Thanks Josh and flash.
Tried different addresses and adding a delay, no difference!! Also all instruments on the LAN are powered down, only the IM3536 is powered On.
Started looking at each individual IM3536 commands (think we found an error, see below) tried them from command line. Found that some don't respond and timeout with a <null> turn.
Here are some examples of timeout commands:
LEV:CHEC:LIM
DCR:LEV
DCR:LEV:VOLT
DCR:LEV:CVOLT
DCR:LIM
DCR:LIM:VOLT
DCR:LIM:CURR
These all timeout however the IM3536 is still connected and can return proper measurement values as shown in Current values tab. BTW these also timeout under USB connection, however USB seems to work with Setup, LAN doesn't!!
We can read and write various commands by command line and they work as expected, however once Setup is invoked the LAN interface to the IM3536 is apparently lost (even *IDN? times out) and doesn't recover after exiting Setup and a "Reconnect" is required.
Something in the Setup initiation is breaking the LAN connection to the IM3536.
Anyway, will keep digging into this as time permits altho somewhat hampered by not being Network, Interface and SCPI savie

Best