The output buffer - after a series of experiments I settled with the V5 - see below. I have not seen a difference using the ADA4523 with compensation there thus I continue with the OPA189.
Below also the latest aging report.
The measurements are already challenging the capabilities of my 34401A pretty much. Without the TC compensation and having the DMM 24x7 on for at least 2-3 days after a week powered off the results will be much hairy, imho.
What I've learned in the last 10 months:
1. the TC of my meter is aprox +0.5ppm/C in 38-46C internal DMM temperature (I have the LM35 sensor glued over the ADC)
2. the humidity (?) after the meter is powered off affects the values +2..+3ppm and slowly settles down after 2-10 days after the meter is powered on
3. there is a random walk which is typically <1.5ppm peak-peak within an overnight measurement, the running STD (over last 100 samples, 100NPLC) shows typically 0.7-1.2uV (that includes temperature changes), in quiet periods 0.6-0.8uV
4. the ADR1001#1's voltage dropped down by some 2-2.5ppm in last 10 months, and its TC looks like -0.14ppm/C
5. the difference between a battery vs an ac_mains powering the Vref is not visible in my case (here the reference pcb sandwich is isolated by ~1.5cm thick foam inside the box, and its internal temperature measured by an LM35 typically rises up by 1.2C when powered off the transformer placed in the same plastic box with air vent slots)
6. I messed with the R1 values from 1k to 47k. I saw none difference in the measurement noisiness except a small drop of the output voltage, perhaps 1-1.5ppm with the 47k.
Also, it would be great if our better equipped contributors here would be kind enough to analyze in detail the performance of the AZ opamp buffer as depicted below, for the stability and noise (and will suggest improvements).. In the next Vref I would put an 1k into Q2 base, and not sure whether to filter out the AZ switching at the opamp's output..).