Author Topic: LeCroy Probus reverse engineering  (Read 31714 times)
Offline ollopaTopic starter
Re: LeCroy Probus reverse engineering
« Reply #25 on: March 09, 2019, 12:31:18 pm »
Offline Mechatrommer
  • Country: my
  • reassessing directives...
« Reply #26 on: March 09, 2019, 12:48:59 pm »
yes i saw that deck mount quick bnc type connector sold in ebay, the price is a bit high for a china brand. and its less neater than the original, the ground connector is "loose" type i'm not sure it will be impedance controlled. that will be the last option.
Offline wingel
  • Christer Weinigel
« Reply #27 on: March 11, 2019, 07:08:28 am »
> for non EU members, can you share the source?
sure, I just wanted to check with the supplier that i was allowed to post their product drawing in a public forum.
I found an edge mounted BNC connector from a Chinese company called "Eastron" (eastron.cn) on Alibaba:
https://www.alibaba.com/product-detail/RF-50ohms-ST-180-degree-edge_60634004992.html
When I contacted them it turns out that they also have a "push-on" variant of the connector without the coupling nut as in the attached picture.  Not too hard for them to do such a variant I guess, since all they have to do is to build a batch of the normal connector and not mount the coupling nut. And they are willing to sell just a handful of them since they still have a few in stock.
The Alibaba page shows a picture of a vector network analzyer (looks like a HP 8753E which goes up to 3GHz or 6GHz depending on options) so they should be able to check the impedance matching of the connectors they manufacture.  I'm hopeful that the connector will be of decent quality.
Please don't buy up all of their stocks though, I want to buy some first. :)
Offline wingel
  • Christer Weinigel
« Reply #28 on: March 11, 2019, 07:23:23 am »
(Updated.  Darn, I can't spell when I have just woken up and should be on my way to work.)
Finding a decent pin header turns out to be much harder.   I haven't found any supplier with any angled connector with 0.03" pins.
If one could just find a straight pin with the proper rounded end and surface finish (a couple of micrometers of gold on nickel) one could probably 3D-print or mill a plastic support to put the pins in.  And then build something like the attached picture with a bulk-head push-on-BNC to SMA adapter that one can find on eBay for a few dollars.
Build a small PCB like that with an I2C EEPROM on it and you'd have something that fits nicely in the LeCroy scope with no risk of inserting the pin header in the wrong direction.
Do you know any decent manufacturer that have an automatic lathe who could manufacture a run of a hunded pins or so and then plate them with nickel and gold?  It shouldn't be that hard to do, I just have no idea where to find a manufacturer who can do it for a decent price.
« Last Edit: March 11, 2019, 07:46:32 am by wingel »
  • Country: my
« Reply #29 on: March 11, 2019, 07:47:12 am »
dont worry, alibaba is a bit inconvenient for me (not sure how to make order (and paypal payment possible?) and look at seller rating) and i'm not in great need, otherwise we still have alternative the china nutted version above. if you want you can buy all and post in ebay for non-EU member as well. the alibaba link you provided has the outer rotating/lock ring, lecroy doesnt. but i'm thinking we can improvise that by removing the outer ring using brute force or something hard. i thought i've linked earlier the pcb side mount of a BNC connector with lock ring. i measure it about 15.5mm tip to pcb edge. i have this a few in stock.
One more thing.  Could someone who has a LeCroy probe measure the pin header and post the exact dimensions.  I'm trying to find a supplier who can manufacture the connectors just to see how much it would cost, and it would be nice to be able to give them some drawings of the part.
not entirely sure what you want, but measuring with my rusty caliper on AP020, the internal of protruding rectangular plastic protecting/guiding the pins is 6.7 W x 5.3 H x 5 D (mm) the pin diameter is about 0.76mm type, the mill-max also have 0.46mm diameter pins like what i bought from china no brand ebay here, which is too loose and not suitable. the linked pogo spring loaded in 1.07mm dia i dont think that will fit. i browsed mill-max catalog i cant find 2mm pitch headers with 0.76mm pin diameter, 0.76mm pin only available in 2.54mm pitch format, so maybe need to improvise again ::)
Online Someone
    • send complaints here
« Reply #30 on: March 11, 2019, 09:03:33 am »
You should be able to assemble a suitable pin array from mil-max 3150-2-00-15-00-00-08-0 or 3330-2-00-15-00-00-03-0. A PCB structure is robust enough to support and align an active interface as in here: https://www.eevblog.com/forum/projects/diy-100mhz-differential-probe/
The following users thanked this post: wingel
Offline wingel
  • Christer Weinigel
« Reply #31 on: March 11, 2019, 10:16:50 am »
Mill-max 3330-2 looks like it could work.  Thanks!
« Reply #32 on: March 13, 2019, 05:07:44 am »
Somewhat related thread on DIY probe construction for the older scopes, epoxy putty based ProBus connector, a couple more EEPROM dumps:  https://www.eevblog.com/forum/testgear/lecroy-pp092-2gssec-adapter/25/
Offline sixtimesseven
    • Flickr
« Reply #33 on: December 12, 2019, 05:34:18 pm »
Did they work out in the end?
Offline MilkmanCDN
« Reply #34 on: March 12, 2020, 01:26:40 pm »
Folks,
Many of the plastic tabs seem to be broken off on my older active probes.   I an effort to solve this problem, I took some calipers and created a 3D model.   .STL attached in the .ZIP file.
To test fit the model, I printed on a FDM (Prusa MK3S) using Prusament Galaxy Black PLA filament.   The results were much better than I expected.   The fit is wonderful and after mounting it to an existing probe chassis, it feels like brand new.   Tight fit.
Side-note: the back-side of the original connector appears to be covered in some kind of metallic paint.   I've found a few options, but haven't tested any of them yet.
I suspect printing on an SLA would produce better results, but alas, I don't have one (yet..).    If some one does, would it be possible for you to print and post pictures?
« Last Edit: March 12, 2020, 02:02:59 pm by MilkmanCDN »
The following users thanked this post: JohnG
Offline MilkmanCDN
« Reply #35 on: March 12, 2020, 01:31:59 pm »
Folks,
I believe someone was asking for 3D models of these two parts.    Pictures and Models attached.
If anyone finds a place to source these, I'm all ears.    I've been looking for quite some time but have always come up empty.
Offline dmderev
« Reply #36 on: October 11, 2020, 06:58:46 am »
Hello, could you please tell what is the marking on the chip located near R23 and R25 in the picture showing location of green jumper in your post. Most likely it should be a DAC? Which probe model this picture is from? Did you ever reverse engineered schematics?
« Reply #37 on: October 11, 2020, 08:30:40 am »
This is the probe AP022.  No, I did not ever try to reverse the schematics.
74HCT573C Octal D type latch
Q1 looks like a 5V regulator
But on the other side is your DAC: DAK703KU 16-bit monolithic DAC
« Last Edit: October 11, 2020, 08:32:17 am by ollopa »
Offline 2N3055
  • Country: hr
« Reply #38 on: October 11, 2020, 08:53:14 am »
HCT573 is transparent octal D latch, PCF8574 is I2C 8 bit I/O expander.  With those 3, you  can clock in 16 bits for D/A converter (DAC703 16 bit voltage mode D/A with built in reference) to set offset with the help of LT1112 dual opamp....
Dr. Richard W. Hamming
Offline dmderev
« Reply #39 on: October 19, 2020, 04:26:23 am »
For those interested, here is how control of the input mode and offset is performed:
( I programmed the EEPROM for AP020 and used a preprocessor that acknowledges all transactions. )
1. The quality of the contacts in the connector matters, poor contact leads to a transition to strange modes (a red bar “Processing ...” appears. When I realized this, the work became predictable. EEPROM address - 50h (all address inputs must be at zero)
2. The binary images on the Internet contain only the valid data. The default programmer leaves FF in an unused area. It confuses the firmware. All free memory must be filled with zeros.
3. The oscilloscope reads entire EEPROM sequentially, regardless of the size of the data block, in one pass. Then it recognizes the probe, and switches the input mode (in my case, DC coupling is turned on and 50 ohms is not turned on - perhaps the resistor in the probe, although in the picture you can see only the place for its installation.
4. When you rotate the offset knob, the oscilloscope sends the following commands to two PCF8574 port expanders - at addresses 0x23 (data port) and 0x21 (command port). First, the most significant byte of the offset is written to the data port (for positive numbers, the most significant bit = 1, for negative numbers = 0). Then bit 0x01 is set in the command port (latch new data in 74573), after that the bit is cleared (data remains in 74573) and then the low byte is written to the data port. Perhaps this leads to a small glitch at the DAC output, but they are OK with it.
5. If these transactions are not acknowledged, the oscilloscope makes 4-6 recording attempts and stops.
6. For example, 0x8000 = offset = 0, 0x7FFE = offset = -2mV, 0x8002 = offset = + 2mV. In this case, the DAC code is likely to scale, since for the sequence 1,2,3,4,5,6mV codes 0x8002,0x8004,0x8006, 0x8007, 0x8009, 0x800A. But I am not 100% positive in scaling.
7. When switching the vertical scale and if the offset is not zero, the oscilloscope sends the correspondingly changed offset - that is, on the 1mV range, the 10mV offset will be replaced by 50mV when switching to 5mV. If offset was zero, the transaction is not done as it stays zero.
8. Depending on the input mode, the command port bits have the following meaning - 0x10 - when set, the DC coupled input. 0x40 = when set, the input is AC coupled (on probes, is this?) When both bits are cleared, the input must be Grounded. That is, switching the mode and writing a new offset corresponds to 0x23 = MSB, 0x21 = 0x11, 0x21 = 0x10, 0x23 = LSB for DC coupled, 0x23 = MSB, 0x21 = 0x01, 0x21 = 0x00, 0x23 = LSB for GROUND, 0x23 = MSB, 0x21 = 0x41, 0x21 = 0x40, 0x23 = LSB for AC coupled. The relay may be switching Ground.
9. I did not find the purpose of the remaining bits in the control port, other buttons and knobs do not affect the samples.
The schematic which illustrates the concept of control in case of AP020,AP022. The DAC appears in complementary offset binary.
Other bit settings may appear needed for other probes. It looks that WavePro recognizes newer probes (ZDxxx, ZSxxx) and lists their names, but does not send offset data to probes on I2C. Newer scopes should be able to do that.
« Last Edit: October 19, 2020, 04:31:05 am by dmderev »
The following users thanked this post: macboy, lukier, tooki, gamalot, 2N3055, Sighound36, dgo_42
Offline MilkmanCDN
« Reply #40 on: January 29, 2022, 06:44:01 pm »
I've been tinkering away a little bit with the calipers as I believe someone asked for dimensions of the 6-pin probe connector.  I've attached a drawing which should give all the details one might need.
Also included is an updated .STEP (AP214) model to allow input into popular CAD programs.
Again, if someone knows where to source these, I'm all ears.
Offline dmderev
« Reply #41 on: May 25, 2022, 07:26:10 am »
For those interested, here is how I solved the problem of round pins on 6-pin 2mm connector.
There is a Mill-Max part number 3330-2-00-15-00-00-03-0 (Mouser part 575-3330200150000030) - and also Mill-Max 3330-1-00-15-00-00-03-0. These are bulk round pins, gold plated. I made a spacer from 3mm thick FR4, and plugged these pins into a right-angle PCB dual-row 2mm header.
I cut a piece of 3 positions from the 40 pin header. Since the pins are thicker than regular 0.5mm pins of such header, one need to push them in rather strong. Once the pins are in the connector, I applied epoxy and pushed in additional FR4 spacer. This creates a nice angle connector which reliably plugs into the probe socket.
When combined with the 3D-printed shell, this allows to build the probe PCB that fits. The picture shows two connectors: angle one to put on PCB, and a straight one which I used for connecting to logic analyzer.
On the side note: I printed the shell from PETG filament. Because the walls are thin, they may split. Applying acrylic solvent cement (dichloromethane or dichlorethane?) "welds" and bonds the shell layers and makes it strong enough to latch on the probe connector.
« Last Edit: May 25, 2022, 07:36:35 am by dmderev »
The following users thanked this post: Someone, tv84, tooki, 2N3055
Offline dmderev
« Reply #42 on: May 30, 2024, 02:30:05 am »
Does anyone know where I can get the mating (receptacle) 3x2 pin 2mm pitch connectors to build the extender for the probus? The standard 2mm connectors have too small holes and the probe pins do not go there freely. I found that a cutout from DVI connector (it has 1.94mm pitch) can be mated, but I am looking for a better solution... Or, alternatively, are there any receptacle pins that mate with Mill-Max round 0.8mm round pins?
Offline Momchilo
« Reply #43 on: May 30, 2024, 09:34:45 am »
They have a lot of different receptacles.
From their website: https://www.mill-max.com/products/match/mating-receptacles/3330
Offline bson
« Reply #44 on: May 30, 2024, 06:34:10 pm »
Are you sure type 04 is a serial number?  It looks like at least sometimes it's an ascii version number, for the newer probes.  Maybe for older probes it's a binary version number?
ZD1000:
04 06 41 30 30 31    'A001'
ZS1000:
04 06 31 34 38 32   '1482'
ZS2500:
04 06 41 30 30 35   'A005'
AP020:
04 06 96 18 00 99
Maybe 99 is 1999 and they changed the encoding at Y2K?  :P
The following users thanked this post: Mechatrommer
Offline bson
« Reply #45 on: May 30, 2024, 07:02:35 pm »
I think 38 is the input coupling and impedance, while while sometimes 39 is only the coupling.  Not sure why both would be there, especially since 39 is often merely the two first characters of 38.  Some probes only have 38.  It seems a bit helter-skelter.
DA18000AC.bin :
38 09 41 43 20 35 30 15 0a   'AC 50
39 05 41 43 0a   'AC
DA1820A.bin :
38 0b 45 78 74 65 72 6e 61 6c 0a   'External
39 05 45 78 0a   'Ex
OE695G.bin :
38 08 58 58 58 6e 6d 0a   'XXXnm
PP090.bin :
38 06 37 35 ea 0a   '75\xea
ZS1000.bin :
38 09 44 43 20 31 4d 15 0a   'DC 1M
39 05 44 43 0a   'DC
I wonder what those 0xea and 0x15 bytes are for, but I'm sure they serve some purpose.
I assume if it says DC or AC the coupling is locked.  Same if it has an impedance.
« Last Edit: May 30, 2024, 07:05:02 pm by bson »
« Reply #46 on: June 01, 2024, 05:13:23 am »
ZD1000:
ZS1000:
ZS2500:
AP020:
04 06 96 18 00 99
I'm not saying that it's an actual unique serial number, but it is in the "serial#" field of the probe info.
Also note that I don't have real dumps from actual probes for all of these, some are EEPROM values used for simulation so they may have SN values with some other meaning to the firmware developers.
« Last Edit: June 01, 2024, 05:38:18 am by ollopa »
« Reply #47 on: June 01, 2024, 08:09:22 am »
Look at the attached picture and you will see that there are two places in the UI where the coupling appears, one location is just 2 bytes and the other is the longer selection string.  This may be a clue as to the reason why, but I'm not sure.  The 0x39 tag doesn't seem to do anything to the UI but if I change the 0x38 tag in the ZS1000 EEPROM then it does change the input channel coupling selection between DC and AC coupling.  And if you look at the second screenshot I can change the text in the UI to something ridiculous like 5GOhm.
It could be that 0x38 is just the UI label and 0x39 controls the hardware setting, with 0x39 defaulting to DC 1M when not specified.
They do.  0xEA is the omega symbol in the extended ASCII table and 0x15 appears to be an alias.
« Reply #48 on: June 01, 2024, 08:13:48 am »
One more experiment.  Look what I did by changing 0x38 and removing 0x39.  I can set arbitrary text for the coupling UI
The following users thanked this post: bson, tooki
Offline heize0
  • Country: cn
« Reply #49 on: April 13, 2025, 02:39:05 am »
Thanks to everyone.
I successfully did a converter to use a Tek TCP202 current probe on my Waveusrfer 44MXS-B oscilloscope.
The following users thanked this post: bson, Martin72