Author Topic: Program that can log/control many multimeters and other devices.  (Read 1787110 times)
0 Members and 21 Guests are viewing this topic.
Online KungFuJosh
  • TEAS is real.
Re: Program that can log/control many multimeters and other devices.
« Reply #7025 on: July 29, 2026, 12:49:22 pm »
I have some suggestions to improve the recently published LCR definitions by KungFuJosh for the Tonghui (+ other brands) and HIOKI series.
Thanks for the suggestions. I have plans to do more with these meters, including possibly monitoring the frequency. However, as HKJ mentioned, it's not necessary for the param sweeper. Both brands use FREQ? for frequency, so pretty simple to use.
I was considering adding it as 5th optional measurement for frequency tracking. That way it won't necessarily impact speed, but it's not a high priority as it isn't that helpful outside of sweeping.
Newer firmware versions on the TongHui models do allow arbitrary frequency input (to some degree - numberInt at least). TH/ST LCRs stuck on older firmware won't be as fun.
Hmm, this gives me an idea @HKJ will love. ;) What do you think about the ability to add a SCPI command to track manually? Something like:
#scpiTrack handle scpi #value
#scpiTrack ST2832 FREQ? #value M5_Freq Hz SI
Thanks,
Josh
Best Continuity Tester Ever
Best Continuity Tester Ever...Got Betterer!
The following users thanked this post: flash2b
Offline flash2b
  • Everything I like about myself is better with you.
« Reply #7026 on: July 29, 2026, 01:35:41 pm »
Well I've experienced the frequency point behavior myself on a Tonghui TH2830 (on a demo somewhere), but it might that newer units or newer firmware removed the annoying frequency points.Techmize does not state it anymore.
You cannot use Param Seeper on cmdSetuo numberInt controls by the way.
My Owon LCR2020 (lowest in the Range) has a resolution of 10 mHz across the programmable continuous spectrum
They say attention is a shovel. It's time to dig 'em out.
Offline HKJTopic starter
    • Tests
« Reply #7027 on: July 29, 2026, 01:47:45 pm »
Correct, but on steps you can put in linear sweeps on int, but is has to be a #interface control.
Online KungFuJosh
  • TEAS is real.
« Reply #7028 on: July 29, 2026, 01:57:30 pm »
Yup, that would have been an older firmware. I don't think they even advertised the change. But you can see in the 2830/2832 thread there was a point when the firmware changed the frequency input limitations.
That's good to know, thanks.
TongHui is weird. Even the TH2832 specifies 20 Hz to 200 kHz, with 15025 points. That's about 13 Hz resolution, which is odd. It appears that reality is also 10mHz resolution:
;; Found Sourcetronic ST2832 on USB Serial Device (COM10) (COM10) sn: Ver 1.0.0
FREQ?
;; 4.00020e+02
That might also be the result of the "newer" firmware.
Hioki specifies 5 digits with a minimum of 10mHz resolution.
So, yeah, numberInt would be wrong regardless. ;)
Thanks,
Josh
Offline flash2b
« Reply #7029 on: July 29, 2026, 03:17:48 pm »
Dear HKJ,
I have added support for the Anbai AT810A and Applent AT810A in my existing LCR definitions. 2 more devices for TC and 12 updated ones.
With #removeMatch it is even more easy to take not supported stuff out of a definition.  :-+
On top of this I also updated all other devices that this definition support and since I converted the TXT file to UTF-8-BOM I now use proper symbols like  Ω, Δ, Δ%, θ_°  and θ_r.
[ Attachment Invalid Or Does Not Exist ]
« Last Edit: August 12, 2026, 08:31:51 am by flash2b »
Offline flash2b
« Reply #7030 on: July 29, 2026, 04:51:21 pm »
The conversion from ANSI to UTF-8-BOM is very easy.
For Notepad++ just load your file, go to Encoding and change ANSI to UTF-8-BOM, and save the fie. I have no idea how other editors do this.
Windows Notepad can do it as well, Save As, Encoding, UTF-8 with BOM.
It enables the use of special symbols in TestController definitions, like Ω and °. #metaDebug save does not support it, but don't be alarmed, internally TC does do it properly. Also some logging does not support it, forgot which.
« Last Edit: July 29, 2026, 04:59:12 pm by flash2b »
    • Tests
« Reply #7031 on: July 29, 2026, 05:40:26 pm »
Dear HKJ,
Updated/added, thanks.
Offline flash2b
« Reply #7032 on: July 29, 2026, 06:41:14 pm »
Oh... I tried using getDeviceInfo(handle) I was sad to see getDeviceInfo(handle).version was not supported directly. I was hoping it returned a space separated string which I could attack with getElement, but it returns a struct instead which needs a var assignment to get it working. I did not want any var in this particular definition so I did not use it.
Online mawyatt
« Reply #7033 on: July 29, 2026, 11:57:20 pm »
If that is the case, then running in Debug mode will hopefully tell us why. On your Windows computer, that's easy to do. There's a file called testControllerDebug.bat in the TC directory that you can run that will automatically start debug mode.
Thanks,
Josh
Yes we are using the TC debug mode per HKJ suggestion. Will look into this more later today.
Best
Got some more time with this using TC Debug on PC. Here's 2 files showing the debug details of when using USB and LAN as the connection. This is the end of invoking Setup from the Command tab, then changing Parameter 4. Note that USB shows a number of lines with a IM3536: Script start while the LAN shows no IM3536: Script start and a few lines after such without the additional HEX data with Tx command.
Anyway, this seems to be why the LAN connection isn't allowing any IM3536 changes (Rx Timeout), while the USB does.
Here's the USB:
IM3536: Tx <MeasType4 Z>
IM3536: Tx <:PAR4 Z>
COM8: Tx: <:PAR4 Z.> 3A 50 41 52 34 20 5A 0A
IM3536: Tx <PAR4?>
COM8: Tx: <PAR4?.> 50 41 52 34 3F 0A
COM8: Rx: <Z> 5A
IM3536: Rx <Z>
IM3536: Tx <PAR4?>
COM8: Rx: <Z> 5A
IM3536: Rx <Z>
IM3536: Script end <>
IM3536: Tx <:PAR4?>
COM8: Tx: <:PAR4?.> 3A 50 41 52 34 3F 0A
COM8: Rx: <Z> 5A
IM3536: Rx <Z>
And the LAN:
IM3536: Tx <MeasType4 X>
IM3536: Tx <:PAR4?>
IM3536: Rx Timeout
Best
« Last Edit: July 30, 2026, 01:07:52 am by mawyatt »
~Wyatt Labs by Mike~
Online KungFuJosh
  • TEAS is real.
« Reply #7034 on: July 30, 2026, 01:29:42 am »
@mawyatt,
Try the attached instead. It seems like your ethernet connection might not be great, and it's timing out. I've increased the #readingDelay from 1 second to 5 seconds. If that's not enough, try increasing that line to 10 and see if that solves the connection issue.
Look for this line in the attached definition (if needed to edit):
#readingDelay 5
I'll make 5 seconds default in the next version if it works for you.
Thanks,
Josh
« Last Edit: July 30, 2026, 03:37:26 pm by KungFuJosh »
The following users thanked this post: mawyatt
Online mawyatt
« Reply #7035 on: July 30, 2026, 02:30:54 am »
@mawyatt,
#readingDelay 5
Thanks,
Josh
That doesn't work, neither does 10 seconds. It apparently has something to do with the Send command structure. See below where the USB command structure is totally different than the LAN.
USB mode issuing a Display Off command
IM3536: Tx <:DISP OFF>
COM8: Tx: <:DISP OFF.> 3A 44 49 53 50 20 4F 46 46 0A
IM3536: Tx <:DISP?>
COM8: Tx: <:DISP?.> 3A 44 49 53 50 3F 0A
COM8: Rx: <OFF> 4F 46 46
IM3536: Rx <OFF>
USB mode issuing a Display On command
IM3536: Tx <:DISP ON>
COM8: Tx: <:DISP ON.> 3A 44 49 53 50 20 4F 4E 0A
IM3536: Tx <:DISP?>
COM8: Rx: <ON> 4F 4E
IM3536: Rx <ON>
LAN mode issuing a Display Off command (Fails)
IM3536: Tx <:DISP?>
IM3536: Rx Timeout
LAN mode issuing a Display On command (Fails)
IM3536: Tx <:DISP?>
IM3536: Rx Timeout
Our limited understanding of how TC works with the IM3536 indicates that when Setup is invoked with LAN, the connection is lost with the IM3536. This explains the return of all 0s [ Debug shows <No data (Timeout?)> ] in the 4 Parameters values under Current Values  tab. However prior to invoking Setup, the communication is working as expected (proper Parameters values).
Best
« Last Edit: July 30, 2026, 02:41:53 am by mawyatt »
Online KungFuJosh
  • TEAS is real.
« Reply #7036 on: July 30, 2026, 02:43:26 am »
IM3536: Tx <:DISP?>
IM3536: Rx Timeout
IM3536: Tx <:DISP?>
IM3536: Rx Timeout
There's no difference in commands structure. What you're seeing is the way TC transmits/echoes data. The problem is, you're not getting anything over ethernet in that case.
I'd try removing the IM3536 from the device list completely (no duplicates or other connection types).
Try a different IP on the IM3536, and make sure there are no IP conflicts.
Then add the IM3536 back to the device list, and see if the connection issue persists.
Here's an example of my LCR with a good ethernet connection:
IM3570: Tx <getallModes?>
IM3570: Tx <none?>
IM3570: Rx <LP1,X2,Q3,Z4>
:DISP OFF
IM3570: Tx <:DISP OFF>
123.456.789.123: Tx: <:DISP OFF.> 3A 44 49 53 50 20 4F 46 46 0A
:DISP ON
IM3570: Tx <:DISP ON>
123.456.789.123: Tx: <:DISP ON.> 3A 44 49 53 50 20 4F 4E 0A
Thanks,
Josh
ETA: PS. You mentioned previously that you didn't have the Gateway set on your IM3536. That could be a problem. The Gateway should have your router's IP.
« Last Edit: July 30, 2026, 02:46:22 am by KungFuJosh »
Online mawyatt
« Reply #7037 on: July 30, 2026, 03:09:12 am »
IM3570: Tx <none?>
:DISP OFF
:DISP ON
Thanks,
Josh
TC can write and read to/from the IM3536 as the TC collects data for the 4 Parameters and displays results. So TC is seeing and communicating with the IM3536, it's only when Setup is invoked the connection is broken???
Best
Online KungFuJosh
  • TEAS is real.
« Reply #7038 on: July 30, 2026, 03:13:53 am »
When you load the setup, it sends ALL the read commands for everything visible in the setup popup. That's a lot of stuff for a bad connection to handle.
Your log from the setup popping up showed that it made it through a bunch of the command? stuff, then started timing out.
Changing the reading timeout to longer should have fixed that if it was just going too quick for the connection, but your connection seems to be inconsistent.
My guess is that the connection quality is bad, or there's some kind of conflict with communication. If you don't want to change the IP, then make sure the IP is reserved in your router for the LCR. You could also try restarting your router and the LCR. Maybe try a different cable with the LCR.
Thanks,
Josh
PS. Sometimes replacing the device in the device list in TC makes fixing things easier.
Offline flash2b
« Reply #7039 on: July 30, 2026, 10:21:50 am »
IM3536: Tx <:DISP?>
IM3536: Rx Timeout
IM3536: Tx <:DISP?>
IM3536: Rx Timeout
Try adding a :updatedelayed: 0.3
in this one
#cmdSetup buttonsOn Display Main
:read: :DISP?
:write: :DISP
:updatedelayed: 0.3
:string:
:nogrid:
:color: blue
Off OFF
On ON
:tip: Enable or disable display.
I have seen SCPI devices before that do not like quick read-write-read
If this improved that function, it can be added to other similar commands.
Do you have a load balancer between the LCR device and the PC/Mac running TC? The other cause can be a network configuration (load balancing, traffic shaping, fallback, firewall) leading to no answer/timeout.
« Last Edit: July 30, 2026, 11:35:29 am by flash2b »
Online mawyatt
« Reply #7040 on: July 30, 2026, 02:03:31 pm »
Thanks Josh and flash.
Tried different addresses and adding a delay, no difference!! Also all instruments on the LAN are powered down, only the IM3536 is powered On.
Started looking at each individual IM3536 commands (think we found an error, see below) tried them from command line. Found that some don't respond and timeout with a <null> turn.
Here are some examples of timeout commands:
LEV:CHEC:LIM
DCR:LEV
DCR:LEV:VOLT
DCR:LEV:CVOLT
DCR:LIM
DCR:LIM:VOLT
DCR:LIM:CURR
These all timeout however the IM3536 is still connected and can return proper measurement values as shown in Current values tab. BTW these also timeout under USB connection, however USB seems to work with Setup, LAN doesn't!!
We can read and write various commands by command line and they work as expected, however once Setup is invoked the LAN interface to the IM3536 is apparently lost (even *IDN? times out) and doesn't recover after exiting Setup and a "Reconnect" is required.
Something in the Setup initiation is breaking the LAN connection to the IM3536.
Anyway, will keep digging into this as time permits altho somewhat hampered by not being Network, Interface and SCPI savie :-\
Best
« Last Edit: July 30, 2026, 02:07:08 pm by mawyatt »
    • Tests
« Reply #7041 on: July 30, 2026, 02:15:44 pm »
LEV:CHEC:LIM
DCR:LEV
DCR:LEV:VOLT
DCR:LEV:CVOLT
DCR:LIM
DCR:LIM:VOLT
DCR:LIM:CURR
TC requires a ? after the command or it will not expect an answer.
The SCPIx driver has ways around that, with a couple of txrx commands to put in front of the actual command.
List of them are here: https://lygte-info.dk/project/TestControllerConfigDevice2%20UK.html#Commands_to_write_and_read_values
Online KungFuJosh
  • TEAS is real.
« Reply #7042 on: July 30, 2026, 03:23:02 pm »
LEV:CHEC:LIM
DCR:LEV
DCR:LEV:VOLT
DCR:LEV:CVOLT
DCR:LIM
DCR:LIM:VOLT
DCR:LIM:CURR
That is odd. Those are valid commands. Screenshot from here attached: https://www.hioki.com/e/products/lcr/application/IM3536/English/index.htm
I will email you.
Thanks,
Josh
Online mawyatt
« Reply #7043 on: July 30, 2026, 03:25:41 pm »
LEV:CHEC:LIM
DCR:LEV
DCR:LEV:VOLT
DCR:LEV:CVOLT
DCR:LIM
DCR:LIM:VOLT
DCR:LIM:CURR
We didn't include the "?" in the list, but was used to get a return value from the command line.
Best
Online mawyatt
« Reply #7044 on: July 30, 2026, 03:33:11 pm »
LEV:CHEC:LIM
DCR:LEV
DCR:LEV:VOLT
DCR:LEV:CVOLT
DCR:LIM
DCR:LIM:VOLT
DCR:LIM:CURR
I will email you.
Thanks,
Josh
Note these are DCR:LIM and DCR:LEV types that time out. The LEV and LIM commands work OK except the mentioned LEV:CHEC:LIM command.
Best
« Last Edit: July 30, 2026, 03:35:06 pm by mawyatt »
Online KungFuJosh
  • TEAS is real.
« Reply #7045 on: July 30, 2026, 03:36:00 pm »
Oops- I overlooked that! I'll review the DCR commands with the manual. Is there anything else that failed to respond on your IM3536?
ETA:
From your logs it looks like only these commands:
:DCR:LEV?
:DCR:LEV:VOLT?
:DCR:LEV:CVOLT?
:DCR:LEV:CCURR?
:DCR:LIM?
:DCR:LIM:VOLT?
:DCR:LIM:CURR?
:LEV:CHEC?
:LEV:CHEC:LIM?
I will review them, and modify the definition appropriately.
Thanks,
Josh
« Last Edit: July 30, 2026, 03:48:41 pm by KungFuJosh »
Online mawyatt
« Reply #7046 on: July 30, 2026, 04:06:05 pm »
SOLVED !!!!!!
Apparently the various DCR commands are not recognized by the IM3536. We deleted these commands from the Hioki 35XX file and now we have LAN connections and various functions working, and Setup behaves as expected :)
Still have some "cleaning up" to do, but things appear to be in order now and working with LAN ;D
Thanks for all the help :-+
Edit: Here's the updated IM35XX file.
Don't use this one, use the one Josh updated and added below, it's "cleaner"!!!
Best
« Last Edit: July 30, 2026, 05:04:46 pm by mawyatt »
Online KungFuJosh
  • TEAS is real.
« Reply #7047 on: July 30, 2026, 04:06:55 pm »
SOLVED !!!!!!
Best
I will release an official version that has the corrections made for device specific limitations.
Thanks,
Josh
« Last Edit: July 30, 2026, 04:14:09 pm by KungFuJosh »
Online KungFuJosh
  • TEAS is real.
« Reply #7048 on: July 30, 2026, 04:08:04 pm »
Here's an updated version of the Hioki IM series LCR meters. This should cover the compatibility issues Mike experienced.
Apparently, only the IM3570 had those extra features. 🤷
@mawyatt, please try this version instead.
ETA: Updated. I missed the other level check. Now this should be 100% compatible.
Thanks,
Josh
« Last Edit: July 30, 2026, 05:04:05 pm by KungFuJosh »
Online KungFuJosh
  • TEAS is real.
« Reply #7049 on: July 30, 2026, 05:07:04 pm »
@mawyatt & @Martin72,
Please test the updated file on my previous post. I replaced it as it still had a level check that needed to be removed for your devices.
Thanks,
Josh
