still got to figure out what connector to use for the temperature sensors of each reference...
Neat build, just few points. BNCs might not be that good if you chance every last microvolt due to thermal EMFs and usually brass outer shell.
Also I'd trim pins of LTZ much shorter, almost flush to the board surface. Having them that long might cause more harm than good, as even with foam you can't avoid air drafts 100%.
I guess the buffer is a LT1010 type?
now it's off to the lab corner for burn-in next to the LM399 reference![]()
That's the way we catch spiders in our house (the glass) - it looks rather appropriate!
Around 2011-2013 I spotted a table of voltage references listed by Paul Rako in a magazine enumerating all known good references ... - ADI, National and LTC were listed.
A lot of ADI stuff was listed but the LTZ1000(A) from LTC was missing as high end reference ... well I wrote the man and his answer was he was only a webmaster for a certain semiconductor company - nothing more, although we knew the man from many technical webinars ... - so his selective memory was at that time not aware of the LTZ1000(A)
Times have changed since a certain semiconductor company has acquired LTC ... ?
Flinstone
So now this is 3rd PCB on which same chip was installed, and it's safe to confirm that jumps are coming from the flakey LTZ itself, not the related circuitry around it. As side note, this data shows that properly working LTZ1000-based refs and behaving 3458A can achieve ~0.8ppm stability in mid-term, even without battery power or controlled temperature setup. Ambient temperature on 7 day graph log vary from +26C to +23C.
I'm about to join the most prestigious clubJust waiting on getting the z202 resistors from Vishay and I should have everything! I would like to practice my soldering skills before I have at it, though.
I recognize those pcb :-)
Thank you so much again for sending me your extras.
@calmachine - please show what you did and why you did it that way.
Whole bank powered by Keithley 2400 at +11VDC.
Hello,
I don´t know where I have seen the circuit from Lars.
It was a LT5400 with 2*100K in parallel and 2*10K in series.
So actually a 7:5 divider.
This together with a TL431 based heater could give a good basis for a 7->10V transfer.
See bottom of page:
http://www.techlib.com/electronics/ovenckts.htm
with best regards
Andreas
Hello Branadic - can you explain me how you obtain with 9k||9k + (1k+1k) (= 2.25) a ratio of 13:9 ... ? Perhaps I am overlooking something ... !
Thanks
Butterfly
Be careful using diffused chip resistor paks for precision voltage boosters - they are quite noisy compared to PWW. If you're going from 7V to ~10V that noise will get gained up also. Being SMT devices they are prone to mechanical stress pickup from the board, be warned.
Also note that if you use a '5400, you won't get a tight TCR ratio match across -two- or more packages. You must do all your tight-tolerance divider ratios with just the 4 resistors provided in -one- single package in order to keep a tight TCR.
Generally there are much better devices for use around LTZ circuits: For instance talk to Edwin - he will make you some PWW for the exact values you need at no extra cost.
For the LTZ1000 that often is 7.1-7.2v the LT5400-3 is suited.
The LT5400-8 is more suited for the LM399 that is specified with a typical 6.95v.
I have a couple of LM399 and LT5400-8 exactly for this purpose but I haven't done tests on them yet. Maybe because I have found a simple temperature compensated AD587JQ normally is as good or better with much less current draw for a DIY 10V reference.