Hi Bram,
having had some time to look further into the behavior related to your observations, we'd like to update a few points we addressed earlier.
Counter limitsRegarding the maximum measurable counter input frequencies, we were able to further optimize the counter behavior with hysteresis enabled. This will allow us to increase the specified frequency limits starting with the next firmware update.
Setting hysteresis to 0% already allows the counter to operate up to the higher frequencies.
Display behavior, averaging and measurement resultsWith a 200 MHz sine wave and approximately 100 mV rms applied to the oscilloscope, I again see something that shouldn't be there at a time base of 50 nsec.
See the two images below with different time base settings.
[...]
200MHz ~100mV RMS and 2nSec time base

.
Same signal and now 50nSec time base, and no there is no modulation on the signal from my HF generator.

And Andre77, wat about this picture witch a showd before, is it possible to correct this.
is it possible to adjust this behavior so that it matches the behavior of other scoops—that is, not to display amplitude modulation when it isn't present?
Thanks for bringing this up again, and especially for pointing out the Vrms measurements. The issue was already brought up and discussed
earlier, but at the time we classified it as a display-only issue related to interpolation-based rendering.
This is right at the transition point where the oscilloscope switches from showing raw samples to using interpolation (in this case Sinc interpolation). In this “edge case,” small features such as narrow spikes can appear unstable or seem to shift because they fall "between display pixels" and the interpolated points vary depending on the time offset.
The Vrms measurements showed that the effect can also have an impact on measurement results, something we had not taken into account in our earlier assessment and which we now want to correct.
In your case, changing the time scale of the main waveform window, using Zoom mode (here: >1x), or even switching the Measurement display mode to "Compact" and thus changing the utilized screen width should change how strongly the observed effect impacts both the displayed waveform and the resulting measurements.
That said, we agree that the waveform shown in your screenshot should not appear that way in the first place, and that the interpolation should not negatively influence the resulting measurements in this manner.
We have therefore addressed this by improving the interpolation behavior for these specific display and measurement scenarios, resolving the observed issue in the upcoming firmware update.
While the observed behavior may become visible when using Average mode in this particular setup, it is not caused by the Averaging mechanism. Averaging operates as a sampling mode, in your case at 1.6 GSa/s, and is internally processed at the maximum interpolated sample rate of up to 819.2 GSa/s, so it is not affected by the limitation discussed above.
Please feel free to bring up any further questions or observations you would like to discuss.
Best regards
Florian
Batronix