humidity after a small run in phase. It's then a question of thermal insulation and oven temperature stability (0.1°C, 0.01°C or 0,001°C). But the oven could be of much smaller size compared to SL935.

-branadic-
« Last Edit: December 16, 2018, 12:55:18 pm by branadic »
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #42 on: December 30, 2018, 11:09:47 pm »
Little update in first post and Reply #28 and Reply #26.

In short, T.C. :

-from Reply #26: -3.9ppm/'C
-from Reply #28: +0.41ppm/'C

I couldn't believe my eves so I double checked everything and I made multiple measurements.
Both layouts were using the same specimen of NOMCA array.
« Last Edit: December 31, 2018, 08:01:38 am by doktor pyta »
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #43 on: December 31, 2018, 12:02:20 am »
Can you rearrange the resistors and repeat the measurement? R7 + R8 as the center resistors, R5 + R6 symmetrical an both sides, followed by R3 + R4 and finally R1 + R2 (see attachement).

-branadic-
« Last Edit: December 31, 2018, 08:49:23 am by branadic »
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #44 on: December 31, 2018, 12:27:53 am »
@branadic

proposed layout is definitely worth testing.
I will do it within one or two days.

P.S. please remove connections on the layout part of the attached drawing so the readers won't be confused.
I will post new layout drawing soon.

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #45 on: December 31, 2018, 12:23:11 pm »
Measurements of the layout suggested by branadic.
Proper drawings below.


T1=25.3'C
T2=63.6'C
mean T.C. of division ratio -5.8ppm/'C

and

division coefficient measured: 14.00017
 
The following users thanked this post: branadic, Andreas

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #46 on: December 31, 2018, 04:34:47 pm »
There are a few possibilities left, like having R7 and R8 on the outer ends and R5 and R6 in the center.

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Online Andreas

  • Super Contributor
  • ***
  • Posts: 2407
  • Country: de
Re: Statistical arrays
« Reply #47 on: December 31, 2018, 11:31:55 pm »
Hmm,

did you use one and the same sample for all these 3 wirings
or are that different samples for each wiring?

with best regards

Andreas
 

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #48 on: January 01, 2019, 12:02:12 am »
I used one sample in these 3 layouts. Otherwise it would not make big sense.
Cheers! Happy New Year!

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 811
  • Country: de
Re: Statistical arrays
« Reply #49 on: January 01, 2019, 08:47:31 pm »
Did you measure the individual TCs?

I wonder if these measurements are due to the position or due to lucky matching of the resistors for the different sides. Or is that already clear and I overlooked something?
 

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #50 on: January 01, 2019, 09:43:59 pm »
It would be good idea to have individual resistors; T.C. measured.
It may contribute our understanding of the problem.
I have to think how to use my gear to make 8x 4wire measurements at two temperatures easiest way.

At this point my theory of best performance of layout v.2 is that resistors R4, R2 have equal temperatures.
Same with R1, R3.


Online antintedo

  • Contributor
  • Posts: 15
  • Country: pl
Re: Statistical arrays
« Reply #51 on: January 14, 2019, 12:58:05 am »
I found some time to perform TCR measurements. So far only 1 sample of 10k NOMCA was tested.
1:2 divider built out of this sample measured 0.4ppm/K which seems to agree with the results below.

Code: [Select]
Sample #1
No  T1      R1          T2      R2          dT      TCRavg  TCR/TCRavg
1   26      10.00243    64      9.99943     38      -7.89   77%
2   26      10.00286    64      9.99750     38      -14.11  138%
3   26      10.00363    64      9.99877     38      -12.79  125%
4   26      10.00189    64      9.99913     38      -7.26   71%
5   26.5    10.00284    63      9.99927     36.5    -9.78   95%
6   26      10.00264    63      9.99840     37      -11.46  112%
7   26      10.00179    64      9.99839     38      -8.95   87%
8   26.5    10.00230    64      9.99862     37.5    -9.81   96%

Sample #2
No  T1      R1          T2      R2          dT      TCR     TCR/TCRavg
1   25      10.00220    67      10.00106    42      -2.71   73%
2   25      10.00235    67      10.00032    42      -4.83   129%
3   25      10.00193    67      10.00057    42      -3.24   87%
4   25      10.00160    67      9.99992     42      -4.00   107%
5   24.5    10.00193    67      10.00021    42.5    -4.05   108%
6   25      10.00160    67      9.99982     42      -4.24   113%
7   25      10.00205    67      10.00074    42      -3.12   83%
8   25      10.00214    67.5    10.00056    42.5    -3.72   99%
« Last Edit: July 04, 2019, 06:44:01 pm by antintedo »
 
The following users thanked this post: Andreas, e61_phil, EmmanuelFaure, doktor pyta, alm, razvan784

Online Andreas

  • Super Contributor
  • ***
  • Posts: 2407
  • Country: de
Re: Statistical arrays
« Reply #52 on: January 14, 2019, 08:25:20 pm »
Thanks for the measurement,
I never had expected that the stray between neighboured resistors is that large.
I had expected more likely a rising T.C. from one side to the other or a center symmetrical behaviour.

with best regards

Andreas
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #53 on: January 14, 2019, 08:41:18 pm »
I would have expected more equal values for t.c. between single resistors as the ceramic area is comparable small so are the resistors and they are manufactured in thinfilm technology, so multiple networks are processed in parallel and films are highly uniform over small areas.

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline Magnificent Bastard

  • Regular Contributor
  • *
  • Posts: 123
  • Country: aq
Re: Statistical arrays
« Reply #54 on: January 14, 2019, 09:34:41 pm »
I found some time to perform TCR measurements. So far only 1 sample of 10k NOMCA was tested.
1:2 divider built out of this sample measured 0.4ppm/K which seems to agree with the results below.

Code: [Select]
No  T1     R1         T2   R2        dT     TCR
1   26     10.00243   64   9.99943   38     -7.89
2   26     10.00286   64   9.99750   38     -14.11
3   26     10.00363   64   9.99877   38     -12.79
4   26     10.00189   64   9.99913   38     -7.26
5   26.5   10.00284   63   9.99927   36.5   -9.78
6   26     10.00264   63   9.99840   37     -11.46
7   26     10.00179   64   9.99839   38     -8.95
8   26.5   10.00230   64   9.99862   37.5   -9.81

These results seem suspiciously variant (not the absolute values, but the relative TCRs).  Can you describe in detail how you made these measurements?
 

Online antintedo

  • Contributor
  • Posts: 15
  • Country: pl
Re: Statistical arrays
« Reply #55 on: January 15, 2019, 12:13:41 am »
These results seem suspiciously variant (not the absolute values, but the relative TCRs).  Can you describe in detail how you made these measurements?
The resistor was soldered on a FR-4 SOIC to DIP converter and attached to a stabilized heater with a pretty large heatspreader in between.
The whole setup was put into a lightly insulated aluminum box.

For every measurement wires were soldered to corresponding DIP holes.
I used 34401A in 4-wire mode to perform 3 measurements: first at ~25C then at ~65C then again at ~25C.
Temperature readout was done with a thermocouple placed near the middle of the package. Plenty of time was given to reach thermal equilibrium after heating/cooling cycle and soldering.

After doing all 8 resistors, I repeated the measurements in the same order as first round of measurements.
I didn't write it down, but I think repeatability of resistor values between two runs was well below 10 ppm, otherwise I would have just discarded the results.
Room temperature measurements after the experiment agreed to 7ppm with measurements taken the day before. I also used VHP101 resistor box as a sanity check.

I think the main downside of this setup was mechanical stability and the need to solder to the board between every measurement, potentially leaving flux and skin contaminants on the surface. I will try to eliminate this and re-test, but it has to wait until the weekend.

I had expected more likely a rising T.C. from one side to the other or a center symmetrical behaviour.
Andreas
I was surprised as well. Out of curiosity I tested one more more sample today and got very different results. TCR was much lower, 3.7 ppm average +/- 1 ppm spread with no obvious patterns. It is certainly possible something was wrong about my initial methodology. This time I paid more attention to handling the board when soldering and fixed the board rigidly using bolts instead of springs.
 

Offline nfmax

  • Frequent Contributor
  • **
  • Posts: 854
  • Country: gb
Re: Statistical arrays
« Reply #56 on: January 22, 2019, 09:29:08 am »
Maybe the effect of thermally-induced strain changes from the mounting arrangement? All resistors are strain gauges!
 
The following users thanked this post: antintedo

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #57 on: April 10, 2019, 04:06:18 pm »
Since I was interested in the question if the NOMCA arrays are useful for a 7V --> 10V boost circuit and noone has yet answered the question by experiment results I've now ordered the boards with my design shown a few posts before. They will receive soon and I will then perform some measurements such as ratio measurements over temperature. If they perform well I can directly use them for my LTZ1000 references.

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 
The following users thanked this post: Andreas, e61_phil, 2N3055, Grandchuck

Offline doktor pyta

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: pl
    • RF Scientific
Re: Statistical arrays
« Reply #58 on: April 13, 2019, 12:03:36 am »
branadic,

I did not have time to test the idea but this is what I came up with (drawing).
Layout for this can be designed symmetrically and I think it could be further improved by paralleling similar dividers.

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #59 on: June 25, 2019, 09:30:32 am »
I was looking at Nomca arrays for the 13:1 divider for a LTZ1000 reference and found similar results as above when using 3x 5K in series with 4x 5K in parallel. Division ratio TC was less than 1 ppm/°C for two units and 2.4 ppm/°C for one unit.

Now, when i look at the measurements presented above by antintedo for the TCs of single resistors in a Nomca array, i'd like to show where one may get by selection. This is only an example, but my conclusion would be that a Nomca array can be used for the 7V to 10V gain stage, if one finds a practical solution for free combinations. As soon as i have time i'll check this with my 5K Nomca arrays.

Regards, Dieter

PS: Sorry for the confusion, but those measurements were not from Magnificent Bastard.
« Last Edit: June 26, 2019, 05:01:13 pm by dietert1 »
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #60 on: June 25, 2019, 10:02:18 am »
Measuring and selecting TC is a slow process. There is an additional difficulty - soldering and board stress can be part of the TC. So it would need TC measurement after soldering and than adapting the layout to the appropriate connections.  :-//
So it is more like a very limited number of option one could chose from by adding bridges, like leaving out one resistor.

Getting 2.4 ppm/K for the divider is quite a bit. The 3 in series and 4 in parallel version is already averaging over 3 and 4 resistors so that chances are low to get such a poor combination with individual resistors at 5 ppm/K spread.

Just for a test one could try re-soldering the bad divider.

The other point is that often the TC is only the easier part. Long term drift may be of similar importance, with little chance to measure and select upfront.
 

Online TiN

  • Super Contributor
  • ***
  • Posts: 4045
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Re: Statistical arrays
« Reply #61 on: June 25, 2019, 10:15:43 am »
Perhaps a socket could be used for testing, to avoid heat stress to the package?
Like this one?
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #62 on: June 26, 2019, 05:22:34 pm »
Measurement of the TCs of the eight resistors in the array requires a 8x multiplexer and one temperature cycle. Yes, maybe the resistor array can't be soldered after the measurement without affecting it's behaviour. That needs to be decided by measurement.
Anyway, results better than 1 ppm/°C without selection may be good enough to try and control the temperature of the resistor array in a way similar to the heated zener reference.

Regards, Dieter
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #63 on: June 26, 2019, 07:53:42 pm »
One normally would not even need 1ppm/K performance in the LTZ1000. Without adjustment (R9) the TC of the LTZ1000 circuit is more like in the 0.2-0.5 ppm/K range which corresponds to some 20-50 ppm/K for the resistors. The individual R9 trimming would also include resistor TC.
However this is only the TC part - it does not help with long term drift.

I don't know how much the soldering effects the TC. Besides TC soldering can effect the short time drift and possibly hysteresis effects. These could be confused with a TC error in an early test. Still a significant drift / error in the TC test is not a good sign, even of not due to actual TC but hysteresis or aging.

I don't think temperature control is helping, as the TC is not the real critical parameter - it's more like aging that is critical and the TC is more used as an indication of stability, with the real thing very difficult to measure. Some burn in after soldering could help - though it would also hide things. Ideally one would measure before burn in and get suspicious if burn in changes too much.
 

Offline imo

  • Super Contributor
  • ***
  • Posts: 1894
  • Country: 00
Re: Statistical arrays
« Reply #64 on: June 26, 2019, 08:11:40 pm »
While looking at the 34470A teardown, at 18:40-22:45 for example, they are using rather common resistor parts around the LTZ1000, imho.
« Last Edit: June 26, 2019, 08:28:14 pm by imo »
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #65 on: June 26, 2019, 08:27:06 pm »
The demands on the resistors around the LTZ1000 are not that high. It is only of one wants really low long time drift (e.g. < 2 ppm/a) that is really takes really stable resistors (still only some 50 ppm/a). AFAIK the 34470 is not the highest stability - more like having the LTZ1000 for low TC and low noise.

The more critical resistors are in a 7 to 10 V stage. Another point could be a gain stage or divider and possibly a resistor in an ADC, that can effect linearity due to self-heating.  Selft heating could increase the temperature of a resistor by something like 0.1-1 K. So an TC of 10 ppm/K could cause an INL error of some 1-10 ppm.
 

Online Andreas

  • Super Contributor
  • ***
  • Posts: 2407
  • Country: de
Re: Statistical arrays
« Reply #66 on: June 26, 2019, 08:29:43 pm »
While looking at the 34470A , at 37:45 for example, they are using rather common resistor parts around the LTZ1000, imho.

Ok as a volt-nut you have to count the vishay VSMP series (around $15 per pop) as rather common.

the TC of the LTZ1000 circuit is more like in the 0.2-0.5 ppm/K range

from where do you have such large values?
my results with long legs on the non-A version are below 0.06 ppm/K

https://www.eevblog.com/forum/metrology/ultra-precision-reference-ltz1000/msg2301414/#msg2301414

with best regards

Andreas
 
The following users thanked this post: imo

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #67 on: June 26, 2019, 08:41:38 pm »
Quote
from where do you have such large values?
my results with long legs on the non-A version are below 0.06 ppm/K

This are numbers I can confirm with 400k resistors unpopulated, short legs.

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Online Andreas

  • Super Contributor
  • ***
  • Posts: 2407
  • Country: de
Re: Statistical arrays
« Reply #68 on: June 26, 2019, 09:13:18 pm »
Quote
from where do you have such large values?
my results with long legs on the non-A version are below 0.06 ppm/K

This are numbers I can confirm with 400k resistors unpopulated, short legs.

-branadic-

ok this confirms my opinion that it is not a good idea to shorten the legs on the non-A Version.
(with half shortened legs I had also a larger T.C. than before).

with best regards

Andreas
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #69 on: July 01, 2019, 02:38:18 pm »
In the meantime i have done some multiplexer measurements on three more Nomca 1603 8x 5K arrays. They got mounted on SMD adapters with gold plated pins and cleaned in ultrasonic bath. Then i setup 4 wire R measurements using a HP3456A at 0,1 mA measurement current with averaging to get below ppm resolution. Temperature span was 18 °C. Error on a single resistor TC measurement is about 0,03 ppm/°C.

Results on TCs for each part (average( 8 ), stdev( 8 ), span( 8 )):

D     4,34 +/- 0,62 ppm/°C, max-min = 1,85 ppm/°C
E   11,23 +/- 0,63 ppm/°C, max-min = 1,74 ppm/°C
F   16,20 +/- 1,55 ppm/°C, max-min = 4,51 ppm/°C

According to my measurements those are pretty serious parts, well suited to produce temperature compensated voltage dividers. And it makes complete sense to measure TCs of individual resistors in order to determine a good combination - better than statistics! Combining  resistors properly each of my three parts should give me a 13:1 divider with less than 0,03 ppm/°C on the division ratio. At least that's the result of a simulation using the measured curves of the individual resistors.

Regards, Dieter

« Last Edit: July 01, 2019, 02:41:29 pm by dietert1 »
 
The following users thanked this post: branadic, antintedo

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #70 on: July 01, 2019, 02:50:52 pm »
Thanks Dieter,

since I'm still waiting for my boards to arrive, because of the inability of DHL or customs the boards were sent back to Elecrow, I ordered a zero force adapter

eBay auction: #282277360455

that received today to perform similar measurements. Need to perpare a cable for multiplexer of my Prema 5017SC.

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #71 on: July 02, 2019, 10:19:48 pm »
Today i got more evidence for the concept of combining selected resistors within a Nomca 1603 array.

The part B that i mentioned above giving 2.4 ppm/°C showed less than 0.5 ppm change during a 18 °C temperature cycle now after rewiring. It's an upper limit because from the multiplexer measurements i expected a residual of -0,015 ppm/°C on the division ratio and i couldn't see anything with our HP 3456A. That part B is the only one in six that exceeds the 5 ppm/°C tracking spec in the datasheet with a span of 6.7 ppm/°C between measured min and max TCs.

Regards, Dieter
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #72 on: July 03, 2019, 06:36:53 am »
Whether selecting the resistors is only practical if soldering / mounting on a board does not change the TC significantly. For PCB use, there are usually also not that many option to choose from. There may be a few jumpers for options but usually only a few, especially if not many of the resistors are unused.

If soldering has only a limited effect it may at least be possible to test the resistors (if the contacts in an affordable socket are good enough) and than decide on the orientation of the chip. This may avoid a worst case (with a suitable use of the resistors) - which could already be quite a gain.
Selecting a suitable resistor combination is much easier with individual resistors - more like hard with an array.
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #73 on: July 03, 2019, 08:23:49 am »
You are not asking, but i will tell you how to do it.

The construction looks like the thin film arrays of the HP 3456A. The Nomca arrays were put onto SMD to DIL Adapters with gold plated pins. On the carrier board there are gold plated socket strips. Wiring is underneath the carrier board with short soldered bridges, that i prefer over the plugged configuration bridges of the HP3456A. This construction isn't ideal in the sense that is has short copper tracks and wires and the pins/sockets in the path that are not tracking well during fast temperature changes, but the arrays are not soldered anymore after characterization.

I used a similar construction before with S102 resistors soldered on top of 4 pin modules so that one can use 4 wire measurements for characterization and not solder them afterwards. In that case the doubled pins give some redundance when the parts are on the carrier board afterwards. I also thought about making an array of S102 resistors and see how they compare. Individual TCs will be smaller, but temperature tracking will be worse, so finally the result may be similar.

Regards, Dieter

PS: One may even think about using TSSOP32 adapters for the Nomca arrays in order to have 4 wire connections.
« Last Edit: July 03, 2019, 08:31:58 am by dietert1 »
 

Offline Edwin G. Pettis

  • Frequent Contributor
  • **
  • Posts: 413
  • Country: us
  • The plural of anecdote is not data.
Re: Statistical arrays
« Reply #74 on: July 03, 2019, 06:40:29 pm »
The NOMCA arrays are not particularly good for certain applications:

!. Ta2N resistors are rather noisy and has 1/f noise as well, all above the theoretical resistor noise level.
2. While short term stability is not bad, long term stability is questionable especially at low PPM.
3. Shelf stability at 25°C is only fair, tolerance drift at up to .01%, ratio drift up to .002% for the first year, Vishay states no specification beyond 1 year.  If these exhibited good long term stability you can be sure Vishay would have put that spec in there.
4. Ratio matching of 5 PPM/°C is hardly impressive, other types can be below 1 PPM/°C including wire wound.
5. All factors affecting the specifications more or less apply to any resistor type being used.  This requires careful design to manage thermals and unequal heating even on a single substrate.
6. Sensitivity to soldering, per the data sheet, possible significant (and permanent) changes can occur, not only to tolerance but TCR and stability.

In the case of the current discussion on a 7V to 10V boost circuit, any deviation of resistor matching will be multiplied by the amplifier gain, this includes noise and drift of resistor specs over time.  While a NOMCA array may have an attractive price, its specifications for this particular application leaves a lot to be desired.  If you are expecting a long term stability and low noise 10V output over months, these are not the resistors to pick.  The Vishay data sheet is pretty clear on the specifications they do give, NOMCAs are not for this application.

S102 resistors also stink for this application.
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #75 on: July 03, 2019, 09:58:03 pm »
Sorry, i would rather prefer learning from industry leaders who have been applying thin film technology in precision measurement for many years.

I got somewhat impressed that 6 $ plus proper wiring of that part gets me a voltage divider with less than 0.5 ppm change of the division ratio during an 18°C oven cycle. Which is less than 0.03 ppm/°C. And this is the worst of six parts and still so good that i can't really measure the residual change. I know that number doesn't appear in the data sheet. Please don't tell me you know how long the adjustment will last if i don't touch the part anymore. I think right now nobody can tell.

Regards, Dieter
 

Offline Edwin G. Pettis

  • Frequent Contributor
  • **
  • Posts: 413
  • Country: us
  • The plural of anecdote is not data.
Re: Statistical arrays
« Reply #76 on: July 04, 2019, 02:32:50 am »
Well dietert1, considering that you haven't been playing with this part for very long, you don't know what the longer term drift is and since Vishay doesn't specify longer term drift, you can accept that if Vishay doesn't provide longer term drift, there is no significant improvement over time or they would have said so.  I know quite well the performance limitations of many resistor technologies and I also know very well how well my precision wire wound resistors perform.

Carefully read the NOMCA data sheet from Vishay, the information I quoted is directly from their own data sheet, so no you don't have to believe what I said because it is in their data sheet quite plainly.

As far as resistors go, I have nearly 46 years in the resistor business, I think I know a tad more about the subject than you think.....I have the instruments that can measure resistance better than most anybody here and to a very high accuracy traceable to NIST, my uncertainty is 0.15 PPM on my primary reference.

However, you are most welcome to believe whatever you wish.
« Last Edit: July 04, 2019, 06:10:44 pm by Edwin G. Pettis »
 

Online TiN

  • Super Contributor
  • ***
  • Posts: 4045
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Re: Statistical arrays
« Reply #77 on: July 04, 2019, 04:46:17 am »
Edwin G. Pettis

Quote
my uncertainty is 0.15 PPM on my primary reference.
Aww, beat me by 0.01 ppm there  ;D

Jokes aside, how possible would it be to order small qty (10-20?) of PWW network for purpose of 7V/10V and such, but with custom values per network? I'd be interested, and probably some others too to make it into a batch? One could find out precise values using calibrated decades like RS925, and then specify final divider, like 7.1426kohm/20.5216kohm as example. Of course, each network's precise value would be different, matched to specific LTZ circuit.  :popcorn:
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
 
The following users thanked this post: Theboel, MiDi

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #78 on: July 04, 2019, 05:48:07 am »
Well, if the TDP / NOMCA networks were really that bad in aspect of longterm stability and noise I'm pretty sure they wouldn't be used in Wavetek/Fluke 7000 and 7001 and if TaN resistor networks were bad in general Fluke wouldn't have used them in 732B and 732C and replaced the former single wirewound resistors by resistor network. So please not another flame war on resistor technology.

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #79 on: July 04, 2019, 06:22:21 am »
Sorry Mr. Pettys, we understand that you prefer wire wound resistors and you are not at all interested in thin film technology or statistical arrays.
At least one of your statements above was plain false. Analog Devices offers the AD587 reference part, which is well known for its good long term stability (Geller reference). Know what, the 10 V to 7 V divider within that part is based on thin film technology. No wire wound resistors there.

Resistor arrays offer exactly what TiN is asking for: A configurable voltage divider that supports many different division ratios and pretty fine adjustment. And yes, this needs implementation and measurement instead of general wisdom and reading between the lines.

Regards, Dieter
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #80 on: July 04, 2019, 02:08:04 pm »
The specs for long term stability (some are shown in the environmental tests) are not that good. However with something like long term aging this is usually only a best guess, nothing easy to measure and check for every unit.  I am not 100% sure one can use the TDP series to directly use for the SMD version. The SMD version is much more sensitive to board stress.
Especially with the DIP versions, Fluke may have done some selection up front, to get better than typical units.

Especially in a cost critical or small size case wire wound resistors may not such a good alternative. Even if only replacing 2 resistors such an SMD array can be cost and size effective. For some cases there may be better alternatives like the MORN series (4 resistor in SSOP8) with better specs.
When using several resistors in a series and / or parallel configuration the relative TC matching is expected to improve (at least on average and in most cases). However one can still have bad luck and get a relatively poor matching.

The resistor excess noise may in some cases become a problem - but at least from the specs (< - 30dBi) the NOMCA arrays are not that bad, despite the small size. For a LM399 based reference the extra noise should not be a big issue - for a LTZ1000 based reference, one may notice some additional noise. For the resistors directly at the LTZ1000 excess noise should not be relevant with thin film resistors.
The TaN Arrays HP was / is using in there DMMs tend to be larger area with thus a chance for lower noise and possible better matching.
 

Offline Edwin G. Pettis

  • Frequent Contributor
  • **
  • Posts: 413
  • Country: us
  • The plural of anecdote is not data.
Re: Statistical arrays
« Reply #81 on: July 04, 2019, 04:02:58 pm »
My statements were directly based on the Vishay NOMCA data sheet, there was no interpretation involved.  Comparing the AD587's internal resistors to the NOMCA is not valid, they are not identical even if they are using the similar Ta2N resistors which is generally used in most ICs.  That is the same as claiming all precision wire wound resistors are the same which they are definitely not.  Just because I specialize in wire wound technology doesn't mean I don't know how the other resistor technologies work.

The networks you refer to in the the 7000 are not identical to the NOMCA arrays, the manufacturing processes are not the same so you cannot compare the two.

I specifically stated that in my opinion, the NOMCAs are not well suited to a voltage booster circuit requiring low PPM performance.  So far all I've seen is mainly measurements on the arrays and none done on an actual booster circuit, those are two entirely different operating conditions.  Let's see what your measurements come up with in an actual booster circuit first before throwing rocks at the windows.

For that matter, I have made matched tracking TCRs of resistors down to <0.2PPM/°C which has been independently verified over long term period.  No it is not easy to achieve, the physical circuitry details are just as important, if not more important, than the resistors themselves.  Creating a very low tracking TCR depends on a lot more than just the resistors themselves, thermals can be very complex to control and can completely wipe out the low tracking TCRs of the resistors.
 

Offline Magnificent Bastard

  • Regular Contributor
  • *
  • Posts: 123
  • Country: aq
Re: Statistical arrays
« Reply #82 on: July 04, 2019, 05:34:41 pm »
The AD587 (et. al.), as well as the LT5400 are made with silicon-chromium resistors that are passivated with silicon-nitride.  The ratio of silicon-chromium to silicon-nitride is adjusted to minimize absolute TCR.  The Fluke 7000 series uses Vishay TDP networks, which are nickel-chromium elements that are passivated with tantalum-nitride.  Same as above, they tweak the ratio of nickel-chromium to tantalum-nitride in order to minimize the absolute TCR.  NOMCA are made with a 100% tantalum-nitride process, and they are self-passivating.  During the tantalum sputtering process, the percentage of nitrogen in the chamber is controlled to minimize the absolute TCR.  Because of the way they are made; where the resistors are sputtered onto a substrate in a controlled environment; adjacent resistors will track very well in TCR (and hence are valuable as a ratio device).  In resistor networks, the resistors have a tendency to track each other over time because resistor drift is driven by temperature, and the resistors on a network are nearly the same temperature.

Most of the evils that engineers have experienced with networked resistors have to do with the thick film process, which is totally different than thin-film or metal film.  The data sheets for these metal/thin film parts are VERY conservative.  The tracking TCR is related to R1 against any other resistor in the network (over 6-sigma); but resistors that are next to each other (and closer to the center of the network) WILL track each other far better than data-sheet specs say they will; and the networks can be selected for additional needed performance.  Once built and operating, the ratio of such a network will track within 1ppm/a, but usually much better than that.  You DO have to follow some common-sense rules when you are designing such a ratio device.  Even though two networks come from the same batch, they often TCR track very well, but you cannot rely on this; only that resistors within an individual network will track.  You have to make certain that each network in a multi-package divider shares responsibility for both the upper and lower legs of the ratio divider.  The resistors closer to the ends of the network (R1/R8 or R1/R7) are not going to track as well as the resistors closer to the center of the network; and this is all due to package strain.  A statistical array of at least 3 networks will perform better than just one network...  (and so on and so forth...)  You can figure out all of these rules on your own through testing; but don't forget to test for humidity, which can have bigger effects than temperature.
 
The following users thanked this post: doktor pyta, razvan784

Offline try

  • Regular Contributor
  • *
  • Posts: 90
  • Country: de
  • Metrology from waste
Re: Statistical arrays
« Reply #83 on: July 04, 2019, 06:44:15 pm »
As far as resistors go, I have nearly 46 years in the resistor business, I think I know a tad more about the subject than you think.....I have the instruments that can measure resistance better than most anybody here and to a very high accuracy traceable to NIST, my uncertainty is 0.15 PPM on my primary reference.

Well, Edwin, claiming 46 years in the resistors business and having instruments that can measure resistance better than most anybody does not seem to be sufficient at all when it comes to delivering 10k Ohm flat with 10 ppm precision to a customer.

OK, mistakes can happen.
But refusing to deliver a free replacement shipment upon delivering the requested measurement proof is pretty disappointing.

« Last Edit: July 04, 2019, 06:47:58 pm by try »
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #84 on: July 04, 2019, 07:23:33 pm »
While TCs within a Nomca package resemble better than TCs between different packages, in my measurements i have not seen that resistors close to each other resemble better. Neither have i seen that "inner" resistors are matching better than resistors near the ends of an array.

Apparently resistance tolerances within each array are very small, like +/- 0,005 %. This means it is very difficult to tune a Nomca divider for division ratio by using a certain combination of it's resistors. So when making a 7 -> 10 V booster with a Nomca array, additional tuning resistors are necessary to get close. A usual 1% 100 ppm/K MF resistor that one may have around is not good enough. If it gets its TC attenuated by 200 it will still ruin anything tuned to 0,1 ppm/K or better.

The reference booster resistor array of the HP 3456A has lots of different resistor values organized like a binary DAC. They get so close that in the end a usual  potentiometer is good enough for tuning.

Regards, Dieter
 

Online antintedo

  • Contributor
  • Posts: 15
  • Country: pl
Re: Statistical arrays
« Reply #85 on: July 04, 2019, 07:34:14 pm »
While TCs within a Nomca package resemble better than TCs between different packages, in my measurements i have not seen that resistors close to each other resemble better. Neither have i seen that "inner" resistors are matching better than resistors near the ends of an array.
Thanks for your input, that was my observation as well. Since chip layout of NOMCA is not symmetrical, did you perhaps observe any patterns? For example in my test of 5 samples the resistor between pins 2 and 15 was almost always an outlier.

I will try to eliminate this and re-test, but it has to wait until the weekend.
So much for that - I'm still trying to setup a decent thermal chamber with a controller. However I tested a few more samples with my previous setup. The TCR pattern between resistors was similar but tracking much better. I edited the original post to include the results of 1 more sample that I had time to check for repeatability.

Regarding recent discussion, I was curious about 1/f noise of thin films, especially NOMCA vs larger models. I did a quick test dividing LTZ1000 voltage by 2 and measuring 0.1-10Hz pp noise within 1 minute period.

LTZ1000: 0.15 ppm
noname THT 3.3k metal film: 0.16 ppm
NOMCA 10k, 3 samples tested, 6 resistors used every time: 0.22-0.25 ppm
S102K (refurbished) 11k, 6 resistors used: 0.16 ppm

Additionally, long time ago I found this publication containing 1/f noise tests of various resistor types and models. Seems like a lot of models achieve good performance with almost no 1/f noise, even in SMD 0402 size.
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #86 on: July 04, 2019, 09:14:40 pm »
My Nomca TC measurement results.
There is one measurement (R7 in array "B") that exhibits hysteresis. So that measurement may have some problem and be the reason why that part violates the 5 ppm tracking spec. I'm already using the part as a 13:1 divider for a LTZ1000 reference, leaving R7 unconnected.

Regards, Dieter
 
The following users thanked this post: antintedo

Offline Edwin G. Pettis

  • Frequent Contributor
  • **
  • Posts: 413
  • Country: us
  • The plural of anecdote is not data.
Re: Statistical arrays
« Reply #87 on: July 04, 2019, 11:15:21 pm »
Since Try can't seem to get on with it:

Appears Try is still suffering from a malady mostly of his own making.  The original order was shipped in December, 2016, and while I tried to talk him out of ordering ±10PPM resistors he insisted, note that he requested no additional special handling of the resistors, even so I still guaranteed them to meet transfer standard specs (ESI) which I later sent him a copy of the specs which using his own numbers was still within spec.  As is standard industry practice, the resistors are guaranteed to meet said initial specifications before leaving the factory.  The initial readings were indeed -2PPM and -6 PPM when shipped.

I received this reply from Try on 1-18-2017: “I measured the one you marked 10k -2ppm. It was either spot on or exactly 10k + 0ppm on my 34401A with four wire measurement.”  He did not mention the other resistor and I was rather surprised at how close his meter read it, that should mean his meter also read the other resistor at -6PPM.  I did not hear again from Try until nearly 14 months later when he wrote to say he thought that both resistors had drifted some but could not get a calibrated reading on them, using several nondescript resistors and an out-of-cal (5 years) 3458A…..that’s okay it should be in the ball park at least.  The figures he gave me was somewhere around 20-30 PPM higher than original.  I offered to re-calibrate them for the cost of postage so he would have accurate readings again, he refused and claimed they were out of spec…..well yes they were probably not within ±10PPM by this time but as they were standard processed resistors that was not unexpected plus there was no way of knowing what had happened to them in the ensuing 14 months.

The standard guarantee for a transfer standard is that the standard will maintain it’s given tolerance for 60 days after manufacture assuming it was handled properly.  There was another batch of e-mails at a much later time which mostly just rehashed what had been said earlier.

I believe the offer of re-calibration was sufficient and reasonable, since Try insisted only on replacement a good year and a half later, I would only offer replacements if the resistors were beyond specifications for a transfer standard of which I had sent him a copy of the standard specifications and guarantee.

He also complained about the pricing, a mere $28.38 for ±10 PPM resistors, I think he got more than he paid for.  All he really needed was calibration of the resistors to accurate known values and he’s back in business with calibrated references.  Now if he thought he was going to get resistors that would hold ±10 PPM indefinitely and for that price, I think he was just off his gourd a bit.

I will still do the re-calibration for the cost of postage even now but I cannot be held responsible for his turning down the offer repeatedly.
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #88 on: July 05, 2019, 06:27:12 am »
Mr. Petty, considering your story, i understand the measurements try showed above are those same two resistors you mentioned and that they exhibited a difference of 3.7 ppm between each other after 18 months while you determined 4 ppm in the start? So this means the resistors - though worse than expected - were still useful for making a voltage divider, where a possible 40 ppm drift cancels.

I guess the same applies to the Nomca parts i tested. I got those from Mouser two weeks ago and we may probably assume they have been in stock for years. And they still exhibit tracking with 0,6 to 1 ppm (stdev of 8 resistors). Isn't that remarkable?

If there is anybody with a similar set of wire wound resistors from Mr. Petty, what is the stdev after some years? The result will depend somewhat on how the parts were treated/used. And i would really like to know the TCs of those two resistors mentioned above ..

Regards, Dieter
 

Offline MiDi

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
Re: Statistical arrays
« Reply #89 on: July 05, 2019, 11:25:52 am »
A bit off-topic, but seems to fit here:

Could a precision DAC used in "multiplying" configuration be a serious replacement for precision custom resistor dividers/networks in boost amplifier (e.g. AD5781/-91)?

Relying upon Datasheets, these DACs seem to outperform even the best resistor dividers/networks and they are ootb trimmable down to ppm...
« Last Edit: July 05, 2019, 11:48:27 am by MiDi »
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #90 on: July 05, 2019, 01:30:30 pm »
Yes, i was waiting for that. Just combine a AD587 reference with a PT1000 temperature sensor, a humidity sensor, a Cortex M0 microcontroller and some smart firmware to automatically fine tune the reference including support for teach-in. That should result in a near perfect voltage standard without heating and with slow aging. The firmware could even predict and compensate aging drift.

So lets stop putting wire wound resistors into dry boxes, TEC boxes etc. and do some real work instead.

Regards, Dieter

PS: I just have a 10 KOhm econistor in the dry box. Room humidity is and was 32 % for hours, while temperature increased about 1 °C. While from the temperature increase i would expect a resistance change of about 2.2 ppm the part exhibits a resistance change of -49 ppm after drying for some hours. Anybody wants to fill their voltage reference box with silica gel?
« Last Edit: July 05, 2019, 02:13:24 pm by dietert1 »
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #91 on: July 05, 2019, 04:22:55 pm »
.....
PS: I just have a 10 KOhm econistor in the dry box. Room humidity is and was 32 % for hours, while temperature increased about 1 °C. While from the temperature increase i would expect a resistance change of about 2.2 ppm the part exhibits a resistance change of -49 ppm after drying for some hours. Anybody wants to fill their voltage reference box with silica gel?

Silca gel is not just for lowering humidity. If left in normal air for some time it mainly works to buffer humidity changes - so that rel humidity would change less/slower, even in a not so hermitic sealed case.

With the NOMCA arrays in an epoxy type case - they may react to humility changes too. With wires wound types it really depends on the materials used - nor wire wounds are equal.
 

Offline Edwin G. Pettis

  • Frequent Contributor
  • **
  • Posts: 413
  • Country: us
  • The plural of anecdote is not data.
Re: Statistical arrays
« Reply #92 on: July 06, 2019, 05:36:23 pm »
This is the data on the SR-1 transfer standards, not only do my resistors exceed these specs but the cost is far lower.  Notice their initial accuracy is ±20PPM and calibration is only accurate to ±10PPM.  You also appear to not understand the definition of initial accuracy: “Initial Accuracy: The specifications stated in the TEGAM instrument catalogs and data sheets are intended as acceptance specifications and are guaranteed for 60 days from the date of shipment.”  You seem to think that initial accuracy is held indefinitely, no transfer standard does that.


http://www.ietlabs.com/pdf/Datasheets/TegamSR1.PDF

Note the initial accuracy is ±20 PPM and long term is ±50 PPM and calibration is only to 10 PPM, according to Try even after 2.5 years they are still within specifications.  He has nothing to complain about and I even offered a free calibration which he refused twice so I think it is time Try moved on.  He never presented any evidence that the resistors were out of spec.  He also bought a lot of other resistors from me and has never complained about those.  My standard off-the-shelf resistor exceeds the SR-1 specifications and for a lot less money and without post operation processing.

Just for your information, some time back Vishay was chasing after me to become a field applications engineer for them, Felix Zandman interviewed me himself, I turned them down in the end for personal reasons.  The job entailed flying all over the country and at that time I could not do that.  Vishay knew that I had the knowledge they needed, they approached me.  Your erroneous claim that I don’t know about Vishay technology is full of hot air.
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #93 on: July 06, 2019, 05:43:33 pm »
Come on, this thread is about statistical arrays and not about Edwin G. Pettis, if you don't have anything to contribute please be silent., Thanks.

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline Edwin G. Pettis

  • Frequent Contributor
  • **
  • Posts: 413
  • Country: us
  • The plural of anecdote is not data.
Re: Statistical arrays
« Reply #94 on: July 06, 2019, 06:08:22 pm »
I'd be happy to Branadic, I didn't start this nonsense, dietert1 and Try interjected all this nonsense so please feel free to jump them first, it is only fair that I get to re-but them.

Measuring resistors on a DVM  is not a valid technique for determining how a given resistor will perfect in a given circuit under actual operating conditions.  Yes you get a bunch of data on how it is performing under test conditions which may have some indirect bearing on the circuit performance but if you want to know how that divider is going to work under actual conditions, make an actual circuit and try it, see what the end result is, then if it isn't performing as expected you can trace the source.  A resistor under actual operating bias and environmental conditions is not the same as hooking it to the front end of a DVM.
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #95 on: July 06, 2019, 06:23:21 pm »
Dear Mr. Pettis, from your table i understand that buying two resistors from you to make a voltage divider will result in a division ratio with up to 10 ppm/K on the division ratio. This worst case happens if one resistor has +5 ppm/K, the other one -5 ppm/K and it means up to 180 ppm change on the division ratio in an 18°C temperature cycle.
I think i demonstrated above that a Nomca thin film resistor array, when wired correctly, will give you a small fraction of a ppm/K on the division ratio and an overall variation of less than 0.5 ppm with a 18°C cycle. The thin film array is a much superior solution, about 360 times better. Hard to understand what you want to contribute.

In the meantime i finished drying the 10 KOhm econistor (my example of a qualified wire wound resistor). The final result of the humidity test going from 32% RH down to about dry was -60 ppm after about 24 hours. Only then the econistor exhibited a hysteresis free 18°C cycle with a TC of 2.67 ppm/K. The negative sign of the 60 ppm resistance change is interesting. It means this huge humidity effect is not surface conductivity, but probably carrier size. Carrier gets bigger by water absorption, wires get stressed to be slightly longer and slightly thinner. Carrier shrinks when dry and resistance shrinks as well.
Started a similar test on Nomca thin film array, but drying is very slow. Effect on resistance is -12 ppm after 8 hours, yet only 0.21 ppm on division ratio (with 13:1 divider example).

Regards, Dieter
« Last Edit: July 06, 2019, 06:29:32 pm by dietert1 »
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #96 on: July 06, 2019, 07:01:46 pm »
With the comments on wire wound resistors a possible important point got slightly hidden. The NOMCA arrays seem to have noticeable noise:
....
Regarding recent discussion, I was curious about 1/f noise of thin films, especially NOMCA vs larger models. I did a quick test dividing LTZ1000 voltage by 2 and measuring 0.1-10Hz pp noise within 1 minute period.

LTZ1000: 0.15 ppm
noname THT 3.3k metal film: 0.16 ppm
NOMCA 10k, 3 samples tested, 6 resistors used every time: 0.22-0.25 ppm
S102K (refurbished) 11k, 6 resistors used: 0.16 ppm

Additionally, long time ago I found this publication containing 1/f noise tests of various resistor types and models. Seems like a lot of models achieve good performance with almost no 1/f noise, even in SMD 0402 size.

This essentially means the arrays are not good enough for a 7 to 10 V step for a LTZ based reference. The resistor noise in this case seems to be about as much as the LTZ, even with using 3 resistors in parallel each.
They may still be OK with a LM399 based reference, or to set the temperature in an LTZ1000 circuit, at least from the noise side. A still open point is aging. Specs here are tricky, as there is no good and easy way to measure upfront and small differences in the protective layer / epoxy could make quite some difference.

Though not specified much different, there are also NOMCT arrays, using NiCr resistors. These may be better with TC matching / noise, but from what I have read so far could show a little more aging.

With wire wound resistors, there are also set available with TC matching better than the absolute TC. Even without this, chances are good that at least the 2 nd order part will compensate. Especially similar value resistors can be made from the same wire spool to give good matching.

With the resistor arrays the TC matching can on average improve with the statistical use of more resistors, but the worst case does not improve that much. Best case there are 4 resistors on each side. This would be on average an improvement by about a factor of 2 (maybe a little more as there is finite number of resistors and thus more gain than the normal square root N). Without individual matching / selecting the expected divider TC is thus more like 2 ppm/K, maybe often better if the initial specs are rather conservative or for a larger temperature range.
 
The following users thanked this post: TiN, Kosmic

Offline Edwin G. Pettis

  • Frequent Contributor
  • **
  • Posts: 413
  • Country: us
  • The plural of anecdote is not data.
Re: Statistical arrays
« Reply #97 on: July 06, 2019, 07:04:03 pm »
No you are totally incorrect, you are confusing tolerance with TCR, the tolerance of those two resistors were within ±10 PPM as requested, the TCR would have been <1.0PPM°C, probably close to 0.85 PPM/°C, but that is not the tracking TCR, since both were made of the same wire the tracking TCR (ignoring external factors such as thermal differentials) would be likely less than 0.5PPM/°C (conservatively under real operating conditions)).  The tracking is affected by external environmental conditions which I have no control over and yes film resistors would have similar effects even on the same substrate but possibly smaller because they are much smaller physically but they are also affected by factors which have little to no effect on wire wounds.  Tracking is a very difficult specification to work out since it involves much more than just the resistors, each situation is different.

Yes you very well may be getting pretty close tracking on your NOMCA chip under those conditions, I am not particularly disputing your results, the question is how stable is it over time, your one off only shows that this one chip has these current specifications at this time.  The TCR and tracking TCR of my resistors do not shift significantly over time unless the resistors are subjected to extraordinary conditions.  Your calculations concerning my parts are quite incorrect.  I have demonstrated 0.1PPM to 0.2PPM ratio tracking over long term time by one of my customers, that is not to say everyone is going to achieve that low tracking value, it takes effort on the part of the customer as well.

Econistors are not equivalent to mine, their specs are not equal and as stated, all precision wire wound resistors are NOT the same.  If you are going to compare something against my resistors, then use my resistors but stop making erroneous comparisons thank you.

The table I posted is NOT mine, it is the SR-1 transfer resistor data table from ESI/Tegam/IET, my specifications are better than those which I said in the other posting, please read it more carefully.

It is this kind of misinterpretation which causes problems, it would be much better if you asked a question directly instead of posting first without comprehension.  Yes resistors, like most other components are much more complex that most people understand and when you're chasing PPMs, it gets even more difficult.
 

Offline Edwin G. Pettis

  • Frequent Contributor
  • **
  • Posts: 413
  • Country: us
  • The plural of anecdote is not data.
Re: Statistical arrays
« Reply #98 on: July 06, 2019, 07:18:23 pm »
In general I agree with you, Kleinstein, those statements are accurate.
 

Online Andreas

  • Super Contributor
  • ***
  • Posts: 2407
  • Country: de
Re: Statistical arrays
« Reply #99 on: July 06, 2019, 08:00:56 pm »
In the meantime i finished drying the 10 KOhm econistor (my example of a qualified wire wound resistor). The final result of the humidity test going from 32% RH down to about dry was -60 ppm after about 24 hours.

Hello,

usually I measure time constants for drying of epoxy in the order of 3-7 days at room temperature.
So it would be interesting what happens after the 24 hours.

Lars also stated 0.5-2ppm change for 8E16 resistors for each % relative humidity.
https://www.eevblog.com/forum/metrology/t-c-measurements-on-precision-resistors/msg1379236/#msg1379236

when I compare the humidity in Colorado with that of my region here I have minimum 30% more humidity.

with best regards

Andreas

 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #100 on: July 06, 2019, 09:33:58 pm »
No, Mr. Pettis, it's not April 1st today. If you present a table with limits you will stick to, that's it. We know from TiNs measurements that your resistors are no better than econistors. My econistor humidity test was to confirm what others found before. Yes, the econistor remains in it's dry pack and i will measure it again after a week or so. I think Mr. Pettis is creating a lot of noise without any signal.

I was not measuring one Nomca part but six, a total of 48 resistors. Sooner or later i will also characterize the other 34 econistors we have, looking at their TCs. And i wrote about bridge measurements were i had one 13:1 Nomca divider outside and one inside the TEC box, both running from the same 10 V reference. That was the measurement which resulted in the +/- 0,5 ppm/K limit for a 18°C cycle. So it wasn't a DVM resistance measurement only, but very close to the indended application.

Concerning resistor noise: Is there a link/reference on how those measurements were done? I mean before someone talks us into using parts with 60 ppm humidity dependence in order to avoid some 0,07 ppm extra noise, i'd like to have a closer look. Maybe a Nomca array just needs to be protected from random ventilation. I mean its thermal mass is very small in comparison to its surface.

Regards, Dieter
« Last Edit: July 08, 2019, 02:12:37 pm by dietert1 »
 
The following users thanked this post: 3roomlab

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #101 on: July 08, 2019, 05:32:52 pm »
Dieter,

I'm about to measure some values soon, still preparing everything. Attached one NOMCA16035001 together with a NCP15WF104F03RC NTC on a ceramic board, configured as a 10k, with the board glued to a BPR10101J resistor as part of an oven.

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 
The following users thanked this post: TiN, MegaVolt, antintedo

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #102 on: July 08, 2019, 05:51:52 pm »
....
Regarding recent discussion, I was curious about 1/f noise of thin films, especially NOMCA vs larger models. I did a quick test dividing LTZ1000 voltage by 2 and measuring 0.1-10Hz pp noise within 1 minute period.

LTZ1000: 0.15 ppm
noname THT 3.3k metal film: 0.16 ppm
NOMCA 10k, 3 samples tested, 6 resistors used every time: 0.22-0.25 ppm
S102K (refurbished) 11k, 6 resistors used: 0.16 ppm

Additionally, long time ago I found this publication containing 1/f noise tests of various resistor types and models. Seems like a lot of models achieve good performance with almost no 1/f noise, even in SMD 0402 size.

The link in this older post is quite good. The measurement for resistor 1/f noise is relatively easy with a bridge from 4 equal resistors, as described in the link. It mainly take a low noise amplifier, some instrument to record the data at a moderately low frequency. The MIL standard uses a slightly different setup, with only one resistor as a DUT and some fill resistor, but its a little more tricky, especially at very low noise.
So a test may be reasonably easy.

From the measurements quoted the NOMCA resistors as a 1:1 divider (3 x 10 K in parallel each) results in noise similar to the LTZ1000 noise in the 0.1 -10 Hz range  (noise power about doubles).

The < -30 dBi noise specs for the NOMCA are not really good. From the divider measurements they don't seem to be much better than specs.
For the temperature setting divider at the LTZ1000 this would still be well good enough, but not for a 7 to 10 V stage.
 

Offline Kosmic

  • Frequent Contributor
  • **
  • Posts: 812
  • Country: ca
Re: Statistical arrays
« Reply #103 on: July 08, 2019, 06:36:12 pm »
Dieter,

I'm about to measure some values soon, still preparing everything. Attached one NOMCA16035001 together with a NCP15WF104F03RC NTC on a ceramic board, configured as a 10k, with the board glued to a BPR10101J resistor as part of an oven.

-branadic-

Nice assembly  :-+

Is it expensive to order some custom ceramic boards like that ?
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #104 on: July 09, 2019, 05:14:17 am »
Yes, this assembly will certainly perform very well. The large resistor will provide distributed heating with small gradients. But this type of work really requires some sturdyness, since it will take a long time, maybe a year or more to learn how useful such a device can be and whether it helps.
I would also like to mention again SEI RTAN resistors that could be used to do something similar, i mean make an array in a thermostat. Those parts appear to be metrology level, too.

The measurements i started on drying Nomca arrays turn out to be much more difficult than the econistors. I have not seen any humidity effect now after several days, which is good news. The temperature effects are much bigger, even with a part ("C") with only 5 ppm/K on average. I also put a 5 KOhm UPW50 resistor into the drypack and did not see a single ppm change after 24 hours. Apparently our econistors are sealed only on one end and open on the other. And their material is somehow softer, while the UPW50 package appears more like usual IC packages - extremely dense plastic.

Regards, Dieter
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Statistical arrays
« Reply #105 on: July 09, 2019, 04:16:30 pm »
Quote
Is it expensive to order some custom ceramic boards like that ?

Can't tell you that as this is a "design" I gave to my colleague to develope the process on a real world application and without the use of individual masks (digital process chain).

-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #106 on: July 21, 2019, 09:14:35 pm »
After about 2 weeks in the drypack a Nomca array exhibits a humidity effect of about 25 ppm on the resistance. Yet the division ratio of a 13:1 divider changed by only 0.7 ppm under these conditions.
After 2 weeks in the drypack, some wirewound resistors have gone mad and exhibit changes of up to 220 ppm (econistor) and about 80 ppm (UPW50). Changes of econistors differ by more than 30 ppm, so useless for metrology grade voltage dividers. Hard to believe, need to check more.

I also found that the TC compensation scheme proposed above using a certain combination of resistors is delicate. I mean the TCs i determined using a 18 °C oven cycle (using box method) are useful but may not give the best possible result at room temperature. Currently i am trying to refine temperature measurements using a PT1000 sensor close to the Nomca array in order to separate humidity effects from temperature effects.
I also suspect that self heating may affect the compensation scheme.

Regards, Dieter
 
The following users thanked this post: doktor pyta, The Soulman

Offline d-smes

  • Regular Contributor
  • *
  • Posts: 88
  • Country: us
Re: Statistical arrays
« Reply #107 on: July 22, 2019, 09:12:25 am »
After about 2 weeks in the drypack ...
By drypack, I assume you mean a desiccant material.  Are you just throwing a few pouches in a metal box?  Or completely encapsulating the resistors in a sealed jar or equivalent?  Any relative humidity readings?  I know with insulation systems, there is such a thing as too dry.  Wondering if that applies here.
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #108 on: July 22, 2019, 10:32:29 am »
When we got our 3D printer, we bought 4 kg of orange silica gel to make a drybox for filaments. That drybox got about 1 kg in it and the hygrometer has been reading 10 or 11 % rH ever since. That's the minimal reading of the hygrometer (some technical or physical limit). I just put some spoons of that silica gel into plastic envelopes together with the resistors. The envelopes are closed except where the cables enter. I think humidity inside has been below 11 % rH for at least 14 days.
The results are somewhat confusing. During the first 24 hours 10 KOhm, when i was watching them regularly, econistors exhibited a resistance change of about -70 ppm, but now they are about 150 ppm above the initial value. I think i will take them out of the drypack and test whether they go back to initial values or whether they have been damaged. The UPW50 is 80 ppm above initial value after 14 days.

Regards, Dieter

PS: Now i got a Sensirion SHT31 Smart Gadget Development Kit that indicates less than 5 % RH inside of the drypack (PE bag).
« Last Edit: August 07, 2019, 10:19:59 am by dietert1 »
 

Offline 3roomlab

  • Frequent Contributor
  • **
  • Posts: 919
  • Country: 00
  • can you tell me how to cal sesame street meter?
Re: Statistical arrays
« Reply #109 on: July 26, 2019, 07:40:49 am »
In the nomca pdf, it states the load life drift at 125C (looks like derated = 100mW?) is 290ppm. what should be the expected/calculated load life drift be at 1mW? 2.9ppm (125C)?
spheres of influence, example linustechtips. can you feel the brainwashing? showing off equipment, etc. were you swayed and baited? with immense popularity (and social "titles"), can you afford to disagree?
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #110 on: July 26, 2019, 08:05:56 am »
For the load life it is likely the temperature that is most important So it would likely not make a difference if with current or with external heating.
 

Offline splin

  • Frequent Contributor
  • **
  • Posts: 806
  • Country: gb
Re: Statistical arrays
« Reply #111 on: July 26, 2019, 01:33:20 pm »
In the nomca pdf, it states the load life drift at 125C (looks like derated = 100mW?) is 290ppm. what should be the expected/calculated load life drift be at 1mW? 2.9ppm (125C)?

Take a look at this Vishay technical note 'Drift Calculation for Thin Film Resistors':

https://www.vishay.com/docs/28809/driftcalculation.pdf

Note that the numbers in 2.4, ' Precision Operation Mode' don't seem to be right. Given from 2.1, drift @ 1000 hours, 125C = .25%.

Therfore at 85C it should be 0.25% / (2^((125-85)/30) = 0.1%, not < .05% as shown.
 
The following users thanked this post: 3roomlab, msliva

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #112 on: July 26, 2019, 03:30:12 pm »
When i make a 10V - 7V divider from a 8x 5KOhm array using six resistors (2 resistors as 2.5 KOhm and 4 resistors as 5 KOhm) it will take 14 mW, which gives about 3 °C temperature increase. So i can run the Nomca at 40 °C. If the equations from the vishay paper apply in this low power regime, then i get an estimate of 290 ppm / 12,70 = 23 ppm, where 12.7 = 2 ** (150 °C - 40 °C) / 30 °C.
A change of -25 ppm has been observed in the drypack, so a 23 ppm number appears plausible to me. During the same drypack tests the division ratio of a divider optimized for low TC exhibited a change of 0,3 ppm of the division ratio.
To me it appears that vishay engineers were more or less focussed on durability in high humidity/high temperature regime, where it is easier to see real effects. In the meantime i can determine TC differences inside a Nomca array with a sigma of 0,08 ppm/K per resistor just from ambient temperature changes (delta T = 27 .. 29 °C),  without using the oven and i am getting results consistent with the oven measurements i did some weeks ago. Still no significant humidity effect on the division ratio after 3 weeks in the drypack.

As others noted before: The Nomca 1603 array in the Fluke/Wavetek 7000 reference means others have done these more detailed tests before and the Nomca part was found to be useful. Of course, as of to today there may be other, even better choices of thin film resistors. For example the Vishay PRA datasheet looks interesting. On its first page they claim "Very low noise < -35 dB and voltage coefficient < 0.01 ppm/V".

Regards, Dieter
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 811
  • Country: de
Re: Statistical arrays
« Reply #113 on: July 26, 2019, 03:56:47 pm »
@Dieter: What do you mean by sigma? The standard deviation of your measurement? Or is it the error of the fit?
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 6308
  • Country: de
Re: Statistical arrays
« Reply #114 on: July 26, 2019, 04:01:47 pm »
The formula in the Vishay note on drift is only an approximation. It may or may not work very well for different types of resistors. Also the extension to much lower temperatures is speculative. Its more like tested from maybe 100 C - 250 C and increasingly uncertain at lower temperatures.

With the epoxy encapsulated parts, there is also another effect. The usual epoxies have a glass transition temperature somewhere in the 100 C - 160 C range. Things change when comparing the range well below (e.g. some 50 K) and above the glass transition temperature. At higher temperature there is some kind of equilibrium that can be reached, well below it is an increasingly slower drift towards a more dense state.
The rate of cool down from high temperature can have some effect - it kind of sets the disorder that is frozen in. Slower cooling from some 150 C (e.g. soldering) would work like a bun in. It's a window some 30-50 K below glass transition that is important. Burn in at higher temperature would cause new disorder, kind of resetting the clock.
Chances are the formula is for the higher temperature part, with the epoxy more or less in equilibrium.

The humidity is expected to mainly effect the epoxy, so there may not be much correlation with the high temperature drift.
 

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #115 on: July 26, 2019, 07:15:42 pm »
By sigma i meant standard deviation. The number given (0.08 ppm/K) is the standard deviation when comparing eight results of the oven measurements with the line fits of room temperature measurements. When i partition my recent measurements, i am getting similar standard deviations. There is no indication that the oven/box method results are different/worse than the ambient temperature results.
My uncertainties mean that the observed 0.3 ppm humidity effect of the divider isn't a real measurement but more like an upper limit.

Regards, Dieter
 
The following users thanked this post: e61_phil

Offline imo

  • Super Contributor
  • ***
  • Posts: 1894
  • Country: 00
Re: Statistical arrays
« Reply #116 on: August 10, 2019, 12:14:27 pm »
Here is a nice online app called "Resistor Network Finder":
http://kirr.homeunix.org/electronics/resistor-network-finder/

As the number of resistors in the network is limited to 6 and I want to use a bag of 10k resistors (+/-25ppm/C) only, I added several serial and parallel combinations of 10k to my list of "stocked resistors".

Interestingly, it looks it can find almost _any_ feasible value with such an approach..  :-+

My current understanding is the resistor's TCs within a lot (?) do follow somehow a "distribution" typical for the lot. Or do not?

« Last Edit: August 10, 2019, 01:03:13 pm by imo »
 
The following users thanked this post: doktor pyta, razvan784, Magnificent Bastard

Offline dietert1

  • Frequent Contributor
  • **
  • Posts: 281
  • Country: de
    • CADT Homepage
Re: Statistical arrays
« Reply #117 on: August 30, 2019, 09:05:06 pm »
Got some more measurements on Nomca arrays. First i repeated measurements on the arrays A and F that have been in drypacks for about two months now, with very stable results. The 6 % reduction in absolute TC results from an improved temperature measurement.
From nine more arrays tested now the parts J, L and O exhibit low TCs and low TC spreads.

Regards, Dieter
 