Hi,
As Josh already pointed out, I wouldn't worry too much about it.
These overshoots are almost impossible to avoid and vary in severity depending on the probe.
For example, the probes I have here:
The Testec TT-HF612RA and LeCroy PP022—both excellent 500 MHz probes that aren’t exactly cheap—have the same impedance values (10 M and 10 pF). And yet, they behave differently when connected to the same input on my SDS3104X HD.
You can reduce the bandwidth to 20 MHz; this doesn't change the overshoot, which is also clearly visible when you use an external calibration signal—which, at less than 5 ns, is lightning-fast compared to the “sluggish” internal one—instead of the internal one. But it doesn't make any difference.
Of course, you could try out dozens of other probes until you find one that works best with the scope's inputs—but I think that's a waste of time, because there's something else to consider:
But I think that overshoot means high frecuencies will be amplified slightly
Probably not; besides, the probe itself throws a wrench in the works with its derating at higher frequencies—and it starts early on.

Martin