MEGGER BMM80/F Insulation DMM RepairGot lucky, after only a couple hours it's all done, working again and buttoned back up !

First, got a good news in the first few seconds : the battery (was thinking Lithium something) is not dead, so no worries finding and buying a replacement..... because there is no (rechargeable) battery in there !
Just a 9V connector and a montage of x6 1.5V LR6 non-rechargeable batteries.
-group-therapy-thread/?action=dlattach;attach=1823395;image)
Corrosion : zero ! Not a drop... the case is quite well sealed.
Problem was simply the battery connector. It clearly had been worked on past. The negative wire was originally crimped, but it must have broken and the owner tried to fix that by soldering the wire onto the lug instead. Wire broke off of the joint.
-group-therapy-thread/?action=dlattach;attach=1823401;image)
Tried to solder it back, no joy : the white plastic of the connector is soooo soft that it melts like butter and falls apart, long before you have time to do any soldering action....
So I replaced that connector altogether, since I have a stock of these.
Meter now powers up !

I did a quick sanity check on all the "normal" ranges of the meter, works fine.
Then I tried the insulation testing feature... not knowing how it works.
It can output 50/100/250 or 500V. 500V is a bit much, but up to 250V is safe I though, to scope with a x10 probe, so I did that.
Was hoping the meter would therefore display 10Meg measurement..... But alas no.
The LCD prints an analog scale, but no "needle", no numerical value to be seen.
-group-therapy-thread/?action=dlattach;attach=1823419;image)
When I press the "TEST" button, it sounds and looks good : I hear a relay, then a loud beep, and on the scope scree I see a nice DC pulse of roughly the appropriate / expected voltage, which lasts 60= ms or so.
-group-therapy-thread/?action=dlattach;attach=1823413;image)
However once the HV pulse is gone, the meter does not display any result on the screen. No needle moving, no numerical value displayed.... nothing.
So I downloaded the user manual and found the problem : one is supposed not just to briefly press the TEST button, but to actually HOLD IT DOWN. The test runs for as long as the user keeps the button pressed.
-group-therapy-thread/?action=dlattach;attach=1823425;image)
Doing so, I can now get a measurement displayed on the screen ,and it reads 10Meg as hoped ! It works !

-group-therapy-thread/?action=dlattach;attach=1823431;image)
My first time with an insulation tester.... it's cool, I want one now

Before calling it a day, I spent some time trying to fix the positive 4mm socket, that the juice comes out of. As you can see it's in very bad shape to say the least. The "shell" of the socket , which is supposed to be a part of the casing, had broekn clean off of said casing, so bye bye socket, meter unusable. Guy apparently tried to secure it back to the casing using a mix of folder paper and some glue. Glue has come unstuck and only the paper was keeping the socket very loosely in place, somehow.
-group-therapy-thread/?action=dlattach;attach=1823443;image)
So I removed all that crap, scrapped the socket and casing clean, and glued the socket back in place with a ton of hot snot. Sounds nasty but actually the result looks much better than the previous repair / butchery, and it actually hold the socket firmly in place. It's actually a tremendous improvement both functionally and cosmetically... I am quite happy.
Also, the thick red HT lead that goes to that socket was at the verge of parting with the socket. It was hanging by only a couple strands. I stripped it and soldered it properly. Much safer and reliable now.
The owner will get not a working meter, but a better put together and more reliable one now. All that for free, as per usual

Hell I will be daring and see if I can extract 30 Euros from the guy, let's give it a try.
Successful repair.