« Last post by Andreas on Today at 03:39:40 pm »Hello,
update on my ADR1399 ageing reference board after having added 100uF to the output of the LDO voltage regulator on day 246.
(HEF4052 multiplexer has been used since day 190 / previously it was a MAX4052A)
Now I have a much lower standard devation especially on the metal can packages (from 0.25-0.3 ppm to 0.1 - 0.15 ppm)
But also the LS8-packages went down from 0.15-0.3 ppm to 0.1 - 0.17 ppm
And again the output voltage shifted. So I fear that I will have to repeat the ageing drift measurements with a new set of references.
With best regards
Andreas