The automatic measurement is bad, this has been written several times before, that's why you have to use the cursor.
If I calculated and measured correctly, then at 1V/Div sensitivity, the maximum measurable voltage is 9.3Vpp measured with the cursor and according to the automatic measurement it is only 8.6Vpp.
In other words, the automatic measurement is faulty for some reason, I think this must be a firmware bug.
In 1V/Div.
The smallest change that the device measures is 0.05V.
In other words, 50mV, 100mV 150mV. There is no value in between.
Theoretically, it is 8-bit stuff, but if I divide 9.3V by 0.05v, it only comes out to 186, not 256.
If I compare it to the voltage accuracy of DMMs, which is ±(%+digit).
so if I measure a wave that is definitely 3.00V with it, but it shows 3.2V, it is only a +4 digit difference, inaccuracy.
According to the description: "DC Gain Accuracy: ±3%"
It doesn't say what % it is, but I don't think it is the same for every time base.
According to the description:
Sweep speed range: 10ns/div – 20s/div
Sample rate range: 1.5Sa/s – 280MSa/s
If you swap the sample rate value, for example, then
Sweep speed range: 10ns/div – 20s/div
Sample rate range: 280MSa/s - 1.5Sa/s
The memory only says:
"record length: Max 128Kbyte"
From this I assume that it is not constant.
Since we can't save the wave as a file with it, only save images, so I don't know.

I think the HD/Peak detect mode just thickens the lines and makes the part you can't see visible.
I tested it with a single trigger, zoomed in to the wave in HD and Peak detect mode, and I didn't see any difference in the faster time base.
But I hope I'm wrong.