Hi Were you aware of the two setup functions for the 1013D — Voltage Calibration and DC Shift Calibration — shown in my attached videos below

With the Voltage Calibration process, the stored values will be lost, so it's important to carry out the Voltage Calibration first, followed by the DC Offset/Shift Calibration.
Going forward, these settings can be stored within the 1013D oscilloscope. Using the Update section on the boot-up page, once you have saved the settings above, they can also be backed up to the microSD card inside the scope. This allows them to be restored later if new firmware is installed, meaning you may not need to repeat these two calibration procedures each time.
To save the calibration data:
1. Boot into the DSO 1013D Updater (bootloader mode).
2. Press Upload Calibration Data.
3. The calibration information will then be stored on the microSD card.
You can view the stored data by connecting the oscilloscope to your computer via USB and selecting the icon within the DSO 1013D Updater.
When needed, you can restore the saved data using the Download icon (located to the left of the Calibration Data option).
https://youtu.be/81gYnOh9r8s &
https://youtu.be/U086LOFFgn0