I had issues with one of the LMH6518s, offset was suddenly larger than spec and made the scope fail self alignment. Replacing the IC fixed it. Note it was just a single channel, 3 sounds unlikely? There was one other poster who had a
Yes, that is strange. But I thought about it and I think I know the reason now.
In most cases I only use two channels. Channel 3/4 are rarely used. But when I use them, I use them often for current measurement. I have a few (selfmade) potential free current probes (DC to 1Mhz). The output of the probe is an OPV+50R with 3.3V power supply and the scope has to measure 0 to 2.5V. This means for my measurement 0 to 25mA. So the scope has to measure 2.5V at max. Nothing special with it isn't?
But there is an important difference! An example is to measure the current consumption of a microcontroller. Than I move the GND-Line to the bottom of the screen and if the microcontroller draws 20mA, most of the time the voltage is arround 2 to 2.5V. So I use the full screen for 0 to 2.5V! And if I am do programming or don't switch anything of at lunch, it means I have probably a max DC Voltage of 1V or close to that at the PGA for hours. And this is what TI said is bad for the LMH6518.
That means you can not measure full screen DC signal on the scope. And probably other scope with the LMH6518!
I have an idea how to check. If I measure a voltage of -2.5V for an hour, it is possible the offset will be better because it destroy the lower half in the LMH6518. But it is now only an idea. I will check it at the weekend.
If I am right it means we can not use the scope to measure full screen DC signal. That is a shame for any scope and especialy for something so expensive! So I wrote a letter to Batronik/R&S and ask them for repairing, because I think it is there (or TIs!) fault and not my one. I should be in warranty of 2days from my last repair, but I have to talk with my lawyer about it.
I they not will repair it, I can probably exchange the LMH6518 by myself, but this opens the question how can I calibrate the offset in the scope with the new LMH?
And even if anything is working again, it is a shame to have a scope that is not usable for full screen DC signal!
I have never opend the RTB until now, but probably the LMH6518 needs a heatsink? I mean it draws arround 1W!
Oh.. and there arise another question. What happens with there new and shiny RTB2? The same hardware with the same problem?
Or did they noticed it, changed the hardware and forget to tell us about the huge problems with the RTB2004? The answer
for this is important to think about buying anything from R&S in the future or not!
BTW: Did TI inform there customer about the problem in the LHM6518?
Olaf