I ordered a SDS1104X-E from Batronix
After you get it it is good to know least one thing, lest with this FW version what is there now. This version is 7.6.1.12 out from package.
If any channel is set for < 5mV/div (1X selected) then first 30 minutes after start, it need note that it do some kind of partial small auto"cal" every 15 second. It can see on the screen as some kind of short pause/"hick-up". Or example if scope is in Roll mode, every 15 second it restart rolling trace. Or if acquisition mode is average, this average resets etc.
If all channels are 5mV/div or more this do not happen at all. After 30min continuous power on, this process end and 500uV to 2mV/div works without this "autocal" or whgat ever it is.
If user do not know at all about this, it can cause real confusion.Btw, this can see also in EEVblog (Dave) made nice tear down video final part where he is playing bit fun with scope. This sudden display cleaning (position 18:57) is just one symptom about this process. It happen because scope was continuously running less than 30 minutes and channel V/div was under 5mV/div.
I think that Siglent might later change this approach. Either by doing it differently or by giving the user the choice to choose whether the oscilloscope performs internal self-tuning during time what is typical for temperature equilibrium (
>30min)