Where did you find a schematic? That is better than the one I have. I have been using the 2236 schematic for the 2235.
There you are, from the Tek itself..
http://www.tek.com/manual/2235-service-manual
You are going to make a fundamental mistake ... I have 40 years of experience in electronic project and repair and I can guarantee it for you.
Your oscilloscope has crashed suddenly and completely, it is not a "wear" of a capacitor that caused this.
If there is something wrong with a capacitor, it is short circuited, not to low capacitance, nor too high ESR....
Without considering that the aluminum electrolytic capacitors are old enough in this model to be worn out, I disagree because:
Normally when pressing the power button the CRT makes that same, kind of static noise that old TV's does, for a second or two. Now my scope makes this noise but repeatedly for only a tenth of a second or so and the power led is also flashing in the same manner.
David knows of what he speaks.
The PSU in the 22xx series is well known for going into "tick mode" when the electrolytics dry out and go high ESR
C925 and C943 are common culprits but it can be pretty much any of them.
Replacing all the capacitors, systematically, without even to look for the origin of the defect, this is, as I wrote, a complete mistake.
It is very unlikely that it would be an electrolytic capacitor that make that kind of noise because once the thin Al²O3 insulation is pierced and destroyed, the space between the electrolyte and the aluminum becomes almost zero and it would be impossible that the power supply can restart to such a hig level that a led can be flashing.
My Marantz 2285B from 1978 that I use everyday has only one capacitor replaced.
And it still has the same performance as when it was new .... worn capacitors? They are 40 years old of good and loyal service and they will probably hold another 10 or 20 years .... It will always be time to think about recap when there will be repeated capacitor breakdowns.
I notice that we have a completely different conception of what is repairing an electronic equipment ... Your conception is rather what is called "reconditioning" and not repairing.
I do not like preventive electronic maintenance because I believe that in electronics, everything is measurable and one can find a defect in a component by measures, even sometimes simple measurements as ripple.
The systematic replacement of components is often a solution for those who do not understand what they are doing (not your case, of course) or beginners .... I am totally opposed to it.
In my experience, the more you want to do (and in particular replace components unnecessarily), the more you create problems.
But for a Tek 2235 oscilloscope that is old-fashioned, which is worth 100 bucks and is likely to be in the trashbin in a few years (nobody want an analog oscilloscope anymore), for me, what must be done is to look for the faulty component, replace it , and check the power supplies, ripple, correct functional operation, and make an approximate calibration and it is good for some more years of use.
I still have some success with vintage audio and I am applying my conception of repair without any problem....all my clients are very satisfied and I had no return of any problem after my repairs.
Voltage drop (mV) | ||||
# | fwd | rev | ||
CR960 | 50 | 50 | ||
CR962 | 50 | 50 | ||
CR961 | 0 | 0 | ||
CR963 | 0 | 0 | ||
CR956 | 337 | 407 | ||
CR957 | 337 | 403 | ||
CR954 | 439 | 2527 | ||
CR955 | 439 | 2526 | ||
CR879 | 492 | Ol | ||
CR907 | 493 | 999 | ||
CR970 | 99 | 157 | ||
CR967 | 99 | 157 |
This is the rate of failure of electrolytic capacitors....
If you replace a capacitor by a new one, you must be sure you are in the wear-out failure segment...And to be sure of that, you must have first several failures of the same brand of capacitor in the same equipment...
If you replace them too soon (during the random failure segments), you will have more failures than if you don't replace them.
So I finally removed the PSU cover and could have at least a visual inspection of all components. I can't see any signs of any leaking caps. But I can't be sure they aren't defective since I don't have the proper equipment.
I did however try and test the diodes around the T948 and it seems as if most of them are shorted. Could this be correct? If so how could that have happened? I tried two multimeters, one fluke 83 and a cheaper uni-t UT132C. They pretty much showed the same results but the uni-t one showed ".Ol" for values over 1.5 V.
If I replace the caps. What should I be looking for in terms of specs and what should I avoid. For instance should they be Japaneese named brands only? Low ESR?
Not having the aluminum electrolytic capacitors in the wear out portion of their reliability curve is hardly going to be a problem in a piece of test equipment which at a minimum is 28 years old. Input capacitors in linear power supplies last the longest because they end up massively derated but even they can fail at that point though corrosion of the lead connections as water or oxygen leak through the rubber seal; eventually they dry out also.
Look for low ESR, high frequency, and high ripple current capacitors intended for switching power supply applications. I like the Nichicon PW series for general use but there are many options.
1) Look at the Service manual of 2235: you have the list of replaceable electrical parts of A1 with all the specifications of the components....none specification of low ESR capacitors...!!!!
One of the rules in repairing is to respect the technical specifications of the manufacturer when you replace a component....Not better, not worse component, only the same as specified.
One of the rules in repairing is to respect the technical specifications of the manufacturer when you replace a component....Not better, not worse component, only the same as specified.
So there is nothing wrong with the diodes?
I am a bit curious as to what I would find if I tried to start the scope. Is there some easy way of discharging the caps after pulling the cord again? Or how long should I wait before it is safe to poke around?
Is nichicon the same as nippon chemi-con? That is the closest brand name I Could find at my local retailer.
So there is nothing wrong with the diodes?
I am a bit curious as to what I would find if I tried to start the scope. Is there some easy way of discharging the caps after pulling the cord again? Or how long should I wait before it is safe to poke around?
Tektronix *never* mentioned ESR or ripple current in their parts lists whether it was important or not. The only way to know is to recognize the manufacturer's part number.
I did just that, in my post below . Not sure what the problem is though. Do those components behavior mean anything to you?