A quick HP 5327B update, I found the A18 500MHz prescaler boards were defective in both of these.
There are some bias voltages for the U2 & U3 custom amplifier & schmitt trigger ICs, pin 3 of U2 should be 0.65V, pin 3 of U3 should be 0.9V (tolerance of ±0.05V for both).
-group-therapy-thread/?action=dlattach;attach=1757216)
The first 5327B I bought, the complete one, both ICs are dead, U2 bias is stuck at 1.8V, can adjust to 1.83V, U3 bias was 1.72V, can only adjust between 1.55V & 2.2V. The pot for the zero adjust is broken too, adjuster is snapped.
-group-therapy-thread/?action=dlattach;attach=1757210)
The 5327B that was missing the DVM control board (replica made a while back) & both top/bottom covers, is slightly better, U2 bias is working & can be adjusted in the range required, U3 bias is again faulty, can be adjusted between 1.32V & 2.6V.
-group-therapy-thread/?action=dlattach;attach=1757204)
Today I tried changing the transistor arrays, just in case they were defective and causing the odd readings, Mr Peak DCA said they were all fine, I fitted sockets as they were a pain to remove from the multilayer board, had to use bare pin sockets due to limited width between other parts. Also tried some CA3046 from a scrap Tek 24xx board, they made no difference, I didn't expect them too.
I've ordered the only 1826-0085 I can find in the world, will update if/when it turns up, hopefully can get one board working, the 1826-0084 is complete unobtainium.
Interesting there were two previous versions of the A18 prescaler board, the older ones look to have a hybrid of similar size to that used in the 5340A, no idea if it's the same part,I haven't checked.
-group-therapy-thread/?action=dlattach;attach=1757228)
-group-therapy-thread/?action=dlattach;attach=1757222)
David