Hello all,
A screenshot of two instruments running. It was straightforward to get a 34465A added as instrument #2.
Thanks again Ian. This is a valuable asset to my home lab.
I have ordered the temperature/humidity sensor.
Hi all,
V1.23 available in post #1.
I have identified a problem reading the .CSV into the Playback Chart.........i.e. it does work, but some CSV's are dual device logs, some are single devices logs.......and it can trip up the code if you don't select properly.....so I need to re-write the sub that loads in the CSV so that it always gets it right.
I have started to do this but it's not finished so for now just select careful.
Ian.
Hi Ian,
The Autoscale is still troublesome in the main program.
If i configure my 34461A as device-2 alone, it wil never work...
And is it posible to load the settings for device-1 and device-2 via a config file like your notepad button but then for 1 instrument.
so i wil have in de directory where your program is, say 10 different cdevice files, like this: 34461a-DC10V-100PLC.conf, 3458A-1V-10PLC, etc.
Than you can quickly chose device settings for a instrument.
Have a nice weekend, and i do not mean a coding weekend
Kind regards,
Bram
I have searched myself crazy for my 34461A to put it on 100PLC mode via GPIB and I can't find it...
FUNCTION 'VOLT:DC'
VOLT:DC:RANGE 10
VOLT:DC:NPLC 100
Nice thread...quick question...why not use labview?
Might it not be an easier solution?
Sent from my SM-N960F using Tapatalk
Hi Ian,
I can do what i want, but it is not posible for me to get any useful information out of the playback chart.
It doesn't matter if I use one of your .csv or one I made myself...
This is a picture where you can see something at the bottom, I can't make anything else with it.
I have tried many settings but none of them provides a useful piece of software.
What am I doing wrong?
Kind regards,
Bram
1,HP3458A,2018-12-05_21:31:49,0.00000000,24.9,30.8
2,HP3458A,2018-12-05_21:31:49,9.999999730,24.9,30.8
3,HP3458A,2018-12-05_21:32:49,9.999999735,24.9,30.6
4,HP3458A,2018-12-05_21:33:49,9.999998419,25.0,30.6
5,HP3458A,2018-12-05_21:34:49,9.99999779,25.0,30.6
6,HP3458A,2018-12-05_21:35:49,9.999996985,25.1,30.7
7,HP3458A,2018-12-05_21:36:49,9.999997014,25.1,30.8
8,HP3458A,2018-12-05_21:37:49,9.999996883,25.1,30.3
9,HP3458A,2018-12-05_21:38:49,9.999996546,25.0,30.3
10,HP3458A,2018-12-05_21:39:49,9.999995932,25.0,31.0
Hi Ian,
Friday night I had your software work on the 34461A and the 3458a.
And it was the second night that the software got stuck.
The software runs here on a 64 bit Windows -7 computer without unnecessary software.
Meanwhile I have also been able to solve all the remarks and errors from the windows logs.
Only the WMI 10 eror that is in almost every Windows-7 computer i cant kill, but this is not important.
The following screenshots contain some Dutch text, but the important information consists of English text.
I hope these pictures helps to find erors in your software.
Kind regards,
Bram
PS
Your remarks workt, see the picture below.
The wave is the temperature dependency of the instruments and a little of my 10V LAB reference. (draf and switching on and off of the heater in my lab)
The LAB reference will come in a second oven of about 40C and the inner oven is about 43C, but that is a job of the beginning of 2019.