Yes , sadly you are correct: when the scope is in run and depending on timebase and signal. But if you use fast segmentation- or ultra acquire-mode etc. these time stamps between Waveform recordings is still 0.021 to 1.5us (depending on the scope) and no matter what timebase and measurements you activated. When you play back the recording, all waveform measurements are calculated afterwards during playback for each waveform. So these in these modes you miss out on as little glitches etc. as possible and still see your actual measurement values.
As everything, it is a compromise.
Yes but I use measurements in normal running mode and that is primary use case for me.
So super fast segmented mode in what is basically a digitizer mode is something I do more rarely.
Also it is definitely not "no matter what timebase you activated". Retrigger time cannot be faster than sweep time + blind time.
If you set scope to 10 ms/div and have 10 div on screen, even if you have scope with 0 blind time, you cannot have more than 10 triggers per second. And blind time will be very dependent on memory depth. Scope will need more time to process 500MS than 1kS.
Catching out little glitches?
It depends on what is frequency of repetition.
And what information you want to extract.
If you need 100% POI (probability of intercept, detection) in say 100ms period, then your best friend is long capture with long memory. In which case you can have other channels connected to some system references to try to correlate where glitch comes from. You find glitches inside that capture with search.
On better scopes (some of which are still cheaper than MXO series) you have Signal/Wave scan (or some other trade name) that will capture, search, and if detected stop and show, or if no detection, keep recapturing until it finds something. In which case it offers you to copy that detection definition into trigger and then let it run and have 100% detection via trigger.
You can also run scope in simple persistence mode for longer time until you capture something, then by looking into waveform setup a trigger or zone trigger to trigger on problems and also capture them 100% of time.
Once you have a trigger definition, you can then use segmented mode to ensure shortest possible retrigger time, but funny enough, that is not really necessary then, because if your glitches are so close, you will see them in plain sight while looking at the screen in bog standard mode.. Very reliably, even on SDS800xHD.
There is some merit to fast retriggering in normal trigger mode when using scope in a standard fashion, and that is that in that case CRT emulation is better and scope shows more "natural" view on fast changing signals that resembles CRT scope.
I also find that pleasing to look at, but not really that important, once you are showing more than some amount of triggers per second, it gets good enough..