No, Mr. Pettis, it's not April 1st today. If you present a table with limits you will stick to, that's it. We know from TiNs measurements that your resistors are no better than econistors. My econistor humidity test was to confirm what others found before. Yes, the econistor remains in it's dry pack and i will measure it again after a week or so. I think Mr. Pettis is creating a lot of noise without any signal.
I was not measuring one Nomca part but six, a total of 48 resistors. Sooner or later i will also characterize the other 34 econistors we have, looking at their TCs. And i wrote about bridge measurements were i had one 13:1 Nomca divider outside and one inside the TEC box, both running from the same 10 V reference. That was the measurement which resulted in the +/- 0,5 ppm/K limit for a 18°C cycle. So it wasn't a DVM resistance measurement only, but very close to the indended application.
Concerning resistor noise: Is there a link/reference on how those measurements were done? I mean before someone talks us into using parts with 60 ppm humidity dependence in order to avoid some 0,07 ppm extra noise, i'd like to have a closer look. Maybe a Nomca array just needs to be protected from random ventilation. I mean its thermal mass is very small in comparison to its surface.
Regards, Dieter