Metrology / Re: measuring 100mA current source with a better than Keithley 2000
« Last post by The Soulman on Today at 05:19:52 pm »Your best bet is the Hioki on the 120mV range in 7,5 digit mode (is the sample rate high enough?) with an external 1 Ohm shunt.
100mA reads 100mv and the least significant digit is 10nA, speed and accuracy may not be great.
Power loss in the shunt 0,1V*0,1A=0,01W or 10mW, easily manageable.
You could add a op amp to the above and amplify the shunt signal 10x or 100x for 1nA or 0,1nA resolution but you would need some range switching in that case.
What is the duration of your current pulse and how accurately do you need to measure it?
And why do you need that much resolution?