Check the mains voltage and schematic.
My 2020 can be changed to 240V by resoldering the connections to the transformer primary. IIRC the 2005 can't.
I've asked the seller. I DO have a soft spot for the 2005A, so I'm maybe not as steadfast as I ought to be.
It looks like it has been converted
It looks like it has been converted
As for the DM8577, I'm avoiding ancient Nat Semi/TI parts, the Soviet clones should be a bit newer. Strange how the Nat Semi parts on ePay-UK are marked as both DM8577N & DM74188AN if they can't be programmed the same way.![]()
David
I was nearly caught out by that. The ones I have are factory marked DM74188AN DM8577N. I was initially using the TI Databook. I have one around here somewhere and it was easier to find on the web. Mostly because I was curious, I did an extensive search and found the DM8577N in a National book. Much to my horror, they are "Pin Compatible" but very different internally.
They are awful, they even put the 74188 first, wonder how many got wasted by folks programming them as 74188 instead of 8577 over the years.
David
....snip....
I would have made a close-up picture of the screen to prove I am not lying about frequency and signal amplitude. Sadly just before doing that, I tried fiddling with the adapter to see if I could shove it in a bit more to make sure the trace does not go unstable right as I am trying to photograph it and... as I was grabbing the adapter, I got ZAPPED !!! Yes this generator tried to kill me !!!
So no close-up picture for you sorry... instead I immediately pulled the plug !
Checked with the DMM for resistance between the two line terminals in the power socket, and chassis/enclosure.... no short, got 5Mohms and slowly climbing.
So I am not touching this generator for now !
Think I will retrofit an IEC power socket hidden at the back (to not ruin the looks of the face plate), and get myself a life saving earth connection....
Will look at the schematic of the generator to see if there is maybe some failing cap that could explain the mishap...
Catching up still......
Looking at your previous posts on the Metrix 920, it seems to have two capacitors to ground (aka the case) and looking at the back of the mains inlet there are two dubious looking capacitors on the sides of the inductors, I would be amazed if these weren't wax paper deathtraps/smoke bombs.
And the DMM will be of no use here, you need a proper DC insulation tester at 250V or 500V to check them, but probably best to just change them for some class Y rated capacitors anyway.
David
Thanks for doing my work Dave !
Just looked C11 and C12 in the parts list... they are indeed paper caps ! They are 10nF 10%.
Will replace them of course !
Will keep the old caps so I can test them later for fun and educational purposes...
Hell I could cobble a crude tester together easily I guess !
After all these caps are simply subject to mains AC ! So all I need to test them in a relevant manner, is to connect the to mains and measure the leakage current, eh ?
Just need to put a series resistor to limit the current to a safe value, I don't know, less than 500mA, since my Metrix meter has a 500mA current range / fuse.
Will open up the generator in the coming days, if I can sort out this problem right now, why wait... there are enough things to do and fix in this lab for me not to post pone repairs I can do quickly and easily !
I'm suprised no one has said this: YOU NEED TO CHECK THE EARTH CONNECTIONS. Even if the capactors are leaky you would not have got a shock if the units mains earth was intact (all the way back to the supply, it could be a faulty extension lead or badly wired socket).
Even if the capacitors are OK if there is no earth the chassis will be at around 120V AC. This is because the two capacitors act as a AC voltage divider. Not much current due to the low values but enough to feel if you touch it or read on a DMM to ground.
Note that replacement capacitors should be Y type rated for mains use.

Mine is too but there is no service manual for it anywhere ! Contacted Metrix directly and they said "yeah we remember that one, but have no info on it as we outsourced its R&D, we didn't design it ourselves ". Bummer
The service manual for the GX5000 is proving to be very elusive! Metrix were (perhaps still are) partnered with Sefram. Sefram appear to have sold the same unit as Model 4450.
Asking for the unit's ID, via the GPIB, returns "OR-X,GX 5000". I believe Or-X are the original manufacturers. They offer the model 550 (sometimes called FP550) which appears to be the same design.
The Metrix 555 appears to have the same functionality but with an upgraded front-panel. The user manual for the 555 details the GPIB commands - all of which seem to work with the GX5000.


Eh ?! Why not...Have you tried asking Sefram or OR-X directly, for a service manual then ??
Power Designs PD2005A on German ebay:
https://www.ebay.de/itm/255327604587?hash=item3b72b60f6b:g:IjgAAOSwup9gHCiI
Nah, I don't think it's worth it
However I have a new problem now... looks like my Metrix Nixie counter has a problem with its LF / 1Meg input ! Signal is only 10MHz so it can take it no problem. However it would not register / count anything at all, only zeros on the Nixies, even though I have good signal amplitude, 5Vpp unloaded, and no matter how I fiddled with the input "sensitivty" pot/control. Plus, when I connect the counter, I can see on the scope that the signal goes haywire. Big drop in amplitude + it becomes completely out of shape, looking like a wiggly triangle, no more a rectangular pulse at all ! If I use the HF / 50R input instead, all is well.
So, the LF front-end has a problem.... some more repairs in the queue !
This counter is not easy to work on I fear, very compact in there. It's not going to be fun I think...
I would try another signal source, as maybe it can't detect the narrow pulse on the low frequency input? , I know some of the old counters I've got can be a bit fussy on the input signal.
David

(...)
OK I spent a few minutes testing the counter.
Result : the /F input CAN go up to 10MHz no problem... IF the conditions are right.
Let me explain... First I connected the pulse generator only to the scope, 10MHz again. Got a fair bit of ringing. Switched the scope to 50R input... much better, looks decent now (see pic)
Have my baseline now, so I can now connect the counter as well as see what happens.
1) if I connect the HF / 50R input of the counter, it adds a little bit of ringing but minimal, still looks decent and can count just fine. (see pic)
2) Then I switch the counter to its LF / 1Meg input. Oh oh... it adds A LOT of ringing, so much that if the pulse is too narrow, it becomes so distorted that it hardly looks like a square anymore. That's when the counter fails to count. But if you increase the pulse width a little bit, at some point the ripple becomes low enough that the counter manages to pick up the pulses.. but it still looks far from a square (see pic).
So the question is why does the LF input distort the generator output SO much ?!
Of course it adds more capacitance, but one would thing it would dampen the signal not make it "ring" ?
Or... the rsie time is too high, too much high-frequency harmonics, and the capacitive load of the 1Meg input low-passes the signal so much that it kills too much of the spectrum / harmonics... and the little that remains... is not enough to make it look anywhere near a square wave.
SO.... I could make one more experiment : keep the same frequency and pulse wigth, but progressively decrese rise time and see if that make the signal look more and more like a square.
Yes that would be interesting !
, hence why the "ringing" / reflection on the cable / transmission line gets worse!

Yeah but the scope is 50R terminated so I didn't think that putting 1Meg in // with the counter, would make that much of a difference.. 1Meg // with 50R is still pretty much 50R
Played with rise times. That did it, was interesting.
Since the beginning the pusle generator was set to its fastest rise time, 5ns.
So what I did was increase the puls width until the 1Meg input would regsiter. It happened with a pulse width of 27ns.
Then I backed off a little, 26ns, so the counter would not register anymore.
Then I started increasing the rise time (and fall time equally) to see how / if it would have an impact on the shape of the waveform, and it the counter would behave any differently.
That was success.
I increased rise time very progressively, in 0.1ns increments. At 7.2ns the counter started flickering, was picking up some pulse.. and at 7.4ns, it was able to read reliably, was stable.
The shape of the signal however is barely changed to thje naked eye.. but clearly the counter does see the difference !
It was like clock work : redude pulse width by one ns : doesn't count anymore. Increase by 1ns.. hey presto works again. Same with rise time. 7.4ns works solidly, 7.2 it's unstable, 7.0ns no counting at all.
It's fun being to control pulse width and rise times, accurately, and see how the counter reacts while looking at the waveform on the scope
An old school buddy asked me about getting his 80's Marantz ST450 AM/FM tuner fixed. Anyone here know about these old tuners ?
https://youtu.be/31kgOCo2WhA Hmmm... it sure does like to sing while it works, doesn't it...?
mnem
I recently updated the controller board on my 3D printer. It now uses a Creality 4.2.7 board which has TMC2225 motor drivers: https://creality3d.shop/products/creality3d-upgrade-silent-4-2-7-1-1-5-mainboard-for-ender-3-ender-3-pro-ender-5-3d-printer
The printer went from being a "singer" to being incredibly quiet.
McBryce.


Flip side of that is analog tuner feature-set, so not a lot of demand...
Why not 74s188(10.5V) or 82s23(17.5V)?
NS manual for DM8577 is a bit uncertain but 23V is mentioned.
The procedure is for DM8574 and DM8577 pages are stuffed in between.
(169.pdf)
Flip side of that is analog tuner feature-set, so not a lot of demand...The Marantz FM tuners of old aren't bad. If the sound and the ability to dig signals out of noise (while dealing with deafening commercial stations a bit too close) counts for something, they clearly are worth the job...
Flip side of that is analog tuner feature-set, so not a lot of demand...The Marantz FM tuners of old aren't bad. If the sound and the ability to dig signals out of noise (while dealing with deafening commercial stations a bit too close) counts for something, they clearly are worth the job. Here, we've just revived the Model 4230 Quad-receiver my father bought back in 1976, but now in Middle Boy's room, together with the AR-6 loudspeakers that were part of the same original deal, and the Thorens TD166 Mk V I bought in the 90s. Very Classy.
There are a few scale illumination lamps out, and the scale is some 350-500KHz off, pots are a bit scratchy, all as can be expected from a 45yo consumer device, albeit a very high quality one.
I have calculated and made a simple dipole which gives full signal level on all relevant stations, so I'm guessing the FM tuner is basically OK. (but we've got excellent RF reception here, on a hill, practically line-of-sight to the main Nacka transmitter. OTOH, I can get the regional FM stations from all around lake Mälaren with my circular dipole that's feeding the rest of the FM reception units at home)
Service manual has been downloaded, and schematic of course was supplied with the unit so I now have a few reasons to learn to use all the measurement things I have.
Well, no... my point was not that they were a bad tuner; more that the street value of a working ST450 means one can't afford to spend a LOT of time trying to resurrect one that's got major problems. Better to shotgun recap a working one and cut your losses.