Author Topic: Hantek - Tekway - DSO hack - get 200MHz bw for free  (Read 2788129 times)
0 Members and 229 Guests are viewing this topic.
Offline bbf
Re: Hantek - Tekway - DSO hack - get 200MHz bw for free
« Reply #1025 on: January 27, 2012, 09:48:57 am »
Hello,
I have a problem with my Tekway DST1062B - the scope hangs with white screen after running for about 30 seconds to 30 minutes. I have tried upgrading the factory firmware to version 2.06.3 (111202.0) - no change. I have not tried to hack the scope yet.
Hardware version shown is 10050x555583e8, serial number T 1G/005 003613.
Do you have any advice? Thank you!
Ivo
Offline aghp
« Reply #1026 on: January 27, 2012, 10:44:28 am »
Hello,
Ivo
HW fail can not repair with upgrading FW!
Nearly sure it is main board fail. I have seen this situation "more than one time".
Take contact seller or if seller do not care then to Hantek service for repair or spare part main board. (It must be free or nearly free to you if warranty is still valid)
EU: Owon oscilloscopes and in Finland also  some Siglent selected equipments. All with our  lab deep Q.C. in Finland.  Surplus test equipments. Repair service. 40 year experience about electric and electronic repair, service and design. Local Owon SDS-Series  repair service for our customers.
Offline tinheadTopic starter
« Reply #1027 on: January 29, 2012, 01:57:32 am »
Backup tool for Hantek Handhelds : DSO1062B/BV/S, DSO1102B/BV/S, DSO1202B/BV/S
- download the attached file and unzip the dso1kb_backup_tool.up (MD5 41873654111d34ef71c6f7ecc10a9951)
- copy the dso1kb_backup_tool.up to flash drive, ensure nothing else on that drive
- insert flash drive into Handheld usb port
- run firmware update (Utility, Page 1/5, Uptdate and F4 to confirm)
Handheld will reboot automatically after "firmware update" is ready (in this case backup of firmware).
Now you can remove flash drive and save the content in a secure location.
« Last Edit: January 30, 2012, 10:04:51 am by tinhead »
Offline pgup62
    • Elec3i Instrumentation et automatisme
« Reply #1028 on: January 31, 2012, 11:25:04 pm »
Hello,
    Ivo
This white screen can come from 3.3V regulator that should perform a voltage between 3.2~3.4. If this voltage is lower than 3.2V,then it may produce this white screen.
Elec3i has met this kind of behaviour already and has changed the power unit in some DSO to avoid this, It is why we test our devices one by one with at least 72 hours burn-in period... ;)
« Last Edit: January 31, 2012, 11:27:35 pm by pgup62 »
Offline bbf
« Reply #1029 on: February 06, 2012, 05:28:54 pm »
Hello pgup62,
thank you very much! It seems that some tuning of the 3.3 V regulator circuit helped. (I hope :) )
Offline pgup62
« Reply #1030 on: February 07, 2012, 11:44:19 am »
Yes, sure it will  8)
For now in Elec3i, we are promoting hand-held DSO1102B/DSO1062B and bench scopes DSO5102B (hw: 10070 0x555583e9) at 489.00 EUR net price !!! for some time. :)
« Reply #1031 on: February 07, 2012, 04:48:10 pm »
Pascal,
as already asked here :
https://www.eevblog.com/forum/product-reviews-photos-and-discussion/hantek-dso1202b-or-owon-sds7102/msg87209/#msg87209
- are these Handhelds with or without LAN?
- are these Handhelds true CATIII certified (KTK-1, KTK10 fuses + protection or simple 250V fuses?)
« Last Edit: February 09, 2012, 01:32:39 pm by tinhead »
Offline stef_teta
« Reply #1032 on: February 08, 2012, 08:40:32 pm »
I recently bought a DST1102B for my hackerspace and few ppl have used recently. After their use (sampling/recording/etc) the oscilloscope is freezing, randomly. Really randomly. On my own opinion, I think there is some broken samples still remaining on the file system, cause when I do a "df" on the debug console, I got :
Filesystem      1K-blocks      Used      Available       Use        Mount
/dev/root        63152          63152          0             100%        /
I would like to know :
- Where are recorded the samples before recording them on an usb key ?
- Is it normal to have the root filesystem used at 100% ?
- Have you ever met this problem before ?
Thanks a lot, waiting for your answer :)
... Later, I will put the hack on the track. Later :)
« Reply #1033 on: February 09, 2012, 01:06:41 am »
heh, you should probably kill some ppl for that
- tmp fs
- no, it should be never like that, typical used space is about 40-50%
- no never ever
You should check root fs for strane files, look in /param/ (and subdirs), /mnt (and subdirs if any) and / itself,
if you don't know if specific files are necessary or not just ask here, i will answer then.
Offline niebieski20
« Reply #1034 on: February 13, 2012, 08:47:35 am »
Hi
I just receive my DSO5102B
First I change it into DSO5202B and everything works great. Thanks for great job with reverese engineering.
Now I have a HW 10070 of PCB and there is a space on it for RJ45 for LAN and I wonder if it is possible to just add this connector and enable this option in firmware???
Offline paul
« Reply #1035 on: February 14, 2012, 03:28:19 pm »
Unfortunately I had the same problem as djeans, my scope locked up at the weekend. Totally unresponsive, powers up ok but just freezes after the hantek logo with the normal scope display, both traces on, run button lit up etc.
I tried the reset tool over USB but that did not work.
I am in the process of making a lead up for the uart, but which files do I delete and can they be backed up first ?
From a earlier post
remove the /param/sav/run* and /param/sav/run*
arent these two the same ? Or am I missing something.
Paul.
« Reply #1036 on: February 14, 2012, 03:49:51 pm »
it seems to be issue with the latest models (what is your version in system info->hw? 555583e8 or 83e9 ?)
yeah, it works as long scope is responding to anything.
If (due broken setting file - the one in /param/sav ...) the dso.exe is unable to startup properly the USB reset tools is not working.
From a earlier post
that was typo, it should be
- remove the /param/sav/run* and /param/sav/udisk/run*
Offline pgup62
« Reply #1037 on: February 14, 2012, 06:12:03 pm »
No, this behaviour can also be found on oldest versions (0x10050 555583e8) !
« Reply #1038 on: February 14, 2012, 06:23:23 pm »
sure, even the hw0 will sometimes write worng values into setting file, especialy when you power off
DSO while it was wrinting this file.
But now it seems to be just too much, i know already 6 ppl having exact same issue and  all 6 have latest hw version
with 83e9 design - within few days twice to number of issues from last 2 yrs ...
What additional unfortunate is the fact that the "default setup" button is not doing anything, before a
DSO with broken setup was always booting and it was enough to push once default setup button .. now ppl
have to disasemble scopes and delete manualy the file ...
Offline pgup62
« Reply #1039 on: February 14, 2012, 06:40:44 pm »
Well a bit tricky: hw or sw or both ?
last 25 devices (0x10070 555583e9) show normal behaviour after booting-up and 72 hours continuous operation. we have 2 old versions (10050 555583e8) that have this after boot-up freeze behaviour, one just after HMI appears (menus and waveforms), another within 20seconds-1hour after boot.
It seems that this can be a hw attitude (power voltage fall for instance from FPGA  or CPU sides).
« Last Edit: February 14, 2012, 06:43:02 pm by pgup62 »
Offline paul
« Reply #1040 on: February 15, 2012, 01:37:39 pm »
I have had the scope for a few months, and it has locked up a few times but always reset itself when I switched it off and on again. It was quite bad with the original firmware 110531.1 but after I upgraded to 110909 it was very stable.
However I upgraded it to 111226 two weeks ago and it started locking up again ! this time I had just changed the memory depth and did not notice that had locked up until after 30 seconds or more, and it does not matter how many times its switched off and on it remains totally locked up.
The hardware version is 83e8
BTW I didnt notice that the header is a 2mm pitch until I tried to plug in my 0.1” lead, and am I right in assuming that the uart is 3.3V ?
Thanks for all of your help
Paul.
« Reply #1041 on: February 15, 2012, 02:09:52 pm »
you can update to 2.06.3(120112.1)
you will find it here :
http://hotfile.com/dl/142100356/1d87527/2.06.3_120112.1.zip.html
(MD5 371c00942b43b88df49eb23ca4ede99d)
right, it is 2mm pitch and 3.3V level UART
Offline paul
« Reply #1042 on: February 15, 2012, 07:18:49 pm »
Thank you tinhead, I now have a working scope again. (DSO5102B  hw 83e8)
Without your help I would be sending it back to china and hoping for a warranty repair, which would probably cost a lot in delivery and take a few weeks for it to be returned.
I only deleted the /param/sav/run file (I only found one) and it booted up OK, if it didnt I was going to try the other files in /param/sav/udisk.
I still think this is a bug in the current firmware 111226, (its still the latest available on the Hantek website) as I didnt have any problems until I upgraded.
Can it be downgrade to a older firmware ? I am a bit reluctant to upgrade again :(  (although I have downloaded the link)
Thank you again
Paul.
« Reply #1043 on: February 15, 2012, 08:13:41 pm »
don't worry, the latest version here is working better as the latest version on Hantek website.
I remember i developed downgrade firmware ... here is it
https://www.eevblog.com/forum/index.php?topic=1571.msg63001#msg63001
It should work with 83e8 hw1007 (i would not recommend to install it on hw1007 83e9)
So you can try it, if you not happy you can still install the latest fw again.
Offline Kampfwurst
« Reply #1044 on: February 15, 2012, 11:13:30 pm »
it would be helpful if all the firmware Updates are for example on the first page.
« Reply #1045 on: February 15, 2012, 11:18:35 pm »
THEY ARE ON FIRST PAGE!
i'm trying always to maintain this thread, deleting old fw/tools, updating all the links on the first page ...
so all ppl have to do is check there ...
Offline sobomax
« Reply #1046 on: February 18, 2012, 01:05:56 am »
Hi
There is an instruction on page 70 on how to connect LAN using $5 USB dongle.
-Maxim
Offline sobomax
« Reply #1047 on: February 18, 2012, 01:11:04 am »
sobomax :
I think I have looked everywhere but I can't seem to find your network adapter supported list. :(
Also, any code updates ?
Hey Douglas, not much happening here really. I don't have any hardware list, but you can just take release notes for the 2.6.13 kernel and see what is supported there. This should give you a pretty good idea. Or you can just take your laptor with the i386 version of 2.6.13 running in the virtual machine to the shop and try to plug few and see which one works.
The kernel itself is now online here: https://github.com/sobomax/hantek_dso_kernel
Right now it's just stock qq2440 code, but in the next hour or so I will be checking in some small fixes required to build it with the recent 4.x gcc. The original code still assumes 3.4.
-Maxim
Offline McCurrent
« Reply #1048 on: February 18, 2012, 03:54:37 pm »
@patz:
Is it possible to add an option to your software which allows me to change back to the 1062b and not only one way to 1202b?
best regards
Johannes
should be possible, I'll have a look at it ASAP.
Sorry, Patz,
I was impatient   :-[ .
I extended your sources to switch between all bandwidthes at will.
Offline McCurrent
« Reply #1049 on: February 18, 2012, 05:18:12 pm »
Hi tinhead, amazing thread!
Thx for sharing your work with us  ;D.
it works! Communication is done through the official USB driver.
I attached the first version as EXE file. Use at your own risk!
Some feedback if it is working would be nice. Tested with Hantek DSO5102B.
I just took patz' sources to turn my Tekway DST1062 into a DST1202.
But in fact I'm not quite sure whether I was really successfull, thus I have a couple of questions.
  • After I changed the bandwidth, TTScope identifies the scope with the new bandwidth, and I may switch to 2ns timebase, but the scopes system information still reads dst1062b.
    Is this the suspected behaviour?
  • Unfortunately I don't have a adequate source to check the actual bandwidth, thus I just "listened" to some USB communication to get an idea of it (poor thing, I know).
    The rise-time improvement was from 5,7ns to 4ns (3.7 at best, 61MHz -> 95Mhz)
    Now I'm considering my probes.
    Pinsonne offers three different accessories probes. One for each bandwidth, but they all seem to be identical.
    Of courxe I have only the 60MHz version.
    http://www.pinsonne-elektronik.de/pi10/pd111.html
    http://www.pinsonne-elektronik.de/pi10/pd108.html
    http://www.pinsonne-elektronik.de/pi10/pd107.html
    Do you have any experience, which bandwidth you could reach with the original probes?
  • How did you test your improvements?
    Did you take the original probes or something different?
cheers
McCurrent