.. Well, after 2 weeks of this torture, the readings in the 3458A suddenly dropped 0.25ppm (when it had been solid, with no discernible drift, for the first part of the year). ...
and the 3458A dropped another 0.25ppm! [The 3458A has never recovered-- even though the humidity has been around 8% for months now, it is still low by 0.5ppm-- so I think it was permanently affected]. ...
My Fluke 732B did not suffer at all through this (as far as I can tell-- I don't yet own a JJA)-- the important resistors in the 732B are all hermetically sealed.
Hi,
If you don't have an JJ, how did you distinguish, or decide, that the 3458A was drifting absolutely and not your 732B?
I assume you noticed the drift by comparing the 3458A against the 732B?
You know, what I mean, the "man with the 1 clock, 2 clocks, or 3 clocks" problem.
Anyhow, if it was really the 3458A, there are 2 possibilities of a permanent shift.
First, the sudden high humidity caused a sudden shift , i.e. an irreversible ageing, of one or two of the bulk metal foil resistors, about 25ppm in total.
Second, the humidity caused the temperature stabilization to fail, heating the LTZ even more than 95°C, where it is supposed to be normally.
Maybe you did not log that event.
This might have caused a reversible but permanent hysteresis, resulting in a smaller reference value, so the 3458A reading should go up afterwards.
Would be interesting what you have observed, and how you argue about the 3458A/732B comparison.
And I am interested, how you basically monitor the 3458As stability , against that single 732B only?
Do you leave the 3458A always on?
What if you switch it off, does it return to its former value?
Frank