Author Topic: Siglent Bench DMMs SDM3065X, SDM3055 and SDM3045X  (Read 392382 times)
etc6849 and 1 Guest are viewing this topic.
Offline tautechTopic starter
Re: Siglent NEW Bench DMM SDM3055
« Reply #100 on: October 08, 2015, 07:02:05 pm »
Continuity test
Results:
Continuity triggering stable to 15 Hz. (0-4V pk squarewave 50% duty)
Minimum pulse width reliably triggered is 15 ms. (@ 1Hz rate)
I noticed that on my unit (with the 15R1 firmware) the continuity test was very slow. However, when I switched off the filter in VDC-mode the continuity mode reacts fast again.  So it seems the filter in VDC-mode influences the speed in other modes.  I assume that during your tests the filter was off.
I would prefer that in a next firmware version, the filter is automatically switched off when in continuity mode.
Correct.
Yes I've noticed some filer on/off dependencies too. You are quite right, why would any filter be needed in continiuty tests.  :-//
« Last Edit: October 08, 2015, 08:17:46 pm by tautech »
Avid Rabid Hobbyist
Offline sanman
« Reply #101 on: October 13, 2015, 12:24:43 am »
Page 96 of the SDM3055 User Manual talks about Application of the Analog Filter and specifies the introduced error.
Is the reading of 0.007mV within spec?
« Last Edit: October 13, 2015, 02:10:12 am by sanman »
Offline Eric-H
« Reply #102 on: October 15, 2015, 06:02:07 pm »
I think you are right, this describes the effect I'm seeing! So it is probably not a digital software filter that is used, but an analog filter in hardware that is activated.
I checked the manual for the other effect I mentioned (jump in measured temperature, see my earlier post). The manual states the dmm measures the cold junction temperature. Could it be that this is measured only once when entering temperature mode? So when you measure for a longer time and the cold junction temperature changes, the dmm doesn't compensate for it while being in the temperature mode. Only when you leave the mode and re-enter it, the cold junction compensation is done again.
Offline rf-loop
« Reply #103 on: October 16, 2015, 05:25:29 am »
Yes.
Most low VDC range is 200mV
Specifications: Accuracy: +/-  0.015% from reading + 0.004% from range.
If reading (input V) is zero.  It can still display +/- 8uV (0.008mV)
(also there is specs for temp and temp coefficient)
If range is 200mV DC and  input voltage is 100mV exactly it may display between 99.977 and 100.023 and it is as specified (if cal temp have been +23 and it is now in +18 to +28 temp and if it have been least half hour on.)
Also it need note that measurement need follow example these "thumb of rules" and something more  in this manual for Basic Fundamentals of measurements.
http://www.keithley.com/knowledgecenter/knowledgecenter_pdf/LowLevMsHandbk.pdf
Look example page 3-3 about avoiding Thermoelectric EMFs as error source  for low voltage measurements.
I drive a LEC (low el. consumption) BEV car. Smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.
Wises must compel the mad barbarians to stop their crimes against humanity. Where have the wises gone?
Siglent America
  • Guest
« Reply #104 on: October 16, 2015, 01:40:35 pm »
Hi Eric.
Your assumptions were indeed correct.
The SDM3055 does use an analog filter that engages when the unit is powered up. The filter can improve the measurement accuracy when measuring DC voltage which has some ripple on it. We calibrate the DMM when the filter is turned off so there is a small error when the filter is turned on. You can, of course, use the Relative function to null this small offset out of you measurement.
On your thermocouple question, we do use a TI IC to measure the ambient (cold junction) temperature and use this value to compensate the measured temperature value. The 3055 continues to look up the cold-junction value throughout the Temperature measurement. We are still looking into this problem to find the reason for the offset but it is proving to be difficult for us to reproduce in our lab. We will continue to work at it in order to solve this offset problem.
Offline Eric-H
« Reply #105 on: October 19, 2015, 04:59:57 pm »
If I can provide any information to help you to reproduce the problem please let me know.
I just measure the temperature for 10+ minutes, then switch to VDC mode and immdiately back to temperature mode. If I measure the temperature for a few hours, I've seen the temperarture jump for as much as 3 deg C. The ambient room temperarture may have changed an equal amount in that time (but I'm not sure of that).
« Last Edit: October 19, 2015, 06:31:11 pm by Eric-H »
Offline Eric-H
« Reply #106 on: October 20, 2015, 01:30:38 pm »
I've checked some values for the temperature jump effect. I recorded the temperature together with the voltage.
After powering up the unit, I get after 5 minutes a temperature value of 15.8 deg C (-5.8 uV) after switching to VDC and back to temperature I get a value of 16.4 deg C (-5.1 uV).
After another 10 minutes,  I get a temperature value of 14.4 deg C (-81.6 uV) after switching to VDC and back to temperature I get a value of 16.0 deg C (-78.4 uV).
So it seems the measured voltage is stable (and correct), but the corresponding temperature has an offset. During the measurement the room temperature was stable at about 16 deg C (checked with another thermometer).
I hope this might help in diagnosing the problem...
« Last Edit: October 21, 2015, 11:54:57 am by Eric-H »
Offline Muttley Snickers
  • Cursed: 679 times
« Reply #107 on: October 22, 2015, 07:41:58 am »
Our dear friend Martin Lorton now has a third video posted on the Siglent SDM-3055 with the new faster firmware.
I'm off to watch it..... :popcorn:
Offline ahnatiw
« Reply #108 on: October 27, 2015, 02:03:29 am »
Hi Everyone,
I have been comparing the SDM-3055 to my Agilent 34461A and noticed that the trend chart is not updating the display correctly when viewing ALL trend data.  Attached are some photos of both units.  I noticed this first with the R13 firmware. I also updated to the latest firmware version (15R1) and the bug is still present.  The firmware is getting more stable, but it looks like there are still a few more major bugs to be squashed.
Has anyone else observed this issue?
I can confirm that each individual sample agrees to within approximately 0.1 mV, it just appears that the routine updating the display is not working correctly.
Cheers,
Al
« Reply #109 on: October 27, 2015, 02:48:39 am »
Welcome to the forum.
Thanks for the feedback, I'll make sure Siglent is informed.
Avid Rabid Hobbyist
Offline ahnatiw
« Reply #110 on: October 28, 2015, 02:47:15 am »
Thanks tautech ... I have been lurking on the forums for years, but never got around to registering or posting stuff.
Following up on the trend chart issue, I also saved the captured data to a USB stick and verified that the data collected is correct, so the bug is isolated to the graph updating routine.
I did notice another issue, this time with the acquisition delay.  If you change it to manual mode and try to set the delay to 10 s by changing the unit from us to ms and then to S, it does not work - the value is not altered or applied (you can watch the trigger indicator at the top of the display).  However, if you stay on the us scale and just keep increasing the number to 1000 us, then 100 000 us, it automatically changes units and applies the delay correctly.  The next problem is that if you try to set the unit prefix back to say ms, it does not work.  The range is stuck in seconds and a reboot is necessary to reset the state.
Sorry if the description is a bit convoluted, but even taking photos does not really help.
Cheers,
Al
Offline Eric-H
« Reply #111 on: October 28, 2015, 08:49:32 am »
Yep, I see the same issue.
I checked using a function generator that outputs a ramp with a period of 15 minutes. On the trend chart I see, after a few hours, that the display "squeezes" the data. The oldest data (on the left) is stretched, while the newest data (on the right) is compressed.
« Reply #112 on: October 28, 2015, 09:28:50 am »
Thanks guys  :-+ keep the info coming.
Avid Rabid Hobbyist
Siglent America
  • Guest
« Reply #113 on: October 28, 2015, 01:51:51 pm »
Hi Everyone,
Cheers,
Al
Hello Al.
Thank you for reporting this.
Engineering is aware of the issue and i am told they are testing for a solution now. If I hear anything soon I will let you know. Otherwise, you might check our FW versions from time to time.
Steve
Siglent America
  • Guest
« Reply #114 on: October 28, 2015, 01:55:00 pm »
Cheers,
Al
I'll report this as well.
Thanks
  • Cursed: 679 times
« Reply #115 on: November 05, 2015, 10:21:14 am »
For those that haven't seen it yet Martin Lorton has a follow up video in regards to the SDM-3055 firmware versions P15R1 and P15R2.
Offline Eric-H
« Reply #116 on: November 05, 2015, 06:41:53 pm »
First a big thanks to Siglent as they fixed my reported problem were the measuring temperature jumped when switching between DCV and temperature. In a new release 15 R2 this problem is fixed  :)
It did not change another effect I see during temperature measurement. See attached screen shot. The measurered value jumpes in 0.5 deg steps up and down before settling on the new value. I suspect it might not be a problem but the effect of the cold junction compensation which might be done in 0.5 deg steps. It is within spec but it looks strange.
What do you think?
Offline MartinX
« Reply #117 on: November 05, 2015, 09:34:41 pm »
Hmmm, as I interpret Martin Lortons video when the filter is activated a capacitor is connected to the input terminals? Thus loading the measured voltage with a substantial capacitance? That sounds really crusty, I hope I got this wrong.
Offline Elrod
« Reply #118 on: November 12, 2015, 02:54:03 pm »
Just received my Siglent SDM3055 yesterday.  I immediately updated the firmware to the latest (1.01.01.15R1).  I left the unit in continuity mode when I shut it down last night.  This morning when I turned it on it went to beeping like crazy with random low resistance readings on the screen.  I assumed the leads just happened to be touching but they were not.  I unplugged the leads and it still kept beeping.   I shut it off and then back on.  It came up in DC volts and everything seemed normal.  Tried to duplicate and it didn't have the issue.   Tried it again and sure enough the problem is back.  As I type it is going crazy.  I will record a short video.  Has anyone else seen this?  Can anyone else check there unit for this behaviour?
Elrod
« Reply #119 on: November 13, 2015, 02:18:03 am »
Elrod
Most unusual.  :o
Can you confirm FW 15R1 is shown in the UI as the installed FW?
I'll also point Siglent to your post.
Avid Rabid Hobbyist
Offline Eric-H
« Reply #120 on: November 13, 2015, 07:49:17 am »
Elrod
I checked my meter and saw the same behaviour. I used firmware 15R2, but also saw it with older versions.
« Reply #121 on: November 13, 2015, 08:33:46 am »
Elrod
Siglent informs me this may be another bug and they are fixing it ATM.
Eric, Where did you get 15R2, I could not find a download link for it?
« Last Edit: November 13, 2015, 08:49:42 am by tautech »
Avid Rabid Hobbyist
Offline Eric-H
« Reply #122 on: November 13, 2015, 10:14:48 am »
I got it directly from Siglent as a pre release test for the fix they made in measuring the tempearure. They said it will be available from their site...
Offline Elrod
« Reply #123 on: November 13, 2015, 01:06:41 pm »
Tautech I can confirm my version is 15R1.
Offline Elrod
« Reply #124 on: November 29, 2015, 04:11:35 am »
Anyone have any news on when an updated firmware might be released for the SDM3055?  The last update was September I believe.