1
Test Equipment / Re: Fun With Low Leakage/Bias Current: Femtompere, Electrometer, Keithley 617
« Last post by Alex Nikitin on Today at 02:24:42 pm »After put back the original BNC connector, I recorded the offset. Is it normal or not? thanks
If that is the drift when switching on from cold, it might be normal. According to the manual, the warm up time to rated accuracy is 2 hours. There are two trimmers on the input board, one to adjust the DC offset (in the Voltage measuring mode) and the second is to compensate the input current (in the Current measuring mode), so it can be adjusted to zero. Both are not easy adjustments, mostly because you need to keep the top cover on to get the temperature stable and correct, than open the cover (thus upsetting the thermal equilibrium) to do the adjustments, only to close it quickly and see if you hit the spot after the temperature settles again . Both the offset and the input current will drift noticeably when you open/close the cover however by several iterations it is possible to get a reasonable result.
Cheers
Alex