The MUX itself should not create much overshoot. It is possible that the buffer creates some obershoot / slow settling. An amplifier with bootstrapped supply is always a bit tricky with the stability / settling. I would expect the overshoot more from the buffer than the mux. Anyway for looking at the noise offset drift of the ADC only, one could use the case where the input signal is not connected to the ADC, so the 4053 channel allways off (modus T).
The integrator waveform looks as expected.
To get low noise, the values for the slow slope (sum) and aux ADC should be reasonable correct.
The slow slope part looks reasonable, though the result of the k1 measurement shown is 20.88 while the other files use 20.90. This is not far off, and should not cause that much different in noise. The part with K2 seems to be not working. So there seems to be something wrong with the residual charge measurement. The reading of the µC internal ADC look suspiciously constant at 506. So there may be a problem with the last amplification step (MCP6002 ?). So far this is kind of only the classicl MS ADC without a working res. charge measurement for the last bit of resolution.
It looks a bit, like the ADC channel that is supposed to read the residual ADC reads something different, like the scaled down input voltage or maybe the average integrator voltage. So there may be a small difference in the plan, like using different ADC channels / pins at the µC.
Which is the schematics for the PCB used ?
An interesting signal to look at is the ouput of the slope amplifier (NE5534). This signal should be about -700mV to 1400 mV near square wave in the runup, and somewhere in the 0 - 400 mV range after the rundown part and slightly, variable ( ~50 mV) between conversions. The trimmer will effect the level where the signal be approximately.
The DA test part seems to be net yet correct/fully implemented in the new Pascal program. I think I never used it with the new program. This part is more like an extra if everything else works. It can be used to check the cap qualtiy and see leakage currents at the integrator.
241: {DA-Test 26 chars}
begin
write('DA result: 0x');
das := '0x';
for n := 0 to 25 do
begin
da[n] := ord(Readcom);
write(IntToHex(da[n]),' ');
das := concat(das, IntToHex(da[n]));
end;
writeln('');
writeln(f, das);
end;
I looked at the schematics, and it looks like the AD0 and AD1 signals are swapped: I have the residual charge signal at AD1, rerouter has the resudual charge signal at AD0. So this would need a slight modification to the software.
As far as I have found there are only 3 accesses to the chanel choice of the µC internal ADC (register ADMUX) in the program. So ideally define a 2nd constant for the ADC setting and than use this instead of "ADMUXval-1" .
p.s. :
the waveforms for the slope amplifier out and MCP6002 out look good - so likely just the ADC channels swapped.
Could you calc the stddev from 800secs up till the end, plz?
..For some reason I don't know the usual DMMs don't inlcude much reference filtering, though this is a relatively simple way to reduce the noise a little. Usually the effect is smaller than here, because not many combine a very low noise ADC with a relatively noisy reference (LM399).