Hi Dr. Frank.
..
On the other hand, there was a design/implementation flaw in 3458A by using that big customize chip which contains many paring resistor with very strict requirements that often fail to meet.
Your 3458A is good today, my 3458A is good today, doesn't mean they will be good always, doesn't mean other 3458As are all good. My 3458A was bought 10 years ago 1st hand from an authorized Agilent dealer, but about two months after, I found out it was drifting all the time, they provided free repair of course - they replaced the big customize chip. I myself once try to fixed a drifting 3458A that ended up with suspecting that chip and can do nothing about. I Think TiN has the similar feeling about it.
Even with 'good' 3458As, they require frequent ACAL to cancel the drift of resistors within the big chip(better hourly), which is not possible for T.C. measurement that often automatic, continuous for several hours.
.
Hello zlymex,
I totally disagree, that this property is a design flaw., that's not correct, if you read about the design goals (see hpj 4-1989) of this instrument.
That is, that it should be a very fast A/D converter, also, or in first instance.
This demands, that the integrating resistors would have to be split in different weights (multiple of 2, in this case).
Correct, the disadvantage is, that now the stability depends on the ratio of these integrating resistors.
And I pray, that my, and also your ASIC will stay hermetically tight, so that there will be no degradation, like TiN encountered in several of his A3 PCBs.
Anyhow, my instrument, and also your instrument, are good enough to make stability measurements on the order of a few tenth ppm.
Still, it would be nice, if you would provide these statistical characterization of the 4910A and 732B, so that we volt-nuts in this thread have a baseline for our DIY references.
Frank