Let's start with very simple things. All we know - Magnova is a fast/responsive scope. Let's check the situation outside the sweet spot!
Please see the attachment. I am lazy, I didn't use a calculator, Excel formulas - hopefully, if there are no math mistakes.
Hi Kaspich,
I'm sorry, but the values you mentioned can't be correct.
Please try the following steps:
1. Settings -> Default Settings
2. Acquire -> Disable Extended Capture (This is a useful feature, but it slows things down by a factor of two, so to ensure the measurement is comparable to those from the other scopes, please disable it.)
3. Apply a 10 MHz test signal to the channel as the trigger source
4. Check whether the trigger level matches the signal.
You should now see approximately 357 wfms/s, as shown in the following screenshot.
If you don't achieve a similar speed after following the steps above, please post a screenshot (or send it to me).
In your table, you wrote only 4 wfms/s at 100 µs/div. This could be because you did not set the memory setting to 'Auto', but rather to a fixed, very high value. This forces the oscilloscope to capture a wide time range. If you set 327 MPts, the oscilloscope would have to record 204 ms at 1.6 GSa/s (327E6 pts/1.6E9 Sa/s = 0.204 s). With that, 4 wfms/s is realistic; even with zero dead time, 5 wfms/s would not be achievable.