OK, I want to make one thing VERY Clear: I -really- do appreciate all the work and generous sharing of information provided by DR Frank, Andreas, Digelent, and everyone else and this is a GREAT discussion, and a GREAT collection of very smart guys. It is an honor to post some data once in a while. My hats off to all of you because I know how much time and patience all of this takes. Why do I tell you to wait a few years and re-test Vref when its settled down? Because that's what we are dealing with here. What I am trying to provide to the discussion is some feedback from the real world, and trust me - I understand the Volt-Nut vibe - its fun and addicting, for sure! But I have to mix the desire for ulta-precision with pleasing customers (enough to pay their invoice) and to make a profit in a reasonable time frame: Truth be told, my main goal is to not get a phone call at 2am because some machine went down halfway around the world, so I have learned to really make the stuff last. Other designers scoff because I tend to over-design, but then again most of the stuff we made in 1985 is still running - except for a brain-dead EPROM here and there.
I'm also old enough and lucky enough to have the stars align to have met the likes of Jim Williams and Bob Pease (working on other projects, not Vrefs) and they really were absolutely brilliant, and great to work with.
I didn't know I was in a "fight" with anyone, that is the very last thing I want to do - I think what I failed to make clear is the difference between a carefully controlled lab at home or a very carefully temperature controlled metcal lab (I think this is Dr Franks expertise) vs. a production environment where you can't trust how abused these things will get, and how long does something last - and that is something I deal with on a daily basis.
As David Jones points out, if you are designing a Vref for a product, the "Ambient temperature" can be "a real trap for young players". And what you will find missing on almost all high end DMM / Vref datatsheets is the RATE of CHANGE of ambient temperature that is allowed, and this will bite you in the butt every time when you deliver a product if you're not ready for it, and no SPICE simulator will handle this at all. This is one of the places where THEORY ends and REALITY begins. For instance, the '3458 is rated for 55°C ambient, but you'll see in a moment why that is only part of the story.
Now: Here are some real world numbers I jotted down from a facility I was at yesterday. I tried to take pictures for you guys but it is not allowed.
The setup:
Test Prober station #6 has 4ea HP/Agilent 3458a's. All are on 90-day recal schedule, and they are staggered so that there is only one new unit running at a time. They range in age from 4yrs old to 14 yrs old. None have had their Vref replaced or modified, but two of the older units have new VFD front panel displays replaced. There is one box with a pretty dim display that will need a repair soon, but as a computer-controlled DMM it is running fine.
These are in a 19" rack, stacked near the middle. Overhead are some test prober relay switcher boards and a couple blade PC's (one is running the prober & DMM's, the other one is dedicated to data logging and sending data to the inventory-control system, because at this station every working device is serial numbered and sorted into quality bin lots). Under the DMM's are some power supplies, more relay boards and more power supplies. Because the prober is known to create a small amount of dust, it is surrounded by a plexiglas glazing safety shield doors and the enclosure is under slight negative pressure to keep the dust inside. The doors are only opened up when a stack of wafers are installed or removed. The whole thing is in a cabinet against a wall, on the other side of the wall is an air compressor and chiller and a bank of vacuum pumps. There is also a high-pressure Osmosis - De-ionized water processing station. In other words the wall is fairly warm to touch. To the right of the cabinet is another machine's motion control cabinet and power supply. It is very warm also.
The room itself is running about 27~29°C on a very warm day outside. You have to wear a Tyvek bunny suit, hood, mask, booties and gloves, and you wish they would have the temperature lower, but that's what we have on the menu today.
Inside the '3458a boxes, with the machine idle, bottom to top, the interior temps are 38, 41, 43 and 40°C.
Now here's the fun part: When the prober starts, and then when the machine on the right starts, and then when you notice more vacuum pumps are online on the other side of the wall... The interior DMM temps are now reading 43, 46, 50 and 44°C when I check again, about 15 minutes after everything starts up. Typically a 5 degree jump in about 15 minutes, or a rate of change of 20°C / hour.
NOW: What makes this work in this situation is the data being collected is 6 or 7 digits, and the software compensation / ACAL routines running inside the '3458 are able to compensate well enough so that the testing operation is still working right. AND the Vref Die temp is running at the higher temperature for stability during temp changes. The manager mentioned that they do get special support from Agilent in the form of customized firmware, and they do calibrations at more like 34 or 35°C.
So this is a different situation that you would encounter in a Volt-Nit or metrology lab situation. BUT as a designer, I have to plan ahead for major abuse.
Bottom Line:
YES, for a cooler, temperature controlled lab, running the LTZ1000(a) at a lower temp makes perfect sense. If your VRef is going into a very abusive environment where the temperature swings are sudden and relatively rapid - well then you really need to raise the Vref die temp to keep your Vref relatively stable, no matter what's going on around it. That's one of the reasons '3458 have the die temp set as they are, and I've never really seen one that has "degraded performance" if the DMM has to work hard for a living, at least not over the 90-day calibration period (recommend by Agilent when these units are used like this - nobody ever expects these to run over a year without recal in this setup). Even then you will need some compensation software to help smooth out the dips and glitches, and the '3458's do a pretty good job at that.
They have tried other DMM's on this rack, but the '3458's run the best while being abused. At least in this setup. I know there are other brands and I will not argue with you that they work well also in other situations.
Oh and by the way: This is something else Vishay leaves off the data sheets of their very expensive magical resistors: See how the tempco acts during high-rate-changes of temperature, and compare that to a wirewound - and then remind me why the Magical Vishays cost 10 times as much. Lets just say I have never seen a 10 times performance increase, but that's just me. I have never seen any sort of cost benefit ratio at all, at least for the stuff I design (like 20 / 24 bits DACs/ ADCS and high-precision current sources). The main point I keep streesing: After 10 or 20 years, these Vrefs tend to settle down and become very good performers even without any voodoo, magical datasheet gimmicks at all.
Your milage will vary, and to all Volt Nuts: I love reading about what you do and how you get there. Honest! The generous sharing of information here is fantastic!