has someone experience with the technique described in this Patent:
http://www.google.de/patents/US5369245
Greetz
Oliver
That's the well known Pickering patent used in the Wavetronik/Fluke 7000 reference, for conditioning the LTZ1000 after power down.
I have collected some experience: I did several tests concerning temperature cycling (refrigerators / heater), but not using that schematic.
This patent at first indicates, that temperature excursions of the LTZ induce hysteresis, and if this hysteresis is not removed, by symmetric cycling, it will in turn induce creeping drift.
From that it directly follows, that any "burn-In" process is bad for the LTZs stability!
I have seen hysteresis of several ppm on both of my LTZ references, but only if the temperature deviation from the nominal value is > +/-25°C. (I accidentially heated one of the LTZs to >90°C.)
If the temp. excursion is less, there is no noteworthy hysteresis, perhaps a few tenths of ppm only.
That means, at 45°C stabilization temperature, as in the Fluke 7000, the conditioning does not improve uncertainty that much.. The reference at this temperature is already very stable without further measures, also in intermittent operation.
I measure that on the HP3458A, each time I power it on again, and two times already in the last 4 years, when I removed my constantly powered LTZ references from the mains for several weeks. Each of those 3 LTZs return to their default value to < 1ppm deviation, every time. Exactly the same goes for the Fluke 5442A calibrator, which uses 2 SZA263.
Only if very high (>70°C) or very low (<15°C) ambient temperature excursions are encountered, this feature might theoretically have some effect...
But then, as the lower conditioning temperature is limited to room temperature, i.e. 20°C, the complete or symmetric conditioning circle, as described, can not be passed through. 0°C would be necessary, to reset a trip to 90°C, as shown in the patent in fig. 2.
Therefore, this patent is nice for theory, but did obviously NOT work in the Fluke 7000.
Perhaps this was one of several reasons, why Fluke terminated that product.
Frank