Author Topic: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes  (Read 622745 times)

0 Members and 7 Guests are viewing this topic.

Offline photomankc

  • Contributor
  • Posts: 35
  • Country: us
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1075 on: December 04, 2018, 04:12:20 pm »
I believe the 1000Z does do this.  I'm pretty sure I've seen a few shots posted here that show it but I can't confirm as mine is in UPS' hands headed back.  I know my 2202 does it.

I agree that there is no real room on the marker flag, but there could be some on the channel "cards" on the right side under the channel number tab indent.  It would be helpful when you make screen shots not to have to include a legend each time explaining what the traces are. 
 

Offline Performa01

  • Super Contributor
  • ***
  • Posts: 1645
  • Country: at
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1076 on: December 04, 2018, 05:42:37 pm »
I agree that there is no real room on the marker flag, but there could be some on the channel "cards" on the right side under the channel number tab indent.  It would be helpful when you make screen shots not to have to include a legend each time explaining what the traces are. 
Yes, it would be a nice feature, no doubt.

Yet in the Siglent UI, there is no space left in the channel tabs as there should be additional icons for bandwidth limit, channel inversion and probe attenuation factor. Since I don't see these in your screenshot, I'm wondering if you have the latest firmware installed (7.6.1.26)?
 

Offline photomankc

  • Contributor
  • Posts: 35
  • Country: us
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1077 on: December 04, 2018, 05:57:56 pm »
Damn.  It was a thought.  Not a big thing.  Just a wish.  I’ll check firmware version tonight.  That was a stolen screenshot from the interwebs. 
« Last Edit: December 04, 2018, 06:05:52 pm by photomankc »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28387
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1078 on: December 04, 2018, 06:28:34 pm »
Damn.  It was a thought.  Not a big thing.  Just a wish.  I’ll check firmware version tonight.  That was a stolen screenshot from the interwebs.
A trace name could probably be implemented and best it would be the same color as the trace and at the 0V position just inside the active display window alongside the left channel indicator. Best it would be character # limited to maybe 5 characters which would suit most needs.
The virtual keyboard like that used for the WiFi PSK would be a good selection to accomplish this.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline photomankc

  • Contributor
  • Posts: 35
  • Country: us
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1079 on: December 04, 2018, 07:11:52 pm »
Damn.  It was a thought.  Not a big thing.  Just a wish.  I’ll check firmware version tonight.  That was a stolen screenshot from the interwebs.
A trace name could probably be implemented and best it would be the same color as the trace and at the 0V position just inside the active display window alongside the left channel indicator. Best it would be character # limited to maybe 5 characters which would suit most needs.
The virtual keyboard like that used for the WiFi PSK would be a good selection to accomplish this.

I would love that if they could do it!  4 or 5 characters would be great.  Just enough that captures are somewhat self-documenting.  There are spots the option could go in the channel menus and stay 2 pages.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28387
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1080 on: December 04, 2018, 07:30:35 pm »
Damn.  It was a thought.  Not a big thing.  Just a wish.  I’ll check firmware version tonight.  That was a stolen screenshot from the interwebs.
A trace name could probably be implemented and best it would be the same color as the trace and at the 0V position just inside the active display window alongside the left channel indicator. Best it would be character # limited to maybe 5 characters which would suit most needs.
The virtual keyboard like that used for the WiFi PSK would be a good selection to accomplish this.

I would love that if they could do it!  4 or 5 characters would be great.  Just enough that captures are somewhat self-documenting.  There are spots the option could go in the channel menus and stay 2 pages.
Yes, that is only the sensible place to put any channel naming scheme.....the channel menu on P2.  :)

Edit
Spelling correction.
« Last Edit: December 04, 2018, 08:17:17 pm by tautech »
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline photomankc

  • Contributor
  • Posts: 35
  • Country: us
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1081 on: December 04, 2018, 08:15:18 pm »
Damn.  It was a thought.  Not a big thing.  Just a wish.  I’ll check firmware version tonight.  That was a stolen screenshot from the interwebs.
A trace name could probably be implemented and best it would be the same color as the trace and at the 0V position just inside the active display window alongside the left channel indicator. Best it would be character # limited to maybe 5 characters which would suit most needs.
The virtual keyboard like that used for the WiFi PSK would be a good selection to accomplish this.

I would love that if they could do it!  4 or 5 characters would be great.  Just enough that captures are somewhat self-documenting.  There are spots the option could go in the channel menus and stay 2 pages.
Yes, that is only the sensible place to put and channel naming scheme.....the channel menu on P2.  :)

Not the label... the option to set the label.  Which other menu makes more sense for that?
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 6663
  • Country: hr
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1082 on: December 04, 2018, 08:24:11 pm »
On DS1000Z it is in channel setup. It is implemented very nicely except the fact that you have to use virtual keyboard to enter text.

I'm telling you, convince Siglent to enable keyboard driver on USB, so you can use keyboard for any numeric or alpha entry.
Touch screen scopes most useful feature is that you gain keyboard. External keyboard (you can get one of those small wireless media thingies) and you are cooking...
 
The following users thanked this post: photomankc

Offline Rerouter

  • Super Contributor
  • ***
  • Posts: 4694
  • Country: au
  • Question Everything... Except This Statement
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1083 on: December 04, 2018, 09:34:21 pm »
Silly question, is the SAG1021 a usb SCPI device, or is it literally a fancy DAC for the scope with no real USB functionality if connected directly to a PC?

Currently away from home until the end of the week, so not able to answer this for myself.
 

Offline katagia

  • Newbie
  • Posts: 7
  • Country: de
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1084 on: December 05, 2018, 06:33:29 pm »
Hi

Has anyone here tried to use a waveform generator from siglent which is not the SAG1021?
Will the SDG830 work for bode plots?
Do I need the additional software license for bode plots? I the manual the option is only mentioned for the "Arbitrary Waveform Generator". The bode plot is mentioned in a complete different chapter.

I would prefer the SDG830 because it will work standalone ans as a complete user interface. It would be nice to use it to record bode plots.

Thank you very much.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28387
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1085 on: December 05, 2018, 06:42:55 pm »
Hi

Has anyone here tried to use a waveform generator from siglent which is not the SAG1021?
Will the SDG830 work for bode plots?
Do I need the additional software license for bode plots? I the manual the option is only mentioned for the "Arbitrary Waveform Generator". The bode plot is mentioned in a complete different chapter.

I would prefer the SDG830 because it will work standalone ans as a complete user interface. It would be nice to use it to record bode plots.

Thank you very much.
Welcome to the forum.

Yes any of the Siglent AWG's will provide Bode plot functionality with zero licensing requirements.....plug and play.
Bode plot is a free feature, all that's needed for it is a sweeping sinewave controlled by the scope.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 
The following users thanked this post: katagia

Offline commongrounder

  • Frequent Contributor
  • **
  • Posts: 372
  • Country: us
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1086 on: December 05, 2018, 07:35:36 pm »
Hi

Has anyone here tried to use a waveform generator from siglent which is not the SAG1021?
Will the SDG830 work for bode plots?
Do I need the additional software license for bode plots? I the manual the option is only mentioned for the "Arbitrary Waveform Generator". The bode plot is mentioned in a complete different chapter.

I would prefer the SDG830 because it will work standalone ans as a complete user interface. It would be nice to use it to record bode plots.

Thank you very much.
Welcome to the forum.

Yes any of the Siglent AWG's will provide Bode plot functionality with zero licensing requirements.....plug and play.
Bode plot is a free feature, all that's needed for it is a sweeping sinewave controlled by the scope.

I have been using a Siglent SDG1005 with my SDS1104x-e for bode plotting, and it works just fine, and, as mentioned, no additional software needed.
 
The following users thanked this post: katagia

Offline katagia

  • Newbie
  • Posts: 7
  • Country: de
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1087 on: December 08, 2018, 12:26:35 pm »
Thank you to both of you for your clearification.

I have bought the SDG830 for 200€ which is compareable to the SAG1021 but has it's own user interface, little higher output frequency and output voltage.

The bode plot feature has worked out of the box.
I tried the AWG interface of the 1204 but it doesn't recognize the SDG830. Is the feature limited to the SAG1021?
The temporary license with 30 uses should be sufficient?
The only thing I am missing in this setup is I can't control the AWG via LAN.

Did I miss anything else which is possible with the SAG1021?

Thank you very much.
 

Offline Performa01

  • Super Contributor
  • ***
  • Posts: 1645
  • Country: at
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1088 on: December 08, 2018, 12:47:58 pm »
I tried the AWG interface of the 1204 but it doesn't recognize the SDG830. Is the feature limited to the SAG1021?
The temporary license with 30 uses should be sufficient?
The only thing I am missing in this setup is I can't control the AWG via LAN.

Did I miss anything else which is possible with the SAG1021?
The AWG control in the SDS1204X-E is for the "faceless" SAG1021 only. You cannot control any other AWG with that.

And to be honest, who would want to do that anyway? Even your cheap SDG830 can do more things than the control interface in the SDS1004X-E would allow you to do, and then a real physical user interface as on the SDG830 is much nicer to use anyway.

Many Siglent AWGs can be controlled via LAN by means of the EasyWaveX application, but the 800 entry level series is limited to USB.
 
The following users thanked this post: katagia

Offline katagia

  • Newbie
  • Posts: 7
  • Country: de
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1089 on: December 08, 2018, 01:10:20 pm »
In my case the 1204 could be used as lan to usb bridge when it wouls support other AWGs that way.
Could be handy for automation.

I aggree you about the SDG830. I think I'm lucky to get it for 200€. It has been delivered with a very ancient firmware version. Seems to be some remaining stock.

I know there are other AWGs with LAN. I have no use for it at the moment.
I just want to analyze the differences.
« Last Edit: December 08, 2018, 01:13:15 pm by katagia »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28387
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1090 on: December 08, 2018, 07:23:21 pm »

I aggre you about the SDG830. I think I'm lucky to get it for 200€. It has been delivered with a very ancient firmware version. Seems to be some remaining stock.
Please see that you update the FW to version 1.08.01.13R6 at your earliest convenience.
https://www.siglenteu.com/service-and-support/firmware-software/waveform-generators/#sdg800-series

Inside that version is provision to recover the SDG800 should it ever experience a boot freeze/failure.
In such cases the USB boot recovery software is available from Siglent.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline Rerouter

  • Super Contributor
  • ***
  • Posts: 4694
  • Country: au
  • Question Everything... Except This Statement
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1091 on: December 09, 2018, 12:08:07 am »
Tautech, you might want the devs to get a spell checker :)

SYSTtem:ERRor?, has 1 too many "t"

MEAASURE_STATUS. 1 too many A's at a guess

DIGITAL:RESET_DDR, has a ? on the short but not the long name
DIGITAL:HIST_MODE, same as ^

And with that I leave what I believe to be every command available on this scope, (could not find where syst:date was, so may have missed some?), mopstly added this time is the decoding, for more digital stuff, decoding/external/ac line triggering.

The 2 different flavors of external triggering have me curious if there is a way to hook up an external trigger.

And sadly no sign of bode plot commands for me.

Code: [Select]
*CAL? Performs complete internal calibration of the instrument.
*CLS Clears all status data registers
*ESE Sets the Standard Event Status Enable register (ESE).
*ESE? Sets the Standard Event Status Enable register (ESE).
*ESR? Reads, clears the Event Status Register (ESR).
*IDN? For identification purposes.
*OPC? Sets the OPC bit in the Event Status Register (ESR).
*RCL Recalls one of five nonvolatile panel setups.
*RSI
*RST The *RST command initiates a device reset.
*SAV Stores current state in nonvolatile internal memory.
*SRE Sets the Service Request Enable register (SRE).
*SRE? Sets the Service Request Enable register (SRE).
*TST? Performas an Internal Self Test
*WAI
ACENBT ACquire_Enhance_BiTs
ACENBT? ACquire_Enhance_BiTs?
ACQW ACQuire_Way Specifies the acquisition mode.
ACQW? ACQuire_Way? Specifies the acquisition mode.
ACSQM ACquire_SeQuence_Mode
ACSQS ACquire_SeQuence_Seg
ALST? ALL_STATUS? Reads and clears the contents of all status registers.
AMEATAT ALL_MEASURE_STATUS
ARM ARM_AQUISITION
ARWV? ARBWAVE? Change arbitrary wave type.
ASET AUTOSETUP Adjusts vertical, time base and trigger parameters.
AVGA AVeraGe_Acquire Selects the average times of average acquisition.
AVGA? AVeraGe_Acquire? Selects the average times of average acquisition.
B1:DCCN B1:DECO_CAN
B1:DCIC B1:DeCo_IiC
B1:DCLN B1:DECO_LIN
B1:DCSP B1:DECO_SPI
B1:DCUT B1:DECO_UART
B2:DCCN B2:DECO_CAN
B2:DCIC B2:DECO_IIC
B2:DCLN B2:DECO_LIN
B2:DCSP B2:DECO_SPI
B2:DCUT B2:DECO_UART
BOARDTEST?
BSWV? BASICWAVE? Set or get basic wave parameters
BTWV? BURSTWAVE? Set or get burst wave parameters
BUZZ BUZZER Controls the built-in piezo-electric buzzer.
BUZZ? BUZZER? Controls the built-in piezo-electric buzzer.
BWL BANDWIDTH_LIMIT Enables/disables the bandwidth-limiting lowpass filter
BWL? BANDWIDTH_LIMIT? Enables/disables the bandwidth-limiting lowpass filter
C1:ARWV C1:ARBWAVE Change arbitrary wave type.
C1:ARWV? C1:ARBWAVE? Change arbitrary wave type.
C1:ATTN C1:ATTENUATION Selects the vertical attenuation factor of the probe
C1:ATTN? C1:ATTENUATION? Selects the vertical attenuation factor of the probe
C1:BSWV C1:BASICWAVE Set or get basic wave parameters
C1:BSWV? C1:BASICWAVE? Set or get basic wave parameters
C1:BTWV C1:BURSTWAVE Set or get burst wave parameters
C1:BTWV? C1:BURSTWAVE? Set or get burst wave parameters
C1:BWL C1:BANDWIDTH_LIMIT Enables/disables the bandwidth-limiting lowpass filter
C1:BWL? C1:BANDWIDTH_LIMIT? Enables/disables the bandwidth-limiting lowpass filter
C1:CMBN C1:COMBINE
C1:CMBN? C1:COMBINE?
C1:CPL C1:COUPLING Selects the specified input channel‘s coupling mode.
C1:CPL? C1:COUPLING? Selects the specified input channel‘s coupling mode.
C1:CRST C1:CURSOR_SET Allows positioning of any one of eight cursors.
C1:CRST? C1:CURSOR_SET? Allows positioning of any one of eight cursors.
C1:CRVA? C1:CURSOR_VALUE? Returns trace values measured by specified cursors.
C1:DAC_ATTEN
C1:DAC_ATTEN_SET
C1:DAC_ATTEN_SET
C1:DACC
C1:DACC?
C1:HADJ C1:HORIZON_ADJ
C1:HADJ? C1:HORIZON_ADJ?
C1:HARM C1:HARMONIC
C1:HARM? C1:HARMONIC?
C1:INVS C1:INVERT_SET Invert the trace or the math waveform of specified source.
C1:INVS? C1:INVERT_SET? Invert the trace or the math waveform of specified source.
C1:INVT C1:INVERT
C1:INVT? C1:INVERT?
C1:MDWV C1:MODULATEWAVE Set or get modulate wave parameters.
C1:MDWV? C1:MODULATEWAVE? Set or get modulate wave parameters.
C1:OFST C1:OFFSET Allows output channel vertical offset adjustment.
C1:OFST? C1:OFFSET? Allows output channel vertical offset adjustment.
C1:OUTP C1:OUTPUT
C1:OUTP? C1:OUTPUT?
C1:PAVA? C1:PARAMETER_VALUE Returns current parameter, mask test values
C1:PRI C1:PRIORITY
C1:PRI? C1:PRIORITY?
C1:SKEW Sets the skew of specified trace
C1:SKEW? Sets the skew of specified trace
C1:SRATE C1:SAMPLERATE
C1:SRATE? C1:SAMPLERATE?
C1:SWWV C1:SWEEPWAVE Set or get sweep wave.
C1:SWWV? C1:SWEEPWAVE? Set or get sweep wave.
C1:SYNC
C1:SYNC?
C1:TRA C1:TRACE Enables or disables the display of a trace.
C1:TRA? C1:TRACE? Enables or disables the display of a trace.
C1:TRCP C1:TRIG_COUPLING Sets the coupling mode of the specified trigger source.
C1:TRCP? C1:TRIG_COUPLING? Sets the coupling mode of the specified trigger source.
C1:TRLV C1:TRIG_LEVEL Adjusts the trigger level of the specified trigger source
C1:TRLV? C1:TRIG_LEVEL? Adjusts the trigger level of the specified trigger source
C1:TRLV2 C1:TRIG_LEVEL2 Adjusts the second trigger level of the specified trigger source.
C1:TRLV2? C1:TRIG_LEVEL2? Adjusts the second trigger level of the specified trigger source.
C1:TRSL C1:TRIG_SLOPE Sets the trigger slope of the specified trigger source
C1:TRSL? C1:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
C1:UNIT Sets the unit of specified trace
C1:UNIT? Sets the unit of specified trace
C1:VDIV C1:VOLT_DIV Sets the vertical sensitivity
C1:VDIV? C1:VOLT_DIV? Sets the vertical sensitivity
C1:VGAC
C1:VGAC?
C1:WF? C1:WAVEFORM? Gets the waveform from the instrument.
C1-C2:MEAD? The MEASURE_DELY command places the instrument in the continuous measurement mode and starts a type of delay measurement.
C1-C3:MEAD? The MEASURE_DELY command places the instrument in the continuous measurement mode and starts a type of delay measurement.
C1-C4:MEAD? The MEASURE_DELY command places the instrument in the continuous measurement mode and starts a type of delay measurement.
C2:ARWV C2:ARBWAVE Change arbitrary wave type.
C2:ARWV? C2:ARBWAVE? Change arbitrary wave type.
C2:ATTN C2:ATTENUATION Selects the vertical attenuation factor of the probe
C2:ATTN? C2:ATTENUATION? Selects the vertical attenuation factor of the probe
C2:BSWV C2:BASICWAVE Set or get basic wave parameters
C2:BSWV? C2:BASICWAVE? Set or get basic wave parameters
C2:BTWV C2:BURSTWAVE Set or get burst wave parameters
C2:BTWV? C2:BURSTWAVE? Set or get burst wave parameters
C2:BWL C2:BANDWIDTH_LIMIT Enables/disables the bandwidth-limiting lowpass filter
C2:BWL? C2:BANDWIDTH_LIMIT? Enables/disables the bandwidth-limiting lowpass filter
C2:CMBN C2:COMBINE
C2:CMBN? C2:COMBINE?
C2:CPL C2:COUPLING Selects the specified input channel‘s coupling mode.
C2:CPL? C2:COUPLING? Selects the specified input channel‘s coupling mode.
C2:CRST C2:CURSOR_SET Allows positioning of any one of eight cursors.
C2:CRST? C2:CURSOR_SET? Allows positioning of any one of eight cursors.
C2:CRVA? C2:CURSOR_VALUE? Returns trace values measured by specified cursors.
C2:DAC_ATTEN
C2:DAC_ATTEN_SET
C2:DACC
C2:DACC?
C2:HADJ C2:HORIZON_ADJ
C2:HADJ? C2:HORIZON_ADJ?
C2:HARM C2:HARMONIC
C2:HARM? C2:HARMONIC?
C2:INVS C2:INVERT_SET Invert the trace or the math waveform of specified source.
C2:INVS? C2:INVERT_SET? Invert the trace or the math waveform of specified source.
C2:INVT C2:INVERT
C2:INVT? C2:INVERT?
C2:MDWV C2:MODULATEWAVE Set or get modulate wave parameters.
C2:MDWV? C2:MODULATEWAVE? Set or get modulate wave parameters.
C2:OUTP C2:OUTPUT
C2:OUTP? C2:OUTPUT?
C2:PAVA? C2:PARAMETER_VALUE Returns current parameter, mask test values
C2:PRI C2:PRIORITY
C2:PRI? C2:PRIORITY?
C2:SKEW Sets the skew of specified trace
C2:SKEW? Sets the skew of specified trace
C2:SRATE C2:SAMPLERATE
C2:SRATE? C2:SAMPLERATE?
C2:SWWV C2:SWEEPWAVE Set or get sweep wave.
C2:SWWV? C2:SWEEPWAVE? Set or get sweep wave.
C2:SYNC
C2:SYNC?
C2:TRA C2:TRACE Enables or disables the display of a trace.
C2:TRA? C2:TRACE? Enables or disables the display of a trace.
C2:TRCP C2:TRIG_COUPLING Sets the coupling mode of the specified trigger source.
C2:TRCP? C2:TRIG_COUPLING? Sets the coupling mode of the specified trigger source.
C2:TRLV C2:TRIG_LEVEL Adjusts the trigger level of the specified trigger source
C2:TRLV? C2:TRIG_LEVEL? Adjusts the trigger level of the specified trigger source
C2:TRLV2 C2:TRIG_LEVEL2 Adjusts the second trigger level of the specified trigger source.
C2:TRLV2? C2:TRIG_LEVEL2? Adjusts the second trigger level of the specified trigger source.
C2:TRSL C2:TRIG_SLOPE Sets the trigger slope of the specified trigger source
C2:TRSL? C2:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
C2:UNIT Sets the unit of specified trace
C2:UNIT? Sets the unit of specified trace
C2:VDIV C2:VOLT_DIV Sets the vertical sensitivity
C2:VDIV? C2:VOLT_DIV? Sets the vertical sensitivity
C2:VGAC
C2:VGAC?
C2:WF? C2:WAVEFORM? Gets the waveform from the instrument.
C2-C3:MEAD? The MEASURE_DELY command places the instrument in the continuous measurement mode and starts a type of delay measurement.
C2-C4:MEAD? The MEASURE_DELY command places the instrument in the continuous measurement mode and starts a type of delay measurement.
C3:ATTN C3:ATTENUATION Selects the vertical attenuation factor of the probe
C3:ATTN? C3:ATTENUATION? Selects the vertical attenuation factor of the probe
C3:BWL C3:BANDWIDTH_LIMIT Enables/disables the bandwidth-limiting lowpass filter
C3:BWL? C3:BANDWIDTH_LIMIT? Enables/disables the bandwidth-limiting lowpass filter
C3:CPL C3:COUPLING Selects the specified input channel‘s coupling mode.
C3:CPL? C3:COUPLING? Selects the specified input channel‘s coupling mode.
C3:CRST C3:CURSOR_SET Allows positioning of any one of eight cursors.
C3:CRST? C3:CURSOR_SET? Allows positioning of any one of eight cursors.
C3:CRVA? C3:CURSOR_VALUE? Returns trace values measured by specified cursors.
C3:DAC_ATTEN
C3:DAC_ATTEN
C3:DAC_ATTEN_SET
C3:DACC
C3:DACC?
C3:HADJ C3:HORIZON_ADJ
C3:HADJ? C3:HORIZON_ADJ?
C3:INVS C3:INVERT_SET Invert the trace or the math waveform of specified source.
C3:INVS? C3:INVERT_SET? Invert the trace or the math waveform of specified source.
C3:PAVA? C3:PARAMETER_VALUE Returns current parameter, mask test values
C3:PRI C3:PRIORITY
C3:PRI? C3:PRIORITY?
C3:SKEW Sets the skew of specified trace
C3:SKEW? Sets the skew of specified trace
C3:TRA C3:TRACE Enables or disables the display of a trace.
C3:TRA? C3:TRACE? Enables or disables the display of a trace.
C3:TRCP C3:TRIG_COUPLING Sets the coupling mode of the specified trigger source.
C3:TRCP? C3:TRIG_COUPLING? Sets the coupling mode of the specified trigger source.
C3:TRLV C3:TRIG_LEVEL Adjusts the trigger level of the specified trigger source
C3:TRLV? C3:TRIG_LEVEL? Adjusts the trigger level of the specified trigger source
C3:TRLV2 C3:TRIG_LEVEL2 Adjusts the second trigger level of the specified trigger source.
C3:TRLV2? C3:TRIG_LEVEL2? Adjusts the second trigger level of the specified trigger source.
C3:TRSL C3:TRIG_SLOPE Sets the trigger slope of the specified trigger source
C3:TRSL? C3:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
C3:UNIT Sets the unit of specified trace
C3:UNIT? Sets the unit of specified trace
C3:VDIV C3:VOLT_DIV Sets the vertical sensitivity
C3:VDIV? C3:VOLT_DIV? Sets the vertical sensitivity
C3:VGAC
C3:VGAC?
C3:WF? C3:WAVEFORM? Gets the waveform from the instrument.
C3-C4:MEAD? The MEASURE_DELY command places the instrument in the continuous measurement mode and starts a type of delay measurement.
C4:ATTN C4:ATTENUATION Selects the vertical attenuation factor of the probe
C4:ATTN? C4:ATTENUATION? Selects the vertical attenuation factor of the probe
C4:BWL C4:BANDWIDTH_LIMIT Enables/disables the bandwidth-limiting lowpass filter
C4:BWL? C4:BANDWIDTH_LIMIT? Enables/disables the bandwidth-limiting lowpass filter
C4:CPL C4:COUPLING Selects the specified input channel‘s coupling mode.
C4:CPL? C4:COUPLING? Selects the specified input channel‘s coupling mode.
C4:CRST C4:CURSOR_SET Allows positioning of any one of eight cursors.
C4:CRST? C4:CURSOR_SET? Allows positioning of any one of eight cursors.
C4:CRVA? C4:CURSOR_VALUE? Returns trace values measured by specified cursors.
C4:DAC_ATTEN
C4:DAC_ATTEN
C4:DAC_ATTEN_SET
C4:DACC
C4:DACC?
C4:HADJ C4:HORIZON_ADJ
C4:HADJ? C4:HORIZON_ADJ?
C4:INVS C4:INVERT_SET Invert the trace or the math waveform of specified source.
C4:INVS? C4:INVERT_SET? Invert the trace or the math waveform of specified source.
C4:PAVA? C4:PARAMETER_VALUE Returns current parameter, mask test values
C4:PRI C4:PRIORITY
C4:PRI? C4:PRIORITY?
C4:SKEW Sets the skew of specified trace
C4:SKEW? Sets the skew of specified trace
C4:TRA C4:TRACE Enables or disables the display of a trace.
C4:TRA? C4:TRACE? Enables or disables the display of a trace.
C4:TRCP C4:TRIG_COUPLING Sets the coupling mode of the specified trigger source.
C4:TRCP? C4:TRIG_COUPLING? Sets the coupling mode of the specified trigger source.
C4:TRLV C4:TRIG_LEVEL Adjusts the trigger level of the specified trigger source
C4:TRLV? C4:TRIG_LEVEL? Adjusts the trigger level of the specified trigger source
C4:TRLV2 C4:TRIG_LEVEL2 Adjusts the second trigger level of the specified trigger source.
C4:TRLV2? C4:TRIG_LEVEL2? Adjusts the second trigger level of the specified trigger source.
C4:TRSL C4:TRIG_SLOPE Sets the trigger slope of the specified trigger source
C4:TRSL? C4:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
C4:UNIT Sets the unit of specified trace
C4:UNIT? Sets the unit of specified trace
C4:VDIV C4:VOLT_DIV Sets the vertical sensitivity
C4:VDIV? C4:VOLT_DIV? Sets the vertical sensitivity
C4:VGAC
C4:VGAC?
C4:WF? C4:WAVEFORM? Gets the waveform from the instrument.
CFLASH CLEARFLASH
CHDR COMM_HEADER Controls formatting of query responses.
CHDR? COMM_HEADER? Controls formatting of query responses.
CHIP
CHIP?
CHMD CH_MODE
CHS?
CLEARCALI CLEARCALIBRATION
CLEARDELTA
CLK:SAVE
CLK:SET
CLSW COLOR_SWITCH Persist Colour Switch Command (state)
CLSW? COLOR_SWITCH? Persist Colour Switch Command (state)
CM:PAVA? CURR_MEASURE:PARAMETER_VALUE? Returns current parameter, mask test values
CMBN COMBINE
CMBN? COMBINE?
CMPA CURR_MEASURE_STAT
CMR? CMR? Reads and clears the Command error Register (CMR).
CONET COMM_NET Specifies network addresses of scope and printers.
CONET? COMM_NET? Specifies network addresses of scope and printers.
CRMS? CURSOR_MEASURE?
CRSC? CURSOR_MSOURCE?
CRTY? CURSOR_TYPE?
CSVS Saves specified waveform data of CSV format to USB device.
CSVS? Saves specified waveform data of CSV format to USB device.
CUS:H8 CUSTOM:H8
CUS:L8 CUSTOM:L8
CYMT? CYMOMETER? Returns the current cymometer value which displaying on the screen
D0:ACT D0:ACTIVATED
D0:PAVA? D0:PARAMETER_VALUE Returns current parameter, mask test values
D0:TRA D0:TRACE? Enables or disables the display of a trace.
D0:TRSL D0:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D0:TRSL? D0:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D0:WF? D0:WAVEFORM? Gets the waveform from the instrument.
D1:ACT D1:ACTIVATED
D1:PAVA? D1:PARAMETER_VALUE Returns current parameter, mask test values
D1:TRA D1:TRACE? Enables or disables the display of a trace.
D1:TRSL D1:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D1:TRSL? D1:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D1:WF? D1:WAVEFORM? Gets the waveform from the instrument.
D10:ACT D10:ACTIVATED
D10:PAVA? D10:PARAMETER_VALUE Returns current parameter, mask test values
D10:TRA D10:TRACE? Enables or disables the display of a trace.
D10:TRSL D10:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D10:TRSL? D10:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D10:WF? D10:WAVEFORM? Gets the waveform from the instrument.
D11:ACT D11:ACTIVATED
D11:PAVA? D11:PARAMETER_VALUE Returns current parameter, mask test values
D11:TRA D11:TRACE? Enables or disables the display of a trace.
D11:TRSL D11:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D11:TRSL? D11:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D11:WF? D11:WAVEFORM? Gets the waveform from the instrument.
D12:ACT D12:ACTIVATED
D12:PAVA? D12:PARAMETER_VALUE Returns current parameter, mask test values
D12:TRA D12:TRACE? Enables or disables the display of a trace.
D12:TRSL D12:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D12:TRSL? D12:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D12:WF? D12:WAVEFORM? Gets the waveform from the instrument.
D13:ACT D13:ACTIVATED
D13:PAVA? D13:PARAMETER_VALUE Returns current parameter, mask test values
D13:TRA D13:TRACE? Enables or disables the display of a trace.
D13:TRSL D13:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D13:TRSL? D13:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D13:WF? D13:WAVEFORM? Gets the waveform from the instrument.
D14:ACT D14:ACTIVATED
D14:PAVA? D14:PARAMETER_VALUE Returns current parameter, mask test values
D14:TRA D14:TRACE? Enables or disables the display of a trace.
D14:TRSL D14:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D14:TRSL? D14:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D14:WF? D14:WAVEFORM? Gets the waveform from the instrument.
D15:ACT D15:ACTIVATED
D15:PAVA? D15:PARAMETER_VALUE Returns current parameter, mask test values
D15:TRA D15:TRACE? Enables or disables the display of a trace.
D15:TRSL D15:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D15:TRSL? D15:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D15:WF? D15:WAVEFORM? Gets the waveform from the instrument.
D2:ACT D2:ACTIVATED
D2:PAVA? D2:PARAMETER_VALUE Returns current parameter, mask test values
D2:TRA D2:TRACE? Enables or disables the display of a trace.
D2:TRSL D2:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D2:TRSL? D2:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D2:WF? D2:WAVEFORM? Gets the waveform from the instrument.
D3:ACT D3:ACTIVATED
D3:PAVA? D3:PARAMETER_VALUE Returns current parameter, mask test values
D3:TRA D3:TRACE? Enables or disables the display of a trace.
D3:TRSL D3:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D3:TRSL? D3:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D3:WF? D3:WAVEFORM? Gets the waveform from the instrument.
D4:ACT D4:ACTIVATED
D4:PAVA? D4:PARAMETER_VALUE Returns current parameter, mask test values
D4:TRA D4:TRACE? Enables or disables the display of a trace.
D4:TRSL D4:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D4:TRSL? D4:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D4:WF? D4:WAVEFORM? Gets the waveform from the instrument.
D5:ACT D5:ACTIVATED
D5:PAVA? D5:PARAMETER_VALUE Returns current parameter, mask test values
D5:TRA D5:TRACE? Enables or disables the display of a trace.
D5:TRSL D5:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D5:TRSL? D5:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D5:WF? D5:WAVEFORM? Gets the waveform from the instrument.
D6:ACT D6:ACTIVATED
D6:PAVA? D6:PARAMETER_VALUE Returns current parameter, mask test values
D6:TRA D6:TRACE? Enables or disables the display of a trace.
D6:TRSL D6:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D6:TRSL? D6:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D6:WF? D6:WAVEFORM? Gets the waveform from the instrument.
D7:ACT D7:ACTIVATED
D7:PAVA? D7:PARAMETER_VALUE Returns current parameter, mask test values
D7:TRA D7:TRACE? Enables or disables the display of a trace.
D7:TRSL D7:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D7:TRSL? D7:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D7:WF? D7:WAVEFORM? Gets the waveform from the instrument.
D8:ACT D8:ACTIVATED
D8:PAVA? D8:PARAMETER_VALUE Returns current parameter, mask test values
D8:TRA D8:TRACE? Enables or disables the display of a trace.
D8:TRSL D8:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D8:TRSL? D8:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D8:WF? D8:WAVEFORM? Gets the waveform from the instrument.
D9:ACT D9:ACTIVATED
D9:PAVA? D9:PARAMETER_VALUE Returns current parameter, mask test values
D9:TRA D9:TRACE? Enables or disables the display of a trace.
D9:TRSL D9:TRIG_SLOPE Sets the trigger slope of the specified trigger source
D9:TRSL? D9:TRIG_SLOPE? Sets the trigger slope of the specified trigger source
D9:WF? D9:WAVEFORM? Gets the waveform from the instrument.
DAC:SAVE
DAC_CODE
DAC_CODE?
DCDI? DeCo_Digital?
DCPA DeCo_Param
DCST DeCo_Status
DCST? DeCo_STatus
DCVL?
DDR Clears the Device Dependent Register (DDR).
DDR? Clears the Device Dependent Register (DDR).
DEF? DEFINE? Specifies math expression for function evaluation
DFKS Default Key
DI:ACQVD? DIGITAL:ACQ_VALID?
DI:ADCM DIGITAL:ADC_MODE
DI:BV? DIGITAL:BOM_VERSION?
DI:CALM DIGITAL:CALI_MODE
DI:CH CHANNEL_HEIGHT
DI:CMF? DIGITAL:CMD_FINISHED?
DI:DES DIGITAL:DESKEW
DI:DTAN DIGITAL:DATA_ANALY
DI:DTANF? DIGITAL:DATA_ANALY_FLAG?
DI:ECLAI DIGITAL:ECLAI_DATA
DI:EXTV DIGITAL:EXTV
DI:FR? DIGITAL:FREQ?
DI:FV? DIGITAL:FPGA_VERSION?
DI:H8 DIGITAL:HIGH8_SWITCH
DI:HISM? DIGITAL:HIST_MODE
DI:IDTANF DIGITAL:INT_DATA_ANALY_FLAG
DI:KC DIGITAL:KNOB_CONTROL
DI:L8 DIGITAL:LOW8_SWICHT
DI:LCALI? DIGITAL:LCALI_DATA?
DI:LTV? DIGITAL:LOCAL_VALID?
DI:MEAST? DIGITAL_MEASURE_STATUS
DI:MSTM DIGITAL:MASTER_MODE
DI:MTHRED DIGITAL:MTHRED
DI:MTHRED? DIGITAL:MTHRED?
DI:PRBW DIGITAL:PRO_BWL
DI:PV? DIGITAL:PCB_VERSION?
DI:RDDR? DIGITAL:RESET_DDR
DI:RDON DIGITAL:READ_DATA_ONCE
DI:SARA? DIGITAL:SAMPLE_RATE? The SARA? query returns the sample rate of the scope.
DI:SKEW DIGITAL:ADC_SKEW
DI:SLFM DIGITAL:SELF_MODE
DI:SRLN DIGITAL:SERIAL_NUMBER
DI:SRLN? DIGITAL:SERIAL_NUMBER?
DI:STOP DIGITAL:STOP
DI:STPA? DIGITAL:STOP_ACQ?
DI:STTA? DIGITAL:START_ACQ?
DI:SV? DIGITAL:SOFTWARE_VERSION?
DI:SW DIGITAL:SWITCH The SWITCH command is used to set the state of digital.
DI:SW? DIGITAL:SWITCH? The SWITCH command is used to set the state of digital.
DI:SWID? DIGITAL_SWITCH_INDEX?
DI:TDIVF DIGITAL:TDIV_FLAG
DI:THAD DIGITAL:THREADJ
DI:THAD? DIGITAL:THREADJ?
DI:TRCP DIGITAL:TRIG_COUPL
DI:UPIN DIGITAL:UPDATE_PREVIEW_INFO
DI:UTC DIGITAL:USBTMC_CONNECTED
DI:WF? DIGITAL:WAVEFORM? Gets the waveform from the instrument.
DI:WSR? DIGITAL:WAIT_STOP?
DMESg?
DSCL DISPLAY_CLEAR
DTJN DOT_JOIN Controls the interpolation lines between data points.
DTJN? DOT_JOIN? Controls the interpolation lines between data points.
DUTYCALI DUTYCALIBRATION
EMOD EDUMODE
EMOD? EDUMODE?
EQN
EX:TRCP EX:TRIG_COUPLING
EX:TRCP? EX:TRIG_COUPLING?
EX:TRLV EX:TRIG_LEVEL
EX:TRLV? EX:TRIG_LEVEL?
EX:TRLV2 EX:TRIG_LEVEL2
EX:TRLV2? EX:TRIG_LEVEL2?
EX:TRSL EX:TRIG_SLOPE
EX:TRSL? EX:TRIG_SLOPE?
EX5:TRCP EX5:TRIG_COUPLING
EX5:TRCP? EX5:TRIG_COUPLING?
EX5:TRLV EX5:TRIG_LEVEL
EX5:TRLV? EX5:TRIG_LEVEL?
EX5:TRLV2 EX5:TRIG_LEVEL2
EX5:TRLV2? EX5:TRIG_LEVEL2?
EX5:TRSL EX5:TRIG_SLOPE
EX5:TRSL? EX5:TRIG_SLOPE?
EXDL EXT_TRIGDELAY
EXDL:SAVE
EXLV EXT_TRIGLEVEL
EXR Reads, clears the Execution error Register (EXR).
EXR? Reads, clears the Execution error Register (EXR).
EXT:SAVE
EXTCT:SAVE
FCNT FREQCOUNTER Sets or gets frequency counter parameters.
FCNT? FREQCOUNTER? Sets or gets frequency counter parameters.
FCNTCALI FREQCOUNTERCALIBRATION
FDBKAJT FEEDBACKAJUST
FFTA FFT_AUTOSET
FFTC FFT_CENTER The FFT_CENTER command sets the center frequency when FFT (Fast Fourier Transform) is selected.
FFTC? FFT_CENTER?
FFTD FFT_DISPLAY
FFTD? FFT_DISPLAY?
FFTF FFT_FULLSCREEN Enables or disables to display the FFT trace full screen.
FFTF? FFT_FULLSCREEN? Enables or disables to display the FFT trace full screen.
FFTP FFT_POS The FFT_POSITION command sets the vertical offset of FFT waveform
FFTP? FFT_POS? The FFT_POSITION command sets the vertical offset of FFT waveform
FFTS FFT_SCALE Selects the vertical scale of FFT trace.
FFTS? FFT_SCALE? Selects the vertical scale of FFT trace.
FFTT FFT_TDIV The FFT_TDIV? query returns current horizontal scale of FFT waveform.
FFTT? FFT_TDIV? The FFT_TDIV? query returns current horizontal scale of FFT waveform.
FFTU FFT_UNIT The FFT_UNIT command sets the vertical scale type of FFT
FFTU? FFT_UNIT? The FFT_UNIT command sets the vertical scale type of FFT
FFTW FFT_WINDOW Selects the window of FFT.
FFTW? FFT_WINDOW? Selects the window of FFT.
FPAR? FRAME_PARAM? Get frame param.
FRAM? FRAME_SET? Get frame param.
FRSPAJT FREQRESPAJUST
FTIM? FRAME_TIME?
GRDS GRID_DISPLAY Selects the type of grid
GRDS? GRID_DISPLAY? Selects the type of grid
H8:CUS H8:CUSTOM The CUSTOM command sets the threshold value by customer for the specified group of channels
H8:CUS? H8:CUSTOM? The CUSTOM command sets the threshold value by customer for the specified group of channels
H8:TSM H8:THRESHOLD_MODE The THRESHOLD_MODE command sets the threshold type for the specified group of channels.
H8:TSM? H8:THRESHOLD_MODE? The THRESHOLD_MODE command sets the threshold type for the specified group of channels.
HARM?
HMAG HOR_MAGNIFY Horizontally expands the selected expansion trace.
HMAG? HOR_MAGNIFY? Horizontally expands the selected expansion trace.
HPOS HOR_POSITION Horizontally positions intensified zone‘s center.
HPOS? HOR_POSITION? Horizontally positions intensified zone‘s center.
HSIN? HISTORY_INTERVAL?
HSLST? HISTORY_LIST? The HISTORY_LIST command is used to set the state of history list.
HSMD? HISTORY_MODE? The HISTORY_MODE command is used to set the state of history mode.
HSR_BOTEST?
HSR_DAC:SET
HSR_EXTCT:SAVE HSR_EXTCENTER:SAVE
HSR_EXTDT:SAVE HSR_EXTDELAY:SAVE
HSST? HISTORY_STATE?
HSTA HOR_STATE
HSTA? HOR_STATE?
IDN-SGLT-AWG?
IDN-SGLT-PRI?
INNEr:PRODuction:MODE
INR Reads, clears INternal state change Register (INR).
INR? Reads, clears INternal state change Register (INR).
INTS INTENSITY Sets the grid or trace/text intensity level.
INTS? INTENSITY? Sets the grid or trace/text intensity level.
INVT INVERT
INVT? INVERT?
K7DUMP                        Related to the program memory for the main FPGA, unsure how its used.
K7DUMP?
L8:CUS L8:CUSTOM
L8:CUS? L8:CUSTOM?
L8:TSM L8:THRESHOLD_MODE
L8:TSM? L8:THRESHOLD_MODE?
LAGG LANGUAGE
LAGG? LANGUAGE?
LANG LANGUAGE
LANG? LANGUAGE?
LCISL LICENCE_INSTALL
LCISL? LICENCE_INSTALL?
LED?
LICENSE
LICENSE?
LICENSE_ID?
LICENSE_KEY
LICENSE_PR?
LINE:TRCP LINE:TRIG_COUPLING
LINE:TRCP? LINE:TRIG_COUPLING?
LINE:TRSL LINE:TRIG_SLOPE
LINE:TRSL? LINE:TRIG_SLOPE?
LOAD:CALI:FILE?
MATH:CRVA? MATH:CURSOR_VALUE? Returns trace values measured by specified cursors.
MATH:INVS MATH:INVERT_SET Invert the trace or the math waveform of specified source.
MATH:INVS? MATH:INVERT_SET? Invert the trace or the math waveform of specified source.
MATH:PAVA? MATH:PARAMETER_VALUE Returns current parameter, mask test values
MATH:WF? MATH:WAVEFORM? Gets the waveform from the instrument.
MCBD? MACHINE_BAND?
MD5_PR?
MD5_SRLN?
MDID MATH_DIFF_DX
MDWV? MODULATEWAVE?
MEASTAT MEAASURE_STATUS
MENU Enables or disables to display the current menu
MENU? Enables or disables to display the current menu
MMEMORY:UPLoad?
MODE?
MSIZ MEMORY_SIZE Returns the maximal memory size
MSIZ? MEMORY_SIZE? Returns the maximal memory size
MSSTAT MEASURE_SOURCE_STAT
MTEN MATH_ENABLE
MTEN? MATH_ENABLE?
MTVD MATH_VERT_DIV Controls the vertical sensitivity of math waveform of specified source.
MTVD? MATH_VERT_DIV? Controls the vertical sensitivity of math waveform of specified source.
MTVP MATH_VERT_POS Controls the vertical position of math waveform of specified source.
MTVP? MATH_VERT_POS Controls the vertical position of math waveform of specified source.
NBFM NUMBER_FORMAT
NBFM? NUMBER_FORMAT?
NET_BRIDGE_ADDR?
OUTP OUTPUT
OUTP? OUTPUT?
PACL PARAMETER_CLEAR Clears all current parameters in Custom, Pass/Fail.
PACP PARACOPY
PACU PARAMETER_CUSTOM Controls parameters with customizable qualifiers.
PASTAT Parameter Statistics switch
PASTAT? Parameter Statistics switch
PECL PERSIST_CLEAR
PESU PERSIST_SETUP Selects display persistence duration.
PESU? PERSIST_SETUP? Selects display persistence duration.
PFBF PF_BUFFER The PF_BUFFER command sets the output mode when the test fails.
PFBF? PF_BUFFER? The PF_BUFFER command sets the output mode when the test fails.
PFCM PF_CREATEM Creates the mask of the pass/fail.
PFDD? PF_DATADIS? Return the number of the pass/fail monitor which can be displayed on the screen
PFDS PF_DISPLAY Enables or disables to display the test and the message options of pass/fail
PFDS? PF_DISPLAY? Enables or disables to display the test and the message options of pass/fail
PFEN PF_ENABLE The PF_ENABLE command enables or disables the Pass/Fail test features.
PFEN? PF_ENABLE? The PF_ENABLE command enables or disables the Pass/Fail test features.
PFFS PF_FAIL_STOP The PF_FAIL_STOP command sets the switch of the ―stop on fail‖ function
PFFS? PF_FAIL_STOP? The PF_FAIL_STOP command sets the switch of the ―stop on fail‖ function
PFOP PF_OPERATION The PF_OPERATION command controls to run or stop Pass/Fail test
PFOP? PF_OPERATION? The PF_OPERATION command controls to run or stop Pass/Fail test
PFSC PF_SOURCE The PF_SOURCE command sets measurement sources for Pass/Fail test.
PFSC? PF_SOURCE? The PF_SOURCE command sets measurement sources for Pass/Fail test.
PFST PS_SET Sets the X mask and the Y mask
PFST? PS_SET? Sets the X mask and the Y mask
PNSU? PANEL_SETUP? Complements the *SAV/*RST commands.
PRBD? PRODUCT_BAND?
PRINTCALI PRINTCALIBRATION
PRO_SRLN?
PROD PRODUCT
PROD? PRODUCT?
RCPN RECALL_PANEL Recalls a front-panel setup from mass storage.
REBOOT
REFA:PAVA? REFA:PARAMETER_VALUE Returns current parameter, mask test values
REFA:REFSC? REFA:REF_SCALE?
REFB:PAVA? REFB:PARAMETER_VALUE Returns current parameter, mask test values
REFC:PAVA? REFC:PARAMETER_VALUE Returns current parameter, mask test values
REFCL REF_CLOSE The REF_CLOSE command closes the Reference function.
REFD:PAVA? REFD:PARAMETER_VALUE Returns current parameter, mask test values
REFDS REF_DISPLAY The REF_DISPLAY command enables or disables the current reference channel shown on the screen.
REFDS? REF_DISPLAY? The REF_DISPLAY command enables or disables the current reference channel shown on the screen.
REFLA REF_LOCATION The REF_LOCATION command selects the current reference channel.
REFLA? REF_LOCATION? The REF_LOCATION command selects the current reference channel.
REFPO REF_POSITION The REF_POSITION command sets the vertical offset of the current reference channel.
REFPO? REF_POSITION? The REF_POSITION command sets the vertical offset of the current reference channel.
REFSA REF_SAVE The REF_SAVE command saves the waveform (screen range) of the specified source as the reference waveform
REFSC REF_SCALE The REF_SCALE command sets the vertical scale of the current reference channel
REFSR REF_SOURCE The REF_SOURCE command sets the reference waveform source.
REFSR? REF_SOURCE? The REF_SOURCE command sets the reference waveform source.
RESETADC
RFLASH READFLASH
RFLASHBLK READFLASHBLOCK
ROLL
ROSC ROSCILLATOR Sets or gets the clock source.
ROSC? ROSCILLATOR? Sets or gets the clock source.
RPSC REGISTER_PRODUCT_SCPI_CMD
RPSC? REGISTER_PRODUCT_SCPI_CMD?
SANU? Sample_Number? Return the number of sampled points available from last acquisition and the trigger position
SARA? Sample_Rate? Return the sample rate of the scope
SAST SAMPLE_STATUS Return the acquisition status of the scope
SAST? SAMPLE_STATUS? Return the acquisition status of the scope
SAVECALI
SCDP Causes a screen dump to controller.
SCFG SYSTEMCONFIG Set or get power on initializing parameter way.
SCFG? SYSTEMCONFIG? Set or get power on initializing parameter way.
SCOPEID? SCOPE_ID?
SCSV SCREEN_SAVE Controls the automatic screen saver.
SCSV? SCREEN_SAVE? Controls the automatic screen saver.
SELFCALI SELFCALIBRATION
SET50
SGLT-TEST-VAST-DATA
SGLT-TEST-VAST-DATA?
SGLT-UPGRADE
SHELLCMD
SRLN
SRLN?
SRLN2MAC
SRLN-AWG
STAT:OPER:COND? STATus:OPERation:CONDition?
STAT:OPER:ENAB STATus:OPERation:ENABle
STAT:OPER[:EVEN]? STATus:OPERation[:EVENt]?
STAT:PRES STATus:PRESet
STAT:QUES:COND? STATus:QUEStionable:CONDition?
STAT:QUES:ENAB STATus:QUEStionable:ENABle
STAT:QUES[:EVEN]? STATus:QUEStionable[:EVENt]?
STL? STORELIST?
STPN STORE_PANEL Stores front-panel setup to mass storage.
SAVEAJUSTDATA
SVNALL?
SWWV? SWEEPWAVE?
SXSA SinXx_Sample Sets the type of the interpolation
SXSA? SinXx_Sample? Sets the type of the interpolation
SY_FP
SYNC
SYNC?
SYST:COMM:LAN:GAT SYSTem:COMMunicate:LAN:GATeway Set the gateway for the signal generator in the network. The gateway will be fetched automatically if the IP assignment is set to DHCP.
SYST:COMM:LAN:GAT? SYSTem:COMMunicate:LAN:GATeway? Set the gateway for the signal generator in the network. The gateway will be fetched automatically if the IP assignment is set to DHCP.
SYST:COMM:LAN:IPAD SYSTem:COMMunicate:LAN:IPADdress Set a host name for the signal generator in network
SYST:COMM:LAN:IPAD? SYSTem:COMMunicate:LAN:IPADdress? Set a host name for the signal generator in network
SYST:COMM:LAN:MAC? SYSTem:COMMunicate:LAN:MAC?
SYST:COMM:LAN:SMAS SYSTem:COMMunicate:LAN:SMASk Set the subnet mask according to the PC network Settings. The subnet mask will be set automatically if the IP assignment is set to DHCP.
SYST:COMM:LAN:SMAS? SYSTem:COMMunicate:LAN:SMASk? Set the subnet mask according to the PC network Settings. The subnet mask will be set automatically if the IP assignment is set to DHCP.
SYST:COMM:LAN:TYPE SYSTem:COMMunicate:LAN:TYPE Toggles the IP assignment Setting between static (manual) and DHCP (dynamic assignment) mode.
SYST:COMM:LAN:TYPE? SYSTem:COMMunicate:LAN:TYPE? Toggles the IP assignment Setting between static (manual) and DHCP (dynamic assignment) mode.
SYST:COUN? SYSTem:COUNt? System power on count
SYST:DATE SYSTem:DATE
SYST:DATE? SYSTem:DATE?
SYST:ERR? SYSTtem:ERRor?
SYST:FDEF SYSTem:FDEFault
SYST:PON:TYPE SYSTem:PON:TYPE Uses command to set signal generator to power on in default, last. Get power on type.
SYST:PON:TYPE? SYSTem:PON:TYPE? Uses command to set signal generator to power on in default, last. Get power on type.
SYST:PRES SYSTem:PRESet Presets all parameters.
SYST:PRES:TYPE SYSTem:PRESet:TYPE Set preset file when preset type is user
SYST:PRES:TYPE? SYSTem:PRESet:TYPE? Set preset file when preset type is user
SYST:TIME SYSTem:TIME
SYST:TIME? SYSTem:TIME?
SYST:VERS? SYSTem:VERSion?
SYSTem:FileSystem:Type? SYSTem:FileSystem:Type?
TDIV TIME_DIV Modifies the time base setting.
TDIV? TIME_DIV? Modifies the time base setting.
TMPL TEMPLATE Produces a complete waveform template copy.
TMPL? TEMPLATE? Produces a complete waveform template copy.
TRCAN TRIG_CAN
TRCAN? TRIG_CAN?
TRCP TRIG_COUPLING
TRCP? TRIG_COUPLING?
TRDL TRig_DeLay Sets the time at which the trigger is to occur.
TRDL? TRig_DeLay? Sets the time at which the trigger is to occur.
TRDP? TRIG_DROP?
TREG? ACQ1
TREG? ACQ2
TREG? MAP
TRFSM TRIG_FSM
TRFSM? TRIG_FSM?
TRIIC TRIG_IIC
TRIIC? TRIG_IIC?
TRLIN TRIG_LIN
TRLIN? TRIG_LIN?
TRLV TRIG_LEVEL
TRLV? TRIG_LEVEL?
TRMD TRIG_MODE The trigger mode.
TRMD? TRIG_MODE? The trigger mode.
TRNR TRIG_NOISEREJ
TRNR? TRIG_NOISEREJ?
TRPA TRIG_PATTERN Sets the condition of the pattern trigger
TRPA? TRIG_PATTERN? Sets the condition of the pattern trigger
TRPH TRIG_PATTERN_HOLDOFF
TRPH? TRIG_PATTERN_HOLDOFF?
TRPH?
TRPT TRIG_PATTERN_LIMIT
TRPT? TRIG_PATTERN_LIMIT?
TRSE TRIG_SELECT Selects the condition that will trigger acquisition.
TRSE? TRIG_SELECT? Selects the condition that will trigger acquisition.
TRSPI TRIG_SPI
TRSPI? TRIG_SPI?
TRSR TRIG_SOURCE Trigger Source
TRSR? TRIG_SOURCE? Trigger Source
TRST TRIG_STATUS
TRST? TRIG_STATUS?
TRTY TRIG_TYPE
TRUART TRIG_UART
TRUART? TRIG_UART?
TRWI TRIG_WINDOW Return relative height of the trigger window
TRWI? TRIG_WINDOW? Return relative height of the trigger window
USBDATA?
VECD VECTORS_DISPLAY
VECD? VECTORS_DISPLAY?
VKEY VIRTUALKEY Set the virtual key.
VOLTPRT? Sets or gets state of over-voltage protection
WEB:PSW
WEB:PSW?
WF WAVEFORM
WFDA? WF_DATA? saves waveform data to user folder on device
WFLASHBLK WRITEFLASHBLOCK
WFSU WAVEFORM_SETUP Specifies amount of waveform data to go to controller
WFSU? WAVEFORM_SETUP? Specifies amount of waveform data to go to controller
WRFLASHBLK WRITEREADFLASHBLOCK
WVDA WAVEDATA Get the wave data of store .
WVPR WAVEPARAMETER The WAVE_PARA? query returns the location, name, frequency, amplitude, and offset of four arbitrary waveforms.
WVPR? WAVEPARAMETER? The WAVE_PARA? query returns the location, name, frequency, amplitude, and offset of four arbitrary waveforms.
XYDS XY_DiSplay Enables or disables to display the XY format
XYDS? XY_DiSplay? Enables or disables to display the XY format
Z1:PAVA? Z1:PARAMETER_VALUE Returns current parameter, mask test values
Z2:PAVA? Z2:PARAMETER_VALUE Returns current parameter, mask test values
Z3:PAVA? Z3:PARAMETER_VALUE Returns current parameter, mask test values
Z4:PAVA? Z4:PARAMETER_VALUE Returns current parameter, mask test values



« Last Edit: December 13, 2018, 11:20:38 am by Rerouter »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28387
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1092 on: December 09, 2018, 01:16:21 am »

And sadly no sign of bode plot commands for me.

Just 3 that pop out for me:
ROSC   ROSCILLATOR   Sets or gets the clock source.
WF   WAVEFORM
SWWV?   SWEEPWAVE?
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline Rerouter

  • Super Contributor
  • ***
  • Posts: 4694
  • Country: au
  • Question Everything... Except This Statement
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1093 on: December 09, 2018, 01:46:37 am »
ROSC is the signal generators clock source, either the internal oscillator, or I would imagine aux port on the SAG1021,

Waveform is the command to export a raw export of a given channel, I am working with EEZstudio to make it directly export to CSV, in reality there is no reason why the webserver could not do this either.

SWWV is one of the operating mode of the signal generator option, in reality the devices bode plot works off BSWV (BasicWave), sets the frequency, measures then sets the next.


What i was after was a command that would put it in a bode plot mode with the parameters exposed so I could set the center, span, etc via scpi,
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28387
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1094 on: December 09, 2018, 03:36:58 am »
ROSC is the signal generators clock source, either the internal oscillator, or I would imagine aux port on the SAG1021,

Waveform is the command to export a raw export of a given channel, I am working with EEZstudio to make it directly export to CSV, in reality there is no reason why the webserver could not do this either.

SWWV is one of the operating mode of the signal generator option, in reality the devices bode plot works off BSWV (BasicWave), sets the frequency, measures then sets the next.
OK.


Quote
What i was after was a command that would put it in a bode plot mode with the parameters exposed so I could set the center, span, etc via scpi,
Ah, gotcha.
So if we look at the SCPI command set from the US website, it's from 2017 and just before SDS1004X-E models were released, version RC01-E02A
https://www.siglentamerica.com/wp-content/uploads/dlm_uploads/2017/10/Programming-Guide-1.pdf

Yet on the Shenzhen site, a later version RC01020-E01C
http://old.siglentamerica.com/USA_website_2014/Documents/Program_Material/Programming_Guide_forSDS1000X-E.pdf
Overlook the file name: WaveAce Remote Control  ;)

Neither of which have what you're looking for.  :(  ::)

They must be still building it.  :-//
I'll ask.


EDIT
Just found another:
http://old.siglentamerica.com/USA_website_2014/Documents/Program_Material/Programming%20Guide.pdf
Still without what you seek.

I'll unbox one and see if there's any different versions on the CD......none at all !  :o
500+ MB free and no Programming guide ?  |O

Siglent, please add the SCPI command programming guide onto the CD.....at least the latest version available at the time of printing......with an embedded link to a online source of any later versions.
« Last Edit: December 09, 2018, 04:03:19 am by tautech »
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline Rerouter

  • Super Contributor
  • ***
  • Posts: 4694
  • Country: au
  • Question Everything... Except This Statement
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1095 on: December 09, 2018, 03:54:02 am »
While your at it, a workaround I was trying also fell short, The "VKEY" (VIRTUALKEY) command does not work, With this I could have essentially typed in the keys to accomplish it in a roundabout way.

Thanks as always tautech.

Page187-188
https://www.circuitspecialists.com/content/238573/sdg810_manual_+_programming_guide_+_quickstart_guide.pdf

Code: [Select]
28 KB_FUNC1
23 KB_FUNC2
18 KB_FUNC3
13 KB_FUNC4
8 KB_FUNC5
3 KB_FUNC6
48 KB_NUMBER_0
49 KB_NUMBER_1
50 KB_NUMBER_2
51 KB_NUMBER_3
52 KB_NUMBER_4
53 KB_NUMBER_5
54 KB_NUMBER_6
55 KB_NUMBER_7
56 KB_NUMBER_8
57 KB_NUMBER_9
46 KB_POINT
43 KB_NEGATIVE
15 KB_MOD
16 KB_SWEEP
17 KB_BURST
5 KB_PARAMETER
11 KB_UTILITY
70 KB_STORE_RECALL
4 KB_WAVES
72 KB_CHANNEL
177 KB_KNOB_LEFT
175 KB_KNOB_RIGHT
176 KB_KNOB_DOWN
44 KB_LEFT
40 KB_RIGHT
153 KB_OUTPUT1
152 KB_OUTPUT2
 
The following users thanked this post: tv84

Online tautech

  • Super Contributor
  • ***
  • Posts: 28387
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1096 on: December 09, 2018, 04:06:38 am »
While your at it, a workaround I was trying also fell short, The "VKEY" (VIRTUALKEY) command does not work, With this I could have essentially typed in the keys to accomplish it in a roundabout way.
Can I ask you to document it and drop it into this thread please:
https://www.eevblog.com/forum/testgear/the-sdg1000-and-sdg800-thread/
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline Rerouter

  • Super Contributor
  • ***
  • Posts: 4694
  • Country: au
  • Question Everything... Except This Statement
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1097 on: December 09, 2018, 04:40:18 am »
I don't actually own a sdg810, it was just the only manual I could find with an example string for the command.

If you still wish me to write it up there, then I shall.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28387
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1098 on: December 09, 2018, 04:51:03 am »
I don't actually own a sdg810, it was just the only manual I could find with an example string for the command.

If you still wish me to write it up there, then I shall.
OK, never mind then as it's best to have bonafide reports from those that actually have such equipment just in case tech support needs further info from the owner.

Back on Bode plot, is any of the code here of use to you ?
https://www.eevblog.com/forum/testgear/siglent-sds1104x-e-and-sds1204x-e-bode-plot-with-non-siglent-awg/
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline Rerouter

  • Super Contributor
  • ***
  • Posts: 4694
  • Country: au
  • Question Everything... Except This Statement
Re: Siglent SDS1104X-E and SDS1204X-E Mixed Signal Oscilloscopes
« Reply #1099 on: December 09, 2018, 06:39:54 am »
What that code does is appear on the network as a siglent signal generator, the scope then sends it commands as it would to an actual device, the PC then converts these commands and spits it back out to the other brand device in the correct format for that device, over USB or ethernet.

The scope pretty much sends commands like "Generate a 10Hz Wave", measure for 1 acquisition period, "Generate a 12Hz Wave, measure... etc,
In the video this is the "C1:BSWV FREQ,10" line where you see the number counting up. I "think" the device then grabs the same value as "C1:PAVA? PKPK" to plot amplitude, and C1-C2:MEAD? to plot phase. 

I've been playing with the SAG1021 to better learn its limits, And at this point I am starting to think up how I might script it, but would still be nice to have commands for it
« Last Edit: December 09, 2018, 06:43:31 am by Rerouter »
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf